Hi all,

Is bioburden monitoring required for a single use surgical instrument that is provided non-sterile? The instrument is processed at the hospital before being used during surgery. This means that the instruments is cleaned and steam sterilized at the hospital according to the instructions in the instructions for use (IFU).

The steam sterilization cycle mentioned in the IFU was validated using the overkill method described in annex D of ISO 17665-1. This standard mentions that this method should be used when "the challenge to the sterilization process is difficult to define and pre-sterilization treatments (such as cleaning) are difficult to validate and control". As the bioburden on the device before sterilization is mainly depended on the cleaning/handling at the hospital, this seems appropriate. Annex D of ISO 17665-1 also does not mention that ongoing monitoring is required, while it does mention this explicitly when another method (the one of Annex B) is used.

Are we now required to perform bioburden monitoring on the surgical instrument before they are being shipped to the customer? In case of yes, can you point me to the applicable regulation or standard that is requiring this? Also in case of yes, is an acceptance criterion of <10^6 CFU/device appropriate as the overkill method used this amount of spores?

In my opinion this makes little sense to perform this monitoring as there is limited relation between the bioburden level at the time of shipment at our facilities and the level before sterilization at the hospital, but maybe I am forgetting something?

KR
 
Elsmar Forum Sponsor

indubioush

Quite Involved in Discussions
#2
ISO 13485:2016:
7.5.2 Cleanliness of product
The organization shall document requirements for cleanliness of product or contamination control of product if:

b) product is supplied non-sterile and is to be subjected to a cleaning process prior to sterilization or its use;
 

planB

Trusted Information Resource
#3
BioCPS,

no, you are typically not required to monitor the bioburden of your surgical tool prior to shipping in case you label a pretty standard (i.e. commonly used) cleaning, disinfection and sterilisation process. Make sure that the protective packaging used for your shipping configuration protects the device from excessive soiling.

HTH,

Gerhard
 
Thread starter Similar threads Forum Replies Date
M Biological Monitoring Program Environmental Bioburden Test Method Needed ISO 13485:2016 - Medical Device Quality Management Systems 1
S Bioburden Monitoring Sample Size Increase ISO 13485:2016 - Medical Device Quality Management Systems 4
J Bioburden Monitoring - 'Off the Shelf' Bioburden/Microbiological Monitoring Software ISO 13485:2016 - Medical Device Quality Management Systems 4
L Monitoring bioburden on clean room gloves Other Medical Device and Orthopedic Related Topics 1
E Bioburden tests duration and Monitoring of gloves ISO 13485:2016 - Medical Device Quality Management Systems 6
M Bioburden Frequency (sterile MD) Other Medical Device Related Standards 2
K When is Bioburden Testing Required? Other Medical Device Related Standards 4
A ISO 11135:2014, B.1.4, BI resistance x product bioburden ISO 13485:2016 - Medical Device Quality Management Systems 6
D Sterility and BioBurden Testing for Plastics ISO 13485:2016 - Medical Device Quality Management Systems 5
C Sterility and Bioburden Testing - LAL passed under 0.5EU/mL Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
B Radiation Sterilization Dose Audit: Bioburden frequency Other Medical Device Related Standards 1
F Bioburden Testing requirements for Non-Sterile Dental Products EU Medical Device Regulations 6
V Cleanliness/Bioburden for Non-Sterile Gas Path Medical Devices Other Medical Device Regulations World-Wide 7
K Bioburden 0 cfu for Dose Setting using the VDmax25 Other Medical Device Related Standards 9
N Must Bioburden Sampling Procedures be Documented? Other Medical Device Related Standards 5
R Simulated product for bioburden testing Other ISO and International Standards and European Regulations 2
C What is the most appropriate Bioburden Test Method for Surgical Gloves ISO 13485:2016 - Medical Device Quality Management Systems 10
M Bioburden numbers and ECA gloves 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Bioburden testing for Class 10000 (Grade C) clean room ISO 13485:2016 - Medical Device Quality Management Systems 1
O Bioburden Testing - Non Sterile Product / Tapes - ISO 11737-1 Other ISO and International Standards and European Regulations 9
P Serial Testing of Bioburden for Medical Devices Other Medical Device Related Standards 2
C What is an acceptable bioburden level or count when working in cleanroom ISO class7 ISO 13485:2016 - Medical Device Quality Management Systems 1
P Frequency of Bioburden Testing EU Medical Device Regulations 3
B BioBurden and EndoToxin Questions Other Medical Device Related Standards 6
M Sterility Testing vs Bioburden Testing - Differences Qualification and Validation (including 21 CFR Part 11) 7
I Bioburden and Endotoxin (LAL) Testing proposal help 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
I Bioburden Count Question - Bioburden is above the alert limit >1 CFU/component avg. 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
L GB15980-2003 Bioburden Count Trend on Operator Hands Other ISO and International Standards and European Regulations 1
0 Guideline/Method for Raw Material Roll Stock Bioburden Sampling Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
Q Bioburden Requirements on Device used on Unbroken Skin Other Medical Device and Orthopedic Related Topics 4
A Microbial Limit (Bioburden)on dry nonwoven that used for making Hand sanitizing wipes ISO 13485:2016 - Medical Device Quality Management Systems 1
D Bioburden and Product Labels - VD Max studies on all the BOMs Other Medical Device and Orthopedic Related Topics 9
H Determining a New Product Bioburden Limit ISO 13485:2016 - Medical Device Quality Management Systems 11
L Validation of transfer materials - Bioburden of items from lab into cleanroom Other Medical Device and Orthopedic Related Topics 2
M Spinal Implant Testing Requirements - Bioburden, TGA & Ion Chromatography 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
C Establishing alert and action limits for clean room bioburden Other Medical Device and Orthopedic Related Topics 9
C ISO 11737 - Bioburden testing for non-sterile products Other ISO and International Standards and European Regulations 10
S What is a reasonable Bioburden limit? ISO 13485:2016 - Medical Device Quality Management Systems 13
I Microbiological Testing - Bioburden and Identification of Micro-Organisms ISO 13485:2016 - Medical Device Quality Management Systems 2
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
C Monitoring and Control Instruments RoHS, REACH, ELV, IMDS and Restricted Substances 3
E Insulation diagram for ECG monitoring and diagnosis system. IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A How to reduce the process SPC monitoring Capability, Accuracy and Stability - Processes, Machines, etc. 3
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
S User evaluation for self monitoring blood glucose test systems US Food and Drug Administration (FDA) 4
S HIPAA-compliant monitoring software (advice needed) Hospitals, Clinics & other Health Care Providers 1
P ISO Class 8 particle count (annual certification vs monitoring) ISO 13485:2016 - Medical Device Quality Management Systems 4
CPhelan Best metrics for monitoring low volume (<1000) parts per vendors Supply Chain Security Management Systems 1
qualprod Monitoring of lead time - Good KPI identification? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
A How to set up Continuous CpK monitoring of an injection mold process Reliability Analysis - Predictions, Testing and Standards 7

Similar threads

Top Bottom