we are an Orthopaedic Implant manufacturing company and manufacture different kinds of plates , screws , nails .We are planning to apply for 510k .Our consultant is saying that for each and every product we need a biocompatibility test done . Now screws are made form 8mm or 6mm dia bars in various lengths ie, from 5mm to 100mm and diameters also vary from 2.7mm to 8mm so he wants a test done for each length and dia . Similar for different type of plates which are made from sheets thickness from 2mm to 6mm. Is this really required ?? or can we just test one size of screw and one type of plate since the raw material is the same ss / ti. We would be having a total of 400 products like this and it would not be economically feasible to do testing on all products . What is the solution .
Thanks
Michelle
Elsmar Forum Sponsor
M
MIREGMGR
#2
1. It's permissible in some circumstances to obtain a 510(k) on a family or bundled basis, when the entire family shares key characteristics, usually including material and manufacturing processes. In such a case, you might obtain biocompatibility data for the largest and smallest family members, or perhaps just a single one.
You should be familiar with http://www.fda.gov/medicaldevices/deviceregulationandguidance/guidancedocuments/ucm089731.htm and ISO 10993-1:2009.
2. If the materials being used are the subject of a Recognized Consensus Standard, it's much more straightforward to establish implant suitability. Per ISO 10993-1:2009 Section 4.1, this may allow provision of a rationale as to why no biocompatibility testing is required.
If your consultant's advice is fairly characterized, it sounds as if perhaps he or she isn't doing a good job for you.
planB
Super Moderator
#3
Adding to the previous poster:
With a portfolio of 400 products, you definitely may want to aim at grouping your products into families based on similarities in
(1) materials
(2) manufacturing processes
(3) intended use (i.e. nature and duration of body contact according to ISO 10993-1:2009)
For each family, you may want to define a worst-case representative in terms of (1), (2) and (3). For these worst-case representatives you then have to perform your material characterization and any biological test you may consider necessary to mitigate biological risks.
Alternatively, you may also want to consider pooling your families. While this may drastically bring down the numbers of test runs you have to perform, you will also loose information. You would have to assess whether pooling of test items is an option in your case.
HTH,
Gerhard
Thread starter Similar threads Forum Replies Date
S ISO 10993-1: External Communicating Device biocompatibility test requirements Other ISO and International Standards and European Regulations 4
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
C Biocompatibility Test for a Non Contact Medical Device US Food and Drug Administration (FDA) 11
R Biocompatibility Test of Blue Gown vs Blue Drape 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Three final products in one biocompatibility test? Other Medical Device and Orthopedic Related Topics 2
S Laminaria Stick (Cervical Dilator) - Biocompatibility Test required? ISO 13485:2016 - Medical Device Quality Management Systems 1
M Biocompatibility Test Methods for Sensitization US Food and Drug Administration (FDA) 5
J SFDA Biocompatibility Test Report Requests (China) ISO 13485:2016 - Medical Device Quality Management Systems 10
N Can I submit 510(k) without Biocompatibility Test Results? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
bio_subbu Choosing a laboratory for biocompatibility test ISO 13485:2016 - Medical Device Quality Management Systems 8
C Biocompatibility test for absorbable ophthalmic suture EU Medical Device Regulations 1
S Biocompatibility requirement for electrode adapters Other Medical Device Related Standards 1
S ASTM F86 x Biocompatibility Issues Medical Device and FDA Regulations and Standards News 1
Judy Abbott The impact of laser power on biocompatibility of the products Manufacturing and Related Processes 6
Watchcat Summary of De Novo Biocompatibility Information, 2015-2018 Other US Medical Device Regulations 0
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
F Biocompatibility evaluation for Hardware/Interface Components? Other Medical Device Related Standards 12
L Process changes and biocompatibility (ISO 10993-1) Other Medical Device Related Standards 1
Beliz Biocompatibility Testing for Laser Epilation Device EU Medical Device Regulations 2
L FDA Biocompatibility Requirements - Transitory Contact Other US Medical Device Regulations 1
K Biocompatibility Testing - Multile products of different sizes and shapes US Food and Drug Administration (FDA) 2
K Biocompatibility evaluation of gas pathways Medical Device and FDA Regulations and Standards News 5
K Diagnostic X-ray devices - Applicability of Biocompatibility Testing per ISO 10993-1 Manufacturing and Related Processes 7
Judy Abbott Activation of complement at 90 min - BSI requested for justifications to support the biocompatibility of the device EU Medical Device Regulations 2
Judy Abbott Can any one help me with an urgent BIOCOMPATIBILITY Impact assessment? Internal Auditing 9
S ISO 10993 Biocompatibility Evaluation - Electronic thermometer Other Medical Device Related Standards 3
shimonv Referring to previously submitted biocompatibility results in a new 510(k) submission Other US Medical Device Regulations 1
I ISO 10993 Biocompatibility Requirements - manufacturing process chemicals Other Medical Device Related Standards 8
R 510(k) Biocompatibility vs Cleaning Validation Other Medical Device Related Standards 4
K Biocompatibility classification for Lancet CE Marking (Conformité Européene) / CB Scheme 7
F Sample Size for Biocompatibility Tests Other Medical Device Related Standards 4
F Biocompatibility Device Categorization: Priming with Blood Other Medical Device Related Standards 1
W Biocompatibility Risk Analysis for Clinical Practitioner 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
F Biocompatibility on aged devices Other Medical Device Related Standards 4
Gamula Biocompatibility GLP conform tests outside China China Medical Device Regulations 1
K Biocompatibility Testing - Component Level or Assembled Medical Device? ISO 13485:2016 - Medical Device Quality Management Systems 7
I Biocompatibility Impact on Autoclaved Silicone Other Medical Device Related Standards 3
D Should Biocompatibility Tests be performed on Production Samples? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
R What is the expectation for Biocompatibility Testing when switching Suppliers? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
R Biocompatibility - ISO 10993 and SVHC Requirements Other Medical Device Related Standards 4
planB Biocompatibility: Blue Book Memorandum #G95-1 superseded by new FDA guidance Other Medical Device Related Standards 1
M Biocompatibility - why are animal studies necessary? Other Medical Device and Orthopedic Related Topics 11
R FDA Biocompatibility questions for Surgical Gown 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Biocompatibility requirements for a Class IIa medical device CE Marking (Conformité Européene) / CB Scheme 7
S Biocompatibility testing for nasal prongs made of medical grade silicone 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
J Biocompatibility tests in AAA LAC Laboratory ISO 13485:2016 - Medical Device Quality Management Systems 6
B Biocompatibility for Packaging Materials Other Medical Device Related Standards 3
J Biocompatibility Evaluation component other than Testing Other Medical Device Regulations World-Wide 1
J Grouping Products and Testing Sterile and Biocompatibility - Validating Products Other Medical Device Regulations World-Wide 3
Similar threads
Top Bottom