Calibratable Thermometer / Hygrometer for Quality Control Room

F

Freddyfred

#1
Does anyone know where I can find a good quality calibratable "Dial" Thermometer/Hygrometer for a Quality Control Room?
Thanks, Fred
 
Elsmar Forum Sponsor

dgriffith

Quite Involved in Discussions
#2
Does it have to be a dial-type?
Fluke/Hart Scientific's DEWK thermo-hygrometer has a 1 or 2 channel option, with high or low accuracy, large digital display with graphing, IR, networkable, over a years' worth of on-board storage, etc.
We have 7 at our facility, 1 for each lab. One probe hangs centrally from the ceiling, the other channel can be taken to a specific bench for monitoring if needed.

Also, the probes have on-board calibration and ID data storage, so another probe can be hot-swapped in when one has to go in for cal. Only the probes get cal'd. The display stays put.
 
Last edited:
Thread starter Similar threads Forum Replies Date
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
S FDA's E submitter - For clinical electronic thermometer Medical Device and FDA Regulations and Standards News 6
R Measurement Uncertainties Budget for Thermometer Measurement Uncertainty (MU) 6
S ISO 10993 Biocompatibility Evaluation - Electronic thermometer Other Medical Device Related Standards 3
M Traceable Calibration Of An Oven Thermometer General Measurement Device and Calibration Topics 4
J Could Fluke thermometer be used in SAT test in accordance to AMS 2750 ? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
O Electronic Fever Thermometer - Why not IEC 62304 Class C? IEC 62304 - Medical Device Software Life Cycle Processes 7
N Calibration Method of Dial Thermometer Gauges General Measurement Device and Calibration Topics 7
C How to Calibrate a Medical Infrared Thermometer General Measurement Device and Calibration Topics 3
Y No permissible error statement on the thermometer calibration report General Measurement Device and Calibration Topics 7
B Mecury-in-Glass Thermometer Calibration Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Y Calibration Frequency Requirements of Liquid in Glass Thermometer General Measurement Device and Calibration Topics 2
B NIST Traceable Thermometer Uncertainty Measurement Uncertainty (MU) 5
J Appropriate Test Environments for In-Glass Liquid Thermometer Calibrations General Measurement Device and Calibration Topics 2
S Calibrating Thermometer with Simulated Output Temperature from Fluke 5520A General Measurement Device and Calibration Topics 6
A Liquid Bath Thermometer Calibration - Chemical Baths (Anodize Line) General Measurement Device and Calibration Topics 3
N Can a Calibrated Digital Thermometer be used to monitor my oven? General Measurement Device and Calibration Topics 12
D Calibrating a Thermometer for use in Cold Rooms - ISO 9001 Clause 7.6 General Measurement Device and Calibration Topics 8
Mohammed Gouda Latest standard for thermocouples & iprt (industrial platinum resistance thermometer) General Measurement Device and Calibration Topics 5
Mohammed Gouda Liquid Bath Depth for Calibration of Glass Immersion Type Thermometer General Measurement Device and Calibration Topics 10
Mohammed Gouda How to "Anneal" a Glass Thermometer General Measurement Device and Calibration Topics 9
D Control of monitoring and measuring equipment - infrared thermometer for cross-check ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Minimum Frequency for Thermometer Calibration Food Safety - ISO 22000, HACCP (21 CFR 120) 5
D Thermometer Calibration and Calibration Interval ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D How to Schedule Mercury-in-Glass Thermometer Calibration Calibration Frequency (Interval) 17
Q Would you Calibrate this Digital Hygrometer / Thermometer? General Measurement Device and Calibration Topics 17
B Thermometer Vendor that can Supply MSDS General Measurement Device and Calibration Topics 3
T Procedure for Half Yearly Electronic Thermometer Ice Point Check General Measurement Device and Calibration Topics 3
C How do I calculate the accuracy of a thermometer? General Measurement Device and Calibration Topics 2
S Mercury in Glass Thermometer Uncertainty Budget Measurement Uncertainty (MU) 1
P Thermo-Hygrometer - Measurement of 10DegC @10%RH ISO 17025 related Discussions 4
M How to read a calibration certificate for thermo hygrometer General Measurement Device and Calibration Topics 2
A Calibration/Verification of thermo-hygrometer in storage areas ISO 13485:2016 - Medical Device Quality Management Systems 9
T Uncertainty in Hygrometer Calibration Certificate General Measurement Device and Calibration Topics 4
D How much Deviation is Acceptable when Calibrating a Hygrometer? General Measurement Device and Calibration Topics 3
N Interpreting Hygrometer Calibration Uncertainty - Temperature Coefficient Measurement Uncertainty (MU) 1
Howard Lee Hygrometer Calibration - Best way to verify accuracy of a 'reference' Hygrometer? General Measurement Device and Calibration Topics 11

Similar threads

Top Bottom