Biocompatibility classification for Lancet

Krishna Priya T

Starting to get Involved
#1
Hi,

lancet is neither skin contact nor contact with blood path. can you please tell me the classification in which it will fall as per ISO 10993-1?
 
Elsmar Forum Sponsor

planB

Super Moderator
#2
Krishna,

what is the nature and duration of body contact your lancet has? What is the intended use of your device?
 

chris1price

Trusted Information Resource
#3
From my understanding, a lancet is usually a spike or needle used to puncture the skin in order to provide a blood sample. They certainly "contact intact skin" and also "Breached or compromised surfaces"; and would be of limited duration.
 

planB

Super Moderator
#4
In case the nature of body contact is really the one that Chris suspects, the category according to ISO 10993-1:2018 is either "Limited contact to breached or compromised surface", or "Limited tissue contact", both with the following endpoints according to Table A.1:
  • Physical and/or chemical information
  • Cytotoxicity
  • Sensitization /Irritation or intracutaneousreactivity
  • Material mediated pyrogenicity
  • Acute systemic toxicity
ISO 10993-1:2018, section 5.3.2 might also be useful in case the contact duration is very short:

Some medical devices with limited exposure (A) have very brief/transitory contact with the body (e.g. lancets, hypodermic needles, capillary tubes that are used for less than one minute). These generally would not require testing to address biocompatibility.
HTH,

Gerhard
 

chris1price

Trusted Information Resource
#5
Hi Gerhard

Thanks for the update about 10993-1:2018; that's good information that I had not seen.

I just spent 10 minutes wondering why I did not know about the 10993-1:2018 version. It is currently only an ISO standard and has not yet been published as EN ISO10993-1 or BS EN ISO....... The 2009 is still the harmonized version.
 

Krishna Priya T

Starting to get Involved
#6
Thanks Gerhard and Chris for the update.

I am not understanding contact to breached or compromised surface wrt Lancet. can you please brief on this.

Whether blood sample will be collected from breached or compromised surface?
 

planB

Super Moderator
#7
Krishna,

breached or compromised skin: opposite of intact or undamaged skin.

In case you do not get in contact with circulating blood, collecting blood samples does not introduce additional biological hazards to the list above.

@Chris: Depending on your notified body, their stance on "state of the art" may over-rule the concept of harmonized standards, since the harmonization process for medical devices essentially came to a stand-still in the past years.

HTH,

Gerhard
 
J

Jnks_Meddev

#8
Hi Everyone,

I am doing the gap analysis on our product to see if there is any change with the updated ISO 10993:2018. Can someone please let me know how do I go about this? Is there a checklist or something that I can refer to ensure any updates required for our product?
I need to make sure that the testing documents has no effect as well as conformance to ER with the updated revision.
 
Thread starter Similar threads Forum Replies Date
Judy Abbott The impact of laser power on biocompatibility of the products Manufacturing and Related Processes 6
Watchcat Summary of De Novo Biocompatibility Information, 2015-2018 Other US Medical Device Regulations 0
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
F Biocompatibility evaluation for Hardware/Interface Components? Other Medical Device Related Standards 12
L Process changes and biocompatibility (ISO 10993-1) Other Medical Device Related Standards 1
Beliz Biocompatibility Testing for Laser Epilation Device EU Medical Device Regulations 2
L FDA Biocompatibility Requirements - Transitory Contact Other US Medical Device Regulations 1
K Biocompatibility Testing - Multile products of different sizes and shapes US Food and Drug Administration (FDA) 2
K Biocompatibility evaluation of gas pathways Medical Device and FDA Regulations and Standards News 5
K Diagnostic X-ray devices - Applicability of Biocompatibility Testing per ISO 10993-1 Manufacturing and Related Processes 7
Judy Abbott Activation of complement at 90 min - BSI requested for justifications to support the biocompatibility of the device EU Medical Device Regulations 2
Judy Abbott Can any one help me with an urgent BIOCOMPATIBILITY Impact assessment? Internal Auditing 9
S ISO 10993 Biocompatibility Evaluation - Electronic thermometer Other Medical Device Related Standards 3
shimonv Referring to previously submitted biocompatibility results in a new 510(k) submission Other US Medical Device Regulations 1
I ISO 10993 Biocompatibility Requirements - manufacturing process chemicals Other Medical Device Related Standards 8
R 510(k) Biocompatibility vs Cleaning Validation Other Medical Device Related Standards 4
F Sample Size for Biocompatibility Tests Other Medical Device Related Standards 4
F Biocompatibility Device Categorization: Priming with Blood Other Medical Device Related Standards 1
W Biocompatibility Risk Analysis for Clinical Practitioner 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
F Biocompatibility on aged devices Other Medical Device Related Standards 3
Gamula Biocompatibility GLP conform tests outside China China Medical Device Regulations 1
K Biocompatibility Testing - Component Level or Assembled Medical Device? ISO 13485:2016 - Medical Device Quality Management Systems 7
I Biocompatibility Impact on Autoclaved Silicone Other Medical Device Related Standards 3
D Should Biocompatibility Tests be performed on Production Samples? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
R What is the expectation for Biocompatibility Testing when switching Suppliers? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
C Biocompatibility Test for a Non Contact Medical Device US Food and Drug Administration (FDA) 11
R Biocompatibility - ISO 10993 and SVHC Requirements Other Medical Device Related Standards 4
planB Biocompatibility: Blue Book Memorandum #G95-1 superseded by new FDA guidance Other Medical Device Related Standards 1
S ISO 10993-1: External Communicating Device biocompatibility test requirements Other ISO and International Standards and European Regulations 4
M Biocompatibility - why are animal studies necessary? Other Medical Device and Orthopedic Related Topics 11
R Biocompatibility Test of Blue Gown vs Blue Drape 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R FDA Biocompatibility questions for Surgical Gown 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Biocompatibility requirements for a Class IIa medical device CE Marking (Conformité Européene) / CB Scheme 7
S Biocompatibility testing for nasal prongs made of medical grade silicone 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
J Biocompatibility tests in AAA LAC Laboratory ISO 13485:2016 - Medical Device Quality Management Systems 6
B Biocompatibility for Packaging Materials Other Medical Device Related Standards 3
J Biocompatibility Evaluation component other than Testing Other Medical Device Regulations World-Wide 1
J Grouping Products and Testing Sterile and Biocompatibility - Validating Products Other Medical Device Regulations World-Wide 3
S Laminaria - Is a Biocompatibility Study Required? EU Medical Device Regulations 6
M Biocompatibility in Respiratory Products - Help interpreting whitepaper IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
S Laminaria - Biocompatibility Study? Imported Legacy Blogs 1
M Three final products in one biocompatibility test? Other Medical Device and Orthopedic Related Topics 2
S Selecting materials for implants to comply with ISO 10993 biocompatibility Other Medical Device Related Standards 4
S Laminaria Stick (Cervical Dilator) - Biocompatibility Test required? ISO 13485:2016 - Medical Device Quality Management Systems 1
T Biocompatibility for "Desktop" used Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 4
M Biocompatibility Test requirements for 510k - Orthopaedic Implant Screws US Food and Drug Administration (FDA) 2
A Biocompatibility omission justification for FDA 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
Ajit Basrur Is Biocompatibility study performed on medical devices for possible contaminants? Other Medical Device Related Standards 4

Similar threads

Top Bottom