BF-type applied part MOPP vs secondary
Hello everyone,
I am designing a medical device with BF-type applied part (surgical power tool with micromotor).
The applied part is so small that I cannot implement more than 1 MOPP (working voltage of 48V) inside its metallic body (barrier C in attached picture).
The power supply is medical grade with BF class insulation. It provides 2 MOPP (4000VAC) from input to output (barrier A) and 1 MOPP (1500VAC) from output to ground (barrier D).
The applied part is driven by a 48V floating circuit with creepage distances between GND and protective earth for 1 MOPP at main voltage (barrier E).
To me, the design above respects clause 8.5.1.1 on the applied part for 2 MOPP at 48V with C + E.
The clause 8.5.2.1 for 1 MOPP at mains voltage is also respected with E.
But an expert asked for 2 MOPP at 48V between the applied part metal body and the secondary voltage (barrier B).
Wouldn't be C + E or C + D sufficient ?
P.S. : Patient leakage current limits are respected in Normal Condition and Single Fault Condition (even with SFC = shorted barrier C).
Hello again,
A short update on my issue. The expert statement is that MOPP C + E (or C + D) cannot be considered together as they are not rated for the same working voltage.
I disagree as barrier D or E are better than C. Patient safety is even better no ?
Peter Selvey
The expert might be used to dealing earthed voltage sources. Most sources are earth referenced, so it's easy to fall into the trap of applying this in all cases.
If the voltage is earthed, you need 2 MOP to prevent currents exceeding the limits from flowing. This is because the operator or patient is assumed to be earthed, providing a return path.
However, if the voltage source is genuinely floating from earth, then you only need 1 MOP, because, well, .... you know ... current always needs two points of contact to flow. Thus, a single fault to a floating circuit doesn't present any risk. Just like a bird happily resting on an 11kV line.
A caveat though is to make sure the circuit is genuinely floating, which can be verified by test. This would involve shorting insulation C and measuring the current. In this case, insulation E is actually forming the second MOP by preventing current from flowing. Note that circuits are rarely 100% floating as they are typically EMC or other capacitors to earth which bridge insulation E. While these may be isolating at dc, there is usually some impedance at 50/60Hz and could even be low impedance at the switching frequencies often found inside secondary circuits. Nevertheless, with C shorted the currents are usually still below the limits.
The argument that they are "not rated for the same working voltage" seems to be based on clause 8.5.4 2nd dash, the working voltage for each MOP in a 2MOP system should be the same as for the whole system. This is correct, but it applies for each voltage source individually. In a device there may be several voltage sources, and a single barrier can play insulating roles for two or more sources. The MOP analysis (insulation diagram) should, In principle, be done for each voltage source individually, rather than lumped together in a single diagram. In this case, the analysis for the 48V source would find that both C and E need to be each assessed based on (the same) working voltage of 48V.
Separately, for the purpose of F-type applied parts (a hypothetical single fault situation that the patient is raised to mains voltage), isolation E needs to be 1MOP for a working voltage of 230V (mains voltage). It's a difference source, so different analysis.
Then, if a single MOP is involved in multiple voltage sources (as is the case for E), we can of course test it only for the most strict requirements. It makes no sense to do a 1500Vrms test for the 230V and then a 500Vdc test for the 48Vdc. Same for creep/clearance. Note however that there are some case where a higher working voltage does not necessarily mean stricter requirements.
D Separation of F-type applied part and remaining parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B F-type applied part - Separation from ALL(?) other parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
K F-TYPE APPLIED PART IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Number of MOPP for Type BF Applied Part (Medical Device Electronics) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Is insulation between F type applied part and other parts a must? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Insulation requirements between F-type applied part and enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Does applied part of battery powered TENS unit need to be designed as type BF? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
R Defining the type of Applied Part - Metal Probe (Applied Part) employs Water Cooling IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Sidney Vianna Outsourced Processes - Type and Extent of Control to be Applied - ISO 9001:2008 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 40
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
W What is the difference between TYPE B and TYPE BF? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Nihls MSA Study Type 1 (CMM) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
A Fabric roll inspection - What type of Control Chart to use? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
A AMSQQP416E Class 2 type 2 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
S Are EC type examinations still being conducted under MDR? EU Medical Device Regulations 5
S Medical watch Class II (AP Type CF) with USB connection IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Johnnymo62 Non Aerospace topics - Anything for military trucks, trailers, Humvee type vehicles? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
Marcel DS GOST-R (Type approval for Vehicles in Russia) Other ISO and International Standards and European Regulations 0
S Type 1 Gage R&R or something else? Reliability Analysis - Predictions, Testing and Standards 6
I Old Time Scatter diagrams for defect type and location- software Quality Tools, Improvement and Analysis 7
P Surgical masks ( type IIR) - Cleanroom Requirements EU Medical Device Regulations 3
W Insulation requirements of BF type infant skin temperature probe IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
Jane's Like-for-like critical raw material change qualification - type of testing/ number of lots required ISO 13485:2016 - Medical Device Quality Management Systems 4
N Infant radiant warmer - Heater Quartz Element type Other Medical Device Related Standards 0
M Definitive answer on Type 1 vs Type 2 vs Type 3 Gage Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
K Validation of new machine (second machine of the same type) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
F Operator Manual and Type B Uncertainty Measurement Uncertainty (MU) 3
M SI #7 8.4.2.1 Type and extent of control_supplemental IATF 16949 - Automotive Quality Systems Standard 3
A Gage type 1 study on CMM Capability, Accuracy and Stability - Processes, Machines, etc. 2
A IATF Sanctioned Interpretation No. 7 - Type and Extent of Control (supplemental) IATF 16949 - Automotive Quality Systems Standard 4
I AS9100 8.4.2 Type and Extent of Control - External provider test reports Manufacturing and Related Processes 24
F Determining what type of scrap to include in my internal PPM calculation Quality Manager and Management Related Issues 5
R In this type of Legal Manufacturer-CMO arrangement, what happens to the DHR? ISO 13485:2016 - Medical Device Quality Management Systems 1
J AQL for Effectiveness Check - What AQL tiers to use on this type of sampling? Nonconformance and Corrective Action 9
T Type BF or CF (Direct conductive contact with the heart) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
G MSA Type 1 - Cg and Cgk ranges Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
P IVDD - IFU related requirements - Type of information/statement expected EU Medical Device Regulations 1
B Documented process for type and extent of control IATF 16949 - Automotive Quality Systems Standard 6
P Definition Business Type - Definition of what we do-Business Type Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 4
L Valid reasons to omit an "ear-tag" type label on a cable? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P 2x MOOP for type B AP - Therapeutic device - Heater/cooler with pump IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S Complying with IATF 16949 Cl. 8.4.2.1 - Type and extent of control IATF 16949 - Automotive Quality Systems Standard 1
P DMF (Drug Master File) type V for Sterilization Processes for USFDA Submission Other US Medical Device Regulations 3
E IATF 16949 Cl. 8.4.2.5 - Supplier Development Type and Extent IATF 16949 - Automotive Quality Systems Standard 2
F IATF 16949 - Cl. 8.4.2.3 - Which type of suppliers could be exempt of ISO 9001 IATF 16949 - Automotive Quality Systems Standard 16
M For a Type 1 Gage R&R I want to use 1 unit and measure it 30 times Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5