Calibrating a Thermometer for use in Cold Rooms - ISO 9001 Clause 7.6

D

DAWELAH

Dear Members,‎

Hi,‎

How do I ensure that requirement 7.6 is implemented in cold rooms? ‎

I use external thermometers to verify temperature taken from cold room readings. Should I go through planned calibration process to ensure that thermometer is giving accurate results? ‎Keeping in mind that thermometer costs $ 20 and calibration fee is around $90 ‎
‎ ‎
Thanks ‎
 
Last edited by a moderator:

somashekar

Leader
Admin
Re: Calibrating a thermometer for use in cold rooms.

Dear Members,‎

Hi,‎

How do I ensure that requirement 7.6 is implemented in cold rooms? ‎

I use external thermometers to verify temperature taken from cold room readings. Should I go ‎throw planed calibration process to ensure that thermometer is giving accurate results? ‎Keeping in mind that thermometer costs $ 20 and calibration fee is around $90 ‎
‎ ‎
Thanks ‎
Cold rooms (2 - 6 'C) are mostly the walk in type and you use this for storage of your temperature sensitive materials (correct me)
You calibrate the controller and check the inside temperature at areas where you store your materials. I guess you have an internal procedure to do this calibration of cold room. You must off course have a valid calibration for your thermometer before you do this.
I do not know much about your calbration costs, however this costs me around $20 here. Irrespective of the costs you have to use a valid instrument to implement your calibration procedure with its traceability. Remember that this is your secondary master instrument traceable to your calibration agency primary master. we use here an Extech 421501.
 
S

somerqc

Forgive me if the needs of 17025 prevent this (i.e. they won't accept it), but, in a former life we simply boiled water and stuck a thermometer in it to verify it (NOTE: Boiling water is 212F or 100C - or it becomes steam).

If it didn't read within specification (we had +/- .2 degrees C), we replaced it with a new one.

John
 

Wes Bucey

Prophet of Profit
FWIW:
I have seen some operations (where temperature maintenance is critical) move up to recording thermometers which maintain a permanent record of readings throughout the day, often coupled with an alarm to signal nonconforming temperature rise or fall beyond specified parameters. Such recording thermometers do need periodic professional checking to determine they are in calibration. Exactly who performs and how that checking proceeds (including frequency) may be matter of discussion between thermometer manufacturer and historical record. Some calibrate as frequently as monthly, others annually, all according to a FMEA (Failure Mode & Effects Analysis) based on how severely a nonconforming temperature may affect the product and experience with a thermometer remaining in or straying from calibration.
 

BradM

Leader
Admin
Sorry I did not reply earlier. I was having thread issues.:D

Calibration is a service. Sometimes services (especially good ones) are not cheap. Sometimes it costs more to calibrate an item than it's worth. But you need to have confidence in the device.

I think you need to calibrate the thermometers prior to use, and prior to disposal. After all... what if the thermometer drifted out prior to disposition? You would have issues, and you would not even know.

I would be much more attuned to getting good calibration information on the thermometers, and extending the calibration recall as far as you can. You might even consider setting up an inter comparison system. Track the variance of your thermometers; after you get some history and data, you can extend the period so you can save money.

As far as ice and boiling water, it sure beats nothing.:D However, my experience is that you miss the cardinal points that you are interested in. Too, there is a lot of variance due to probe location, heat source, uniformity, etc. when using boiling water. A properly build ice bath is a good standard, assuming it is constructed properly.

As far as needing calibration, of course, this is coming from a calibration guy.:tg: If the correct temperature reading in important or relevant, then it should be calibrated.
 
S

step30044

As far as ice and boiling water, it sure beats nothing.:D However, my experience is that you miss the cardinal points that you are interested in. Too, there is a lot of variance due to probe location, heat source, uniformity, etc. when using boiling water. A properly build ice bath is a good standard, assuming it is constructed properly.

I would consider this a verification not a calibration. I would also suspect with a $20 thermometer you would want data specific to your process tolerances not data on the extremes. Brad you make a good point on the construction of the ice bath, you would be surprised at how many people dont understand how to create a proper ice bath.
 
Last edited by a moderator:

BradM

Leader
Admin
I would consider this a verification not a calibration. I would also suspect with a $20 thermometer you would want data specific to your process tolerances not data on the extremes. Brad you make a good point on the construction of the ice bath, you would be surprised at how many people don't understand how to create a proper ice bath.

Good post; agreed.:agree1:

There is an attachment here at the Cove about how to build a proper ice bath, FWIW.
 

harry

Trusted Information Resource
Old story but need to repeat - what is the tolerance and what is the risk involved?

Many requirements relating to temperature in the food industry (especially storage) are guided by rules of thumb and +/- 1 to 2 degrees are often acceptable. That being the case, an ice bath not to technical details but with plenty of ice and enough water would normally be accurate up to +/- 1/2 to 1 degree and deemed good enough for general usage. The same goes to those portable thermometer if used for verification purpose.
 
