From 21 CFR Part 820.72 (a)
"Each manufacturer shall ensure that all inspection, measuring, and test equipment, including mechanical, automated, or electronic inspection and test equipment, is suitable for its intended purposes and is capable of providing valid results. Each manufacturer shall establish and maintain procedures to ensure that equipment is routinely calibrated, inspected, checked, and maintained. The procedures shall include provisions for handling, preservation, and storage of equipment, so that its accuracy and fitness for use are maintained. These activities shall be documented."
From 21 CFR Part 820.72 (b)
"Calibration procedures shall include specific directions and limits for accuracy and precision. When accuracy and precision limits are not met, there shall be provisions for remedial action to reestablish the limits and evaluate whether there was any adverse effect on the device's quality. These activities shall be documented."
When you make changes to a calibrated device without first ascertaining its current accuracy, you have lost traceability for that instrument clear back to the last time it was calibrated. You are assuming that nothing changed between the time of the last calibration and the modification, but you can't realistically prove it. The same principle applies when calibrated devices are being retired or placed on an inactive status. The device should be calibrated immediately prior to storage or disposal so that traceability extends all of the way to that point. That way if you relied on any measurements that were possibly inaccurate, you know about it and you can evaluate any possible adverse effects on your product. CYA:ca: