Calibrating a Gage Block In-house with Calibrated Digital Calipers
O
ollie107
Hi there,
I have just moved to a lab where they have recently had their calipers callibrated. Now they have sourced a rectangular piece of stainless steel (home made gage block), drawn on lines in permanent marker so that there are 3 locations of differing measurement that are always made in the same place, then carried out repeated measurements on this home made gage block with their recently calibrated calipers. This has given them consistant readings, and are now proposing to use this supposedly calibrated gage block to calibrate the same calipers in the future.
The calipers have been calibrated and have shown a measurement of uncertainty of +/- 0.0161mm. The item that we want to measure with the calipers are sieves, with the smallest diameter to measure is 4mm, and according to the procedure needs to be measured within an accuracy of 5%, which is 0.20mm. Now usually I go by the theory that anything that calibrates an item needs to have a count of at least 1/10th of the item being calibrated. However in this case, we dont need the calipers to be that accurate as they only need to be accurate to 0.20mm.
Can someone please confirm whether this method is acceptable in ISO 17205, or as I suspect we need to either need to send the block off to be calibrated.
Any comments are appreciated.:)
:thanx:
Welcome to the Cove. :bigwave:
If I understand this correctly, you are calibrating calipers, then using them to verify a custom standard, and then using that standard to calibrate the calipers. The circularity of this strategy negates the traceability of the original calibration of the calipers, assuming that such traceability existed to begin with. A piece of measuring equipment can't be traceable to itself.
O
ollie107
Many thanks for that,
O.K so if we continue to have the calipers calibrated by an external lab annually, can we use this home made gage block for intermediate checks throughout the year?
THen if we find the calipers to be out in respect to the gage block between annual calibrations, then we send the calipers to be calibtrated. What do you think? How does everyone else check that their calipers are within limits between calibrations?
:confused:
:confused:
Commercial gage blocks.
bobdoering
Stop X-bar/R Madness!!
It has to be calibrated traceable to a national standard. It would also have to have tolerances to be used as a gage, and they have to be either 10:1 or 4:1 (depending on your take on calibration) to the caliper.
What do you think?
Not such a good idea, unless you have a heck of a superfinishing machine. You need parallelism and surface finish on that block that it pretty extraordinary to be of much value. Plus, it needs to be made of material that will not wear and is thermally stable to be of any value.
Either commercial gage blocks, as Jim mentions, or set masters
Hershal
Metrologist-Auditor
Hi there,
:thanx:
The home made gage block can be turned into a check device, used solely for a sanity check.
If the item is to be used as a gage block for calibration, then it needs to go to an accredited calibration laboratory, and the cert needs to report the uncertainties for each of the measurement points. Then make sure it is kept in a secure and clean area, and human hands never touch it again by using cloves. Then it can be used for calibration.
Hope this helps.
bobdoering
I can't imagine they would touch it without some kind of specification, similar to the specifications used to identify grades of gage blocks. They need to know its nominal size and its specified tolerance. That tolerance needs to relate correctly to its use as a standard for that calibration (to make the excercise worth the cost). I sense that has not been prepared.
jfgunn
I think the more important question should be:
Is a caliper the appropriate measuring tool to measure the diameter of the hole in a sieve?
I believe the answer to this is NO. I believe an optical comparator would be the correct measuring instrument for this purpose.
Measuring an internal diameter with flat ID jaws of a caliper does not work well.
If you are going to use this method I would buy a couple of real gage blocks for $50 - $100, then pay the $20-$30/year to have your caliper calibrated and the $20-$30 to have the few gage blocks calibrated as well. (including uncertainties of course).