Thread starter Similar threads Forum Replies Date
S Calibrating Thermometer with Simulated Output Temperature from Fluke 5520A General Measurement Device and Calibration Topics 6
R Calibrating Solid Thread Rings with HILO master General Measurement Device and Calibration Topics 1
Slo_Glenn Calibrating Clocks General Measurement Device and Calibration Topics 13
I Calibrating the monitoring device but not the unit General Measurement Device and Calibration Topics 7
F Calibrating Taper Sleev/Plug Gauge General Measurement Device and Calibration Topics 0
F Calibrating Gagemaker Products General Measurement Device and Calibration Topics 1
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
R Feeler Gauge Calibration - What tolerance is used for calibrating feeler gauges? General Measurement Device and Calibration Topics 4
I Tool maintenance and calibration procedure - Calibrating tools/equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Crusader Interval for Self-Calibrating Measurement Device? Calibration Frequency (Interval) 15
A Calibrating a PRT sensor - Temperature Sensitivity Coefficient General Measurement Device and Calibration Topics 2
M Calibrating Digital Pressure Gages In-House General Measurement Device and Calibration Topics 10
F OEM has no procedure for calibrating their device General Measurement Device and Calibration Topics 4
N Verifying Calibration of Hardness Gauge vs Actually Calibrating It General Measurement Device and Calibration Topics 2
B Problem Calibrating a Krautkramer DMS 2 Ultrasonic Thickness Meter Calibration Frequency (Interval) 2
J Calibrating an 80,000 psi Fluid Pressure Gauge General Measurement Device and Calibration Topics 6
hogheavenfarm Is there a relatively inexpensive method of Calibrating our Torque Wrenches? General Measurement Device and Calibration Topics 7
L Calibrating Tool Makers Personal Tools General Measurement Device and Calibration Topics 9
P Exempt from Calibrating our Tape Measures and Micrometers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 32
L Calibrating a Template Gage General Measurement Device and Calibration Topics 7
C Best Method for Calibrating Multimeters General Measurement Device and Calibration Topics 8
D How much Deviation is Acceptable when Calibrating a Hygrometer? General Measurement Device and Calibration Topics 3
K Calibrating Plain Tapered Bores at the Small End Diameter General Measurement Device and Calibration Topics 7
M Calibrating a Weight Scale in a Laboratory Calibration Frequency (Interval) 4
S Calibrating Panel Mounted Pressure Gauges located on Hydraulic Test Stands General Measurement Device and Calibration Topics 3
T Hands-On training on calibrating equipment General Measurement Device and Calibration Topics 2
L Calibrating Press Brakes, How to do? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
H Calibrating a Matheson Flowmeter-help appreciated General Measurement Device and Calibration Topics 2
Jerry Eldred Calibrating Mitutoyo SJ-201P Surface Roughness Testers General Measurement Device and Calibration Topics 4
O Calibrating a Gage Block In-house with Calibrated Digital Calipers General Measurement Device and Calibration Topics 7
L Smallest Gage block - What is the smallest gage block for calibrating micrometers? General Measurement Device and Calibration Topics 10
S Calibrating a 1/8-27 NPT thread gage General Measurement Device and Calibration Topics 4
J Calibrating Electronic Equipment Tester Software - ISO 9001 Clause 7.6 Benchmarking 1
M Calibrating 1/2-20 UNF 2B STI Threaded Plug Gage General Measurement Device and Calibration Topics 3
A Calibrating 4" and larger Micrometers General Measurement Device and Calibration Topics 6
R Checking (Calibrating) Truncated Thread Set Plugs General Measurement Device and Calibration Topics 11
G Calibrating a reflectometer - Scale uncertainty with reading? Measurement Uncertainty (MU) 1
D Calibrating Lufkin ID Micrometer ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
Z Calibrating Sunnen PG800 precision bore gage & PG700 setting fixtures General Measurement Device and Calibration Topics 1
C Gage pin calibration procedure for accurately calibrating gage pins General Measurement Device and Calibration Topics 18
Crusader Self-Calibrating Furnace does not require calibration General Measurement Device and Calibration Topics 15
B Using & Calibrating Ultrasonic Micrometer General Measurement Device and Calibration Topics 4
M Calibrating Angle Gage Blocks General Measurement Device and Calibration Topics 2
R Validation of A Procedure for Calibrating Dial Calipers using Traceable Gauge Blocks General Measurement Device and Calibration Topics 6
J Measurement Uncertainty for Calibrating Custom Check Fixtures with a Portable CMM Measurement Uncertainty (MU) 3
S Calibrating Biomed equipment in an ISO 17025 accredited calibration laboratory ISO 17025 related Discussions 5
T Calibrating or Verifying Metal or Steel Tape Measures General Measurement Device and Calibration Topics 210
T Calibrating the calibrator (torque test) - Electronic torque tester CDI 4001-0-ETT General Measurement Device and Calibration Topics 3
R Calibrating Leak Testers - Old fashioned water tank type tester General Measurement Device and Calibration Topics 2
P Calibrating Density Cups & Viscosity Cups In-House - Seeking Procedure / Help General Measurement Device and Calibration Topics 2

Similar threads

Top Bottom