Bringing Production In house

Ed Panek

QA RA Small Med Dev Company
Staff member
Super Moderator
#1
Working as QA RA lead on a project bringing our prod in house (ISO 13485 FDA Class II) from a contract manu. The lead engineer is insisting I determine what the requirements are for the machines and tool and the environment we use. I am stating that aside from ISO 14121 there are none. The requirements are the tools are safe/effective have been through V&V. The environment is not prescribed and just states we have to determine the env controls and implement and monitor them. Is this correct?
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#3
Working as QA RA lead on a project bringing our prod in house (ISO 13485 FDA Class II) from a contract manu. The lead engineer is insisting I determine what the requirements are for the machines and tool and the environment we use. I am stating that aside from ISO 14121 there are none. The requirements are the tools are safe/effective have been through V&V. The environment is not prescribed and just states we have to determine the env controls and implement and monitor them. Is this correct?
These questions are a subset of what you need to look at when moving manufacturing in house, from a QA/RA (FDA & CE?) perspective.
I would take a more holistic approach to managing and addressing applicable requirements.
I sent you a PM some time ago - not sure you've seen it.
All the best,
Ronen.
 

somashekar

Staff member
Super Moderator
#4
Working as QA RA lead on a project bringing our prod in house (ISO 13485 FDA Class II) from a contract manu. The lead engineer is insisting I determine what the requirements are for the machines and tool and the environment we use. I am stating that aside from ISO 14121 there are none. The requirements are the tools are safe/effective have been through V&V. The environment is not prescribed and just states we have to determine the env controls and implement and monitor them. Is this correct?
Please tell that lead engineer that he at least maintains the same machines and tools and the environment as was approved at the CM.
If you are planning to get those same machines and tools from your CM ... OR you plan to buy the same type or better machines and tools., still you need to perform all your validations exercises (IQ/OQ/PQ) including operator qualification as your case needs.
About environment, if your device requirement does not need any special conditions to be created, and no such was also in place at the CM, your environment is ambient. But this needs a deeper levels of inputs and analysis considering your device and the locations of your CM and yourself.
 

Ed Panek

QA RA Small Med Dev Company
Staff member
Super Moderator
#5
For IQ OQ PQ we plan on developing that for each internally developed and externally sourced tool. The overall implementation plan is much larger of course and this is a single action item.
 

yodon

Staff member
Super Moderator
#6
The point @Ronen E makes about a holistic assessment is likely what your lead engineer is asking about. Do you have adequate space for the equipment, are there new HVAC requirements as a result, does the equipment require special flooring, do you have sufficient utilities (power, air, water, etc.) to support the equipment, etc.

Realize also that the QMS is probably impacted - infrastructure and work environment.
 

Ed Panek

QA RA Small Med Dev Company
Staff member
Super Moderator
#7
The QMS is absolutely impacted. We have performed a gap analysis and are working on the required documentation updates and risk analysis. I have internal (contracted) audits scheduled prior to production.

I am trying to state that the QA group doesn't drive the production method requirements - there are probably dozens of methods to execute our production; we are agnostic. Engineering does that work. Our process is very simple. It requires no more than 4 people start to finish and can be done around a picnic sized table - this is a Med Device startup with 15 employees living on investors funding. We want to demonstrate to our investors and new investors a more robust operation with production in house.

QA will be there to review and help develop plans to make sure requirements regarding floorspace, environmental controls, verification/validation, SOPs & Forms and other requirements. How we decide to build production is for us as a business to decide. I have to assure that whatever the method I can adjust it to meet the regulations we are required to meet.

My largest current concern is Im having trouble picturing the material movement workflow to prevent mix-ups. At other larger companies that segregation was simply required with the large inventory we managed. With a smaller operation creating that same spacial segregation is more challenging without having that much space.
 
Last edited:

Ed Panek

QA RA Small Med Dev Company
Staff member
Super Moderator
#8
Update: Making some progress. I created a "warehouse specification" to provide regulatory guidance on how ISO and QSR expect things to run. Should I entertain the idea of a "design review" of the floor layout to help catch issues before we have nearly completed the task of warehouse development? Does the "prevent mix up" clauses of QSR and ISO require any documentation?
 

Al Rosen

Staff member
Super Moderator
#9
Does the "prevent mix up" clauses of QSR and ISO require any documentation?
It's required.
Subpart L--Handling, Storage, Distribution, and Installation
Sec. 820.140 Handling.
Each manufacturer shall establish and maintain procedures to ensure that mixups, damage, deterioration, contamination, or other adverse effects to product do not occur during handling.
 

Ed Panek

QA RA Small Med Dev Company
Staff member
Super Moderator
#10
We will, of course, write our SOPs with that in mind but the actual floor layout...does it need a control document?

