Burst Test Acceptability Criteria

Philip B

Involved In Discussions
#1
Where you carry out burst testing of packaging what acceptability criteria have you used? (This is for sealed pouches opened via a tear notch, not peelable pouches.) If you establish an average burst test result of, say, 600 mbar, is this good, bad or ok? I feel that the results need to relate to a real world condition but we have always struggled to come up with something. Any help appreciated...
 
Elsmar Forum Sponsor

chris1price

Trusted Information Resource
#2
There isn't a set pressure that is acceptable as the configuration, test equipment etc make for many variables. I'd make samples that are poor and use those to determine the acceptable pressure. Then make sure you are operating higher, setting appropriate action & alert levels.
 

Philip B

Involved In Discussions
#3
There isn't a set pressure that is acceptable as the configuration, test equipment etc make for many variables. I'd make samples that are poor and use those to determine the acceptable pressure. Then make sure you are operating higher, setting appropriate action & alert levels.
Thanks Chris
 
Thread starter Similar threads Forum Replies Date
N Is Air Burst test for oven treated condoms still required under ISO 4074? Other Medical Device Related Standards 1
A Burst Test equipment to perform ASTM F1140 Manufacturing and Related Processes 6
R Seal Packaging - Burst and Leak Specifications for a Foil Seal applied to a PTFE well Manufacturing and Related Processes 4
P ASTM F2054 specification for burst testing on sealed foil & laminate pouch Manufacturing and Related Processes 4
P ASTM D 380 Clause 16.1 - Methods of Burst Testing Rubber and Plastic Hoses Various Other Specifications, Standards, and related Requirements 3
B How to test operator to see if they can catch the defective parts? Manufacturing and Related Processes 6
G 60601-1 8.8.3. short circuit of both side during insulation test IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Bacterial Endotoxin test for Purified water Manufacturing and Related Processes 2
shimonv Test Method Validation ISO 13485:2016 - Medical Device Quality Management Systems 10
C Test compulsory for Sutures? Manufacturing and Related Processes 6
dgrainger Informational DRAFT: The Medical Devices (Coronavirus Test Device Approvals) (Amendment) Regulations 2021 UK Medical Device Regulations 1
K Identification test in ph.eur. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 1
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 6
R Cost per test tool/formula Oil and Gas Industry Standards and Regulations 1
R Cost per test tool/formula Oil and Gas Industry Standards and Regulations 1
A R chart from ANOVA gage R&R test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A 60601-1 clause 8.11.3.6 Test Clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
E Tensile test machine, tinius olsen 602 Manufacturing and Related Processes 1
M Test failure Root cause not found Customer Complaints 10
Moumen H Variations between ASTM A29 Standard for steel bars and Mill test certificates specs Manufacturing and Related Processes 1
D Tip test 10° -> 5° IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Sponsor name presented in the test report EU Medical Device Regulations 4
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 1
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Non parametric test for semi-quantitative data. Statistical Analysis Tools, Techniques and SPC 5
Ed Panek Disaster Recovery Test IT (Information Technology) Service Management 3
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Z Is Op Amp Common-Mode Rejection Ratio (CMRR) test applicable to battery operated devices? Other Medical Device Related Standards 0
E When to perform accelerated aging studies and transport test(OQ or PQ) Qualification and Validation (including 21 CFR Part 11) 4
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1
A Test requirements for marketing in US when RATED values are not of US IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H Secondary Circuits, Voltage Test and Clearance/Creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Fogging test as per DIN 75201(50ug/g ) VDA Standards - Germany's Automotive Standards 3
R HV Test of HF Surgical Unit IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E 60950-1 resistance to abnormal heat test Various Other Specifications, Standards, and related Requirements 0
S User evaluation for self monitoring blood glucose test systems US Food and Drug Administration (FDA) 4
E MSA for Push and Pull test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9

Similar threads

Top Bottom