Biocompatibility Testing - Multile products of different sizes and shapes

#1
We have several products with the same indication that are different sizes and shapes. If the contact material is the same in all of them (material, manufacturer and process) can we do the largest surface area product and cite that data in all of the PMA submissions?

Thanks,
Kathy
 
Elsmar Forum Sponsor

William55401

Involved In Discussions
#2
I think you are on the right track. I suggest reading (and applying) "Use of International Standard ISO 10993-1, "Biological evaluation of medical devices - Part 1: Evaluation and testing within a risk management process" Guidance for Industry and Food and Drug Administration Staff". It does include "worst case" language.

https://www.fda.gov/media/85865/download
 

monoj mon

Involved In Discussions
#3
In addition to what William has said about the "Worst Case" from the context of "Largest Surface Area", I would like to mention about the "Worst Case" from the context of "Difficulty to Sterilize" (guessing your device as a sterile one) while doing the evaluation for the "Representative Device" to conduct the biocompatibility testing and document the decision making process.
 
Thread starter Similar threads Forum Replies Date
K Diagnostic X-ray devices - Applicability of Biocompatibility Testing per ISO 10993-1 Manufacturing and Related Processes 5
K Biocompatibility Testing - Component Level or Assembled Medical Device? ISO 13485:2016 - Medical Device Quality Management Systems 7
R What is the expectation for Biocompatibility Testing when switching Suppliers? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
S Biocompatibility testing for nasal prongs made of medical grade silicone 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
J Biocompatibility Evaluation component other than Testing Other Medical Device Regulations World-Wide 1
J Grouping Products and Testing Sterile and Biocompatibility - Validating Products Other Medical Device Regulations World-Wide 3
R Is Biocompatibility Testing required for Implantable Class III Devices sent to China China Medical Device Regulations 3
C Biocompatibility Evaluation Testing on Single Use Tube Sets Other Medical Device Related Standards 9
I Biocompatibility Testing Requirements for a Class I device - Does not contact patient 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
C Do we have to use an FDA recognized lab for Biocompatibility testing of 510k devices? Other US Medical Device Regulations 1
S Biocompatibility - Complement System Testing US Food and Drug Administration (FDA) 1
W Biocompatibility of Implantable Component - What testing, if any? Other Medical Device and Orthopedic Related Topics 5
M Medical Device Biocompatibility Testing - End Product or Components - ISO 10993 ISO 13485:2016 - Medical Device Quality Management Systems 22
T Animals for biocompatibility testing? Other Medical Device Related Standards 10
V Biocompatibility testing centre in Montreal QC ? Other Medical Device and Orthopedic Related Topics 3
A Biocompatibility testing & Registration in China ISO 13485:2016 - Medical Device Quality Management Systems 5
M Predicate Device - Question on Biocompatibility and IEC 60601-1 Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T Biocompatibility Testing for Implantable Medical Devices according to ISO 10993 ISO 13485:2016 - Medical Device Quality Management Systems 4
L FDA Biocompatibility Requirements - Transitory Contact Other US Medical Device Regulations 1
K Biocompatibility evaluation of gas pathways Medical Device and FDA Regulations and Standards News 5
Judy Abbott Activation of complement at 90 min - BSI requested for justifications to support the biocompatibility of the device EU Medical Device Regulations 2
Judy Abbott Can any one help me with an urgent BIOCOMPATIBILITY Impact assessment? Internal Auditing 9
S ISO 10993 Biocompatibility Evaluation - Electronic thermometer Other Medical Device Related Standards 3
shimonv Referring to previously submitted biocompatibility results in a new 510(k) submission Other US Medical Device Regulations 1
I ISO 10993 Biocompatibility Requirements - manufacturing process chemicals Other Medical Device Related Standards 8
R 510(k) Biocompatibility vs Cleaning Validation Other Medical Device Related Standards 4
K Biocompatibility classification for Lancet CE Marking (Conformité Européene) / CB Scheme 7
F Sample Size for Biocompatibility Tests Other Medical Device Related Standards 4
F Biocompatibility Device Categorization: Priming with Blood Other Medical Device Related Standards 1
W Biocompatibility Risk Analysis for Clinical Practitioner 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
F Biocompatibility on aged devices Other Medical Device Related Standards 3
Gamula Biocompatibility GLP conform tests outside China China Medical Device Regulations 1
I Biocompatibility Impact on Autoclaved Silicone Other Medical Device Related Standards 3
D Should Biocompatibility Tests be performed on Production Samples? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
C Biocompatibility Test for a Non Contact Medical Device US Food and Drug Administration (FDA) 11
R Biocompatibility - ISO 10993 and SVHC Requirements Other Medical Device Related Standards 4
planB Biocompatibility: Blue Book Memorandum #G95-1 superseded by new FDA guidance Other Medical Device Related Standards 1
S ISO 10993-1: External Communicating Device biocompatibility test requirements Other ISO and International Standards and European Regulations 4
M Biocompatibility - why are animal studies necessary? Other Medical Device and Orthopedic Related Topics 11
R Biocompatibility Test of Blue Gown vs Blue Drape 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R FDA Biocompatibility questions for Surgical Gown 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Biocompatibility requirements for a Class IIa medical device CE Marking (Conformité Européene) / CB Scheme 7
J Biocompatibility tests in AAA LAC Laboratory ISO 13485:2016 - Medical Device Quality Management Systems 6
B Biocompatibility for Packaging Materials Other Medical Device Related Standards 3
S Laminaria - Is a Biocompatibility Study Required? EU Medical Device Regulations 6
M Biocompatibility in Respiratory Products - Help interpreting whitepaper IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
S Laminaria - Biocompatibility Study? Imported Legacy Blogs 1
M Three final products in one biocompatibility test? Other Medical Device and Orthopedic Related Topics 2
S Selecting materials for implants to comply with ISO 10993 biocompatibility Other Medical Device Related Standards 4
S Laminaria Stick (Cervical Dilator) - Biocompatibility Test required? ISO 13485:2016 - Medical Device Quality Management Systems 1
Similar threads