It's required.
Subpart L--Handling, Storage, Distribution, and Installation
Sec. 820.140 Handling.
Each manufacturer shall establish and maintain procedures to ensure that mixups, damage, deterioration, contamination, or other adverse effects to product do not occur during handling.
 
Thread starter Similar threads Forum Replies Date
Ed Panek Bringing Production in house from Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 3
Y 510k requirement for bringing in gloves Medical Device and FDA Regulations and Standards News 0
T Training recommendations? Bringing our RoHS and REACH compliance efforts in-house REACH and RoHS Conversations 2
M Bringing blending in-house - Verification or full validation? (Disinfectants) ISO 13485:2016 - Medical Device Quality Management Systems 3
Sidney Vianna IATF bringing stakeholders together to ensure Transition by the 2018-09-14 deadline IATF 16949 - Automotive Quality Systems Standard 10
Q Bringing Medical Device Prototypes into the US Other US Medical Device Regulations 3
T AS9104/3 is bringing in a requirement for an AA to qualify as AEA based on Audit Days AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Bringing a Medical Device Sample from a Foreign Country into the US 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
Q Gap Analysis - Bringing a smaller subsidiary company up to IS0 9001:2008 standard ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
C Bringing a Second Office Location into Registration Scope ISO 14001:2015 Specific Discussions 12
C Bringing a CQA Primer Copy for an exam Professional Certifications and Degrees 14
C Bringing Simple Calibrations Back In-House General Measurement Device and Calibration Topics 16
T QMS in place - How to illustrate? Small matter of bringing it into action ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Wes Bucey To sue, or not to sue: That is the question until you are the person bringing suit... Coffee Break and Water Cooler Discussions 73
C Validation of process for production and servicing 5.7.1.5 API Spec Q1, 9th Edition Oil and Gas Industry Standards and Regulations 3
O Production run inventory sell-by expiration date? APQP and PPAP 3
A ISO 9001 Internal Audits - No production right now due to furloughs Internal Auditing 5
DuncanGibbons Process flow & PFMEA for production planning and simulation activities? Process Maps, Process Mapping and Turtle Diagrams 7
D 6.1.2.3 Contingency Plan - Dispositions to restart production after an unplanned stoppage IATF 16949 - Automotive Quality Systems Standard 2
S Production Work Order Procedure 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Techilady Ford Special Characteristics Communication and Agreement Form (SCCAF) provides production requirements? FMEA and Control Plans 4
thisby_ Post Market/Production Risk Assessment ISO 14971 - Medical Device Risk Management 0
Felony Melony Project Milestone Plan-Development to Mass Production APQP and PPAP 2
M Difference between "Production Trial Run" and "Run at Rate" IATF 16949 - Automotive Quality Systems Standard 8
M Material incoming to the production process reflected in PFMEA FMEA and Control Plans 9
C Scrapping on production floor - Setup parts 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
G Dealing with non conformity caused by Supplier Components detected in the production line IATF 16949 - Automotive Quality Systems Standard 14
T IATF Rules for sharing production space with another company IATF 16949 - Automotive Quality Systems Standard 10
qualprod Best practice to ensure inputting of data in production Lean in Manufacturing and Service Industries 19
R Monitor production quality - Internal KPIs Manufacturing and Related Processes 5
E Routine production tests - dielectric voltage withstand test IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P Conformity assessment based on a quality management system or production quality assurance EU Medical Device Regulations 3
Anonymous16-2 Dietary Supplement Raw Material/Component Storage and Production Manufacturing and Related Processes 1
L Significant Production Run - How 300 was determined to be the minimum quantity APQP and PPAP 2
Q ISO 9001 8.5.1 - Control of production and service performance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
C Production and Post Production feedback - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 4
DuncanGibbons How is the arrangement between Design and Production organisation envisaged? EASA and JAA Aviation Standards and Requirements 4
J Production Line Test Brasil - High Voltage Dielectric Strenght Test Other Medical Device Regulations World-Wide 5
J Iterative design and production for custom made products ISO 13485:2016 - Medical Device Quality Management Systems 3
M Production approval testing - Alternative ideas for Validation Reliability Analysis - Predictions, Testing and Standards 4
S How do you record your Nonconforming product? Batch production of staged processes Nonconformance and Corrective Action 2
A When there is a 2 year lapse in production, is a full FAI required? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
P Definition of production tooling in IATF16949 clause 8.5.1.6 IATF 16949 - Automotive Quality Systems Standard 4
Moncia Production Planning procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D Pre-Production Capability Assesment Reliability Analysis - Predictions, Testing and Standards 5
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
A Production of raw materials in university labs? Lean in Manufacturing and Service Industries 11
D How long should we keep the spare parts available for our medical device, after we have stopped the production? ISO 13485:2016 - Medical Device Quality Management Systems 0
V ISO 9001: 2015 Production drawing Control ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D Determining Calibration Frequency schedule for items used in production Manufacturing and Related Processes 2

Similar threads

Top Bottom