L Calibrating a Template Gage General Measurement Device and Calibration Topics 7
S Calibrating a 1/8-27 NPT thread gage General Measurement Device and Calibration Topics 4
M Calibrating 1/2-20 UNF 2B STI Threaded Plug Gage General Measurement Device and Calibration Topics 3
Z Calibrating Sunnen PG800 precision bore gage & PG700 setting fixtures General Measurement Device and Calibration Topics 1
C Gage pin calibration procedure for accurately calibrating gage pins General Measurement Device and Calibration Topics 18
M Calibrating Angle Gage Blocks General Measurement Device and Calibration Topics 2
J Thread Wire Chart to use when Calibrating a particular size Thread Gage General Measurement Device and Calibration Topics 16
J Gage Pin Quality - Buying new pin sets cheaper than calibrating our current sets General Measurement Device and Calibration Topics 6
F Calibrating Gagemaker Products General Measurement Device and Calibration Topics 1
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
I Tool maintenance and calibration procedure - Calibrating tools/equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Crusader Interval for Self-Calibrating Measurement Device? Calibration Frequency (Interval) 15
A Calibrating a PRT sensor - Temperature Sensitivity Coefficient General Measurement Device and Calibration Topics 2
M Calibrating Digital Pressure Gages In-House General Measurement Device and Calibration Topics 10
F OEM has no procedure for calibrating their device General Measurement Device and Calibration Topics 4
B Problem Calibrating a Krautkramer DMS 2 Ultrasonic Thickness Meter Calibration Frequency (Interval) 2
J Calibrating an 80,000 psi Fluid Pressure Gauge General Measurement Device and Calibration Topics 6
S Calibrating Thermometer with Simulated Output Temperature from Fluke 5520A General Measurement Device and Calibration Topics 6
hogheavenfarm Is there a relatively inexpensive method of Calibrating our Torque Wrenches? General Measurement Device and Calibration Topics 7
L Calibrating Tool Makers Personal Tools General Measurement Device and Calibration Topics 9
P Exempt from Calibrating our Tape Measures and Micrometers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 32
C Best Method for Calibrating Multimeters General Measurement Device and Calibration Topics 8
D How much Deviation is Acceptable when Calibrating a Hygrometer? General Measurement Device and Calibration Topics 3
K Calibrating Plain Tapered Bores at the Small End Diameter General Measurement Device and Calibration Topics 7
M Calibrating a Weight Scale in a Laboratory Calibration Frequency (Interval) 4
S Calibrating Panel Mounted Pressure Gauges located on Hydraulic Test Stands General Measurement Device and Calibration Topics 3
D Calibrating a Thermometer for use in Cold Rooms - ISO 9001 Clause 7.6 General Measurement Device and Calibration Topics 8
T Hands-On training on calibrating equipment General Measurement Device and Calibration Topics 2
L Calibrating Press Brakes, How to do? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
H Calibrating a Matheson Flowmeter-help appreciated General Measurement Device and Calibration Topics 2
Jerry Eldred Calibrating Mitutoyo SJ-201P Surface Roughness Testers General Measurement Device and Calibration Topics 4
J Calibrating Electronic Equipment Tester Software - ISO 9001 Clause 7.6 Benchmarking 1
A Calibrating 4" and larger Micrometers General Measurement Device and Calibration Topics 6
R Checking (Calibrating) Truncated Thread Set Plugs General Measurement Device and Calibration Topics 11
G Calibrating a reflectometer - Scale uncertainty with reading? Measurement Uncertainty (MU) 1
D Calibrating Lufkin ID Micrometer ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
Crusader Self-Calibrating Furnace does not require calibration General Measurement Device and Calibration Topics 15
B Using & Calibrating Ultrasonic Micrometer General Measurement Device and Calibration Topics 4
R Validation of A Procedure for Calibrating Dial Calipers using Traceable Gauge Blocks General Measurement Device and Calibration Topics 6
J Measurement Uncertainty for Calibrating Custom Check Fixtures with a Portable CMM Measurement Uncertainty (MU) 3
S Calibrating Biomed equipment in an ISO 17025 accredited calibration laboratory ISO 17025 related Discussions 5
T Calibrating or Verifying Metal or Steel Tape Measures General Measurement Device and Calibration Topics 195
T Calibrating the calibrator (torque test) - Electronic torque tester CDI 4001-0-ETT General Measurement Device and Calibration Topics 3
R Calibrating Leak Testers - Old fashioned water tank type tester General Measurement Device and Calibration Topics 2
P Calibrating Density Cups & Viscosity Cups In-House - Seeking Procedure / Help General Measurement Device and Calibration Topics 2
R Micrometer resolution for calibrating thread plug gages General Measurement Device and Calibration Topics 7