Biocompatbility testing on Class 1 device requirements

Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
Not true. As long as it comes in contact with human skin/tissue and is being used for a medical application - it requires biocompatibility testing, unless you can provide a very good rationale it is not required.
 

planB

Super Moderator
#3
Concurring with shimov: body-contacting medical devices require a biocompatibility evaluation demonstrating that the device does not cause an inacceptable biological response. In case you have enough data at hand to show that, you may omit testing. Otherwise, you will have to close these data gaps by testing.

Hope this helps,
 
#4
Thank you for your input, but Is there an FDA document that clearly states theat being 510(k) exempt does not exempt the manufacturer from biocomp testing?
 

Janie

Registered
#6
Yes, but in the Introduction to that document, it states that the guidance is to assist in preparing premarket applicatons HDEs, IDEs, 510ks, and De Novo request. The trach tube holder is exempt.
 

shimonv

Trusted Information Resource
#7
Janie,
I looked in the 2003 and 2018 edition of ISO 10993-1 and there is no mentioning of premarket application.
Either way, you are expert from it, but not from applicable standards.
 

planB

Super Moderator
#8
Janie,

have a look into FDA's guidance on biocompatibility section II, Scope:
The scope of this document and accompanying attachments is limited to the biological evaluation of sterile and non-sterile medical devices that come into direct or indirect contact with the human body.
Has your device body contact?

If yes, just because your device is exempt from pre-market approval, does not imply that you are exempt from demonstrating documented biocompatibility for your body-contacting device according to its defined intended use.

If no, well then the topic of biocompatbility does not affect your device.

Hope this help,
 

Watchcat

Trusted Information Resource
#10
Is a tracheostomy tube holder exempt from biocompatiblity testing requirements simply because it does not require a 510(k)?
IMO, Exemption from the requirement to submit a 510(k) notification to FDA is limited to just that. You don't have to submit a notification to FDA. It doesn't exempt you from anything else, including, IMO, the need to have data showing substantial equivalence to another legally marketed device that is exempt from the requirement to submit a 510(k) notification to FDA. I don't think it is a device that is exempt from anything just because you say so. It has to be SE to another device that is exempt from 510(k) notification, or it has to have been classified as 510(k) exempt device via De novo.

Regardless, like all medical devices regulated by FDA, it is subject to general controls. IMO, if it is dangerous to health, it is misbranded, and, if you haven't established biocompatibility, you don't know if it is dangerous to health, so a claim that it is safe for its intended use is false and misleading.
General Controls for Medical Devices

I repeat IMO a lot here because I'm not sure even FDA knows what it thinks about all of this. I don't know that it's had enough exposure to devices claiming to be 510(k) exempt to have even developed an opinion. Especially an informed opinion. What I'm pretty sure of? If a device the manufacturer claims is 510(k) exempt causes problems due to a lack of biocompatibility, and the manufacturer never evaluated it for biocompatibility, FDA will quickly develop a very strong opinion about that particular device and manufacturer.
 
Thread starter Similar threads Forum Replies Date
T Shrinkage Testing for EPDM Material Various Other Specifications, Standards, and related Requirements 1
J Managing design verification regression testing of design changes Design and Development of Products and Processes 0
D Conductivity Water Samples testing temperature Qualification and Validation (including 21 CFR Part 11) 2
G Pricing Structure for Physical Mechanical Testing of Labs Manufacturing and Related Processes 1
S Skipped/Reduced testing in Pharma for Raw materials and Intermediates Manufacturing and Related Processes 0
S Pharmaceutical Shelf Life Testing Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
D How to testing of Pressure relief valve as per API 6D ? Oil and Gas Industry Standards and Regulations 1
J Impact Testing--Is conversion between Izod and Sharpy Possible? General Measurement Device and Calibration Topics 3
O Should a Covid vaccine and testing policy be included as part of ISO9001 or AS9100 risk management? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
V How to select the tests for proficiency testing Measurement Uncertainty (MU) 0
B Customer Preference Testing Customer and Company Specific Requirements 2
N Pre-Clinical Performance Testing Design and Development of Products and Processes 3
I Software (SaMD) mobile application verification testing: objective evidence Medical Information Technology, Medical Software and Health Informatics 2
I How to find required testing for a specific device? Other US Medical Device Regulations 3
J CQI-9v4 Alternative TUS Testing Methods Various Other Specifications, Standards, and related Requirements 0
M ASTM F1717 Static and Dynamic Testing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T Mold Limits for Microbiology Testing Other Medical Device Related Standards 3
S Orthopedic implant sterility testing Medical Device and FDA Regulations and Standards News 1
H Independence between the development and testing IEC 62304 - Medical Device Software Life Cycle Processes 6
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Testing criteria - where to place Design and Development of Products and Processes 2
E 60601-1 - Tilt testing - Tensile safety factor IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
V Setup for testing against ISO14708 clause 16 (protection of the patient from herms caused by heat) Other Medical Device Related Standards 0
gramps What do you think about automated QA testing For software app industry? Misc. Quality Assurance and Business Systems Related Topics 5
B In house NIOSH pre Testing accepted by NIOSH? US Food and Drug Administration (FDA) 1
M Bacteriostasis/Fungistasis Testing Other Medical Device and Orthopedic Related Topics 6
P Sample Size for Distribution Simulation Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
N EN 813, EN 12277, EN 1497 - Testing some harness prototypes to an EN standard Various Other Specifications, Standards, and related Requirements 0
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 10
B NIOSH Approval for Surgical N95 Respirators - Required testing US Food and Drug Administration (FDA) 2
M ECG lead leakage currents - How to specify ECG leads during electrical safety testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C Stress / Challenge Conditions for Design Verification Testing to Reduce Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
D CFR Title 14: Aeronautics and Space PART 120—DRUG AND ALCOHOL TESTING PROGRAM Federal Aviation Administration (FAA) Standards and Requirements 3
lanley liao Purchase Acceptance Criteria - Tensile testing Oil and Gas Industry Standards and Regulations 2
M Device mounted at IV pole - what about mechanical stability testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Cochlear Implant Testing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Class I (exempt) testing requirements Other Medical Device Related Standards 1
JoCam Electrical Testing for Japan, PSE or CB Scheme Other Medical Device Regulations World-Wide 0
M Who are the go to companies for non-destructive hardness testing? General Measurement Device and Calibration Topics 3
M Determining if an Insulin Pen Testing Machine is a Medical Device? EU Medical Device Regulations 4
P Testing cloud-based backups IT (Information Technology) Service Management 7
I IATF Lab Scope Testing Qualification and Competency Documentation IATF 16949 - Automotive Quality Systems Standard 3
N Chemical Testing on Medical Devices - Solutions in a container closure system (bag) EU Medical Device Regulations 1
M Comparing data from destructive testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
T Flammability testing Reliability Analysis - Predictions, Testing and Standards 0
E Manufacturers should develop a testing device for covid19 Service Industry Specific Topics 0
K When is Bioburden Testing Required? Other Medical Device Related Standards 8

Similar threads

Top Bottom