Hi Folks! I have worshiped you all from afar for awhile now, but just recently joined and this is my first post.

My question is this: is there any way to justify not calibrating the meters and thingamabobs on welding machines?

My Test Lab Manager, who oversees this calibration and for whom I have a great deal of respect based upon his experience and qualifications does not see this as a value-adding activity. I've received additional comments from the Manufacturing Manager, QC Manager, Welding Manager and Maintenence Manager all of whom claim to have a lot of experience in this area.

We are a custom manufacturer of power distribution switchgear and utilize welded construction. We certify our welders in-house within the criteria of AWS and do periodic destructive testing but nothing hardcore like x-rays and such.

I do not know a lot about welding yet, so I'm relying on what others tell me. Plus I'm a newbie to the QA field (went from Production Engineer to QA Manager in March and had to transition our entire QMS beast for the upgrade audit this week because the past two managers didn't do a lick of work) so I haven't had any time to address/research/think/dream about this particular issue.

I'm told that power flucuations(sp?) affect the machines and therefore welders use their own experience and rarely rely on the meters on the machine. I've told everyone we are maintaining the status quo and continuing calibration until I can feret out a solution, if any. Any advice and/or suggestions are welcome.
 
Elsmar Forum Sponsor

SteelMaiden

Super Moderator
Super Moderator
#2
Don't know if this will be of any use, but here goes. Most welders do not go by the gages, except for a set up kind of thing, once they strike an arc, they readjust by their experience, and really don't pay that much attention.
 
P

p_tww

#3
I'm told that power flucuations(sp?) affect the machines and therefore welders use their own experience and rarely rely on the meters on the machine. I've told everyone we are maintaining the status quo and continuing calibration until I can feret out a solution, if any. Any advice and/or suggestions are welcome.[/QUOTE]
Hi, Welding is a special process, you must validate all the manufacturing element such as operator, equipment/meters,process,material,environment.
If your welding process is a continuous processing and defect will be consequent detected on the line, you do not need calibrate related meters used for welding machine.
 
M

ML Retcher

#4
is there any way to justify not calibrating the meters and thingamabobs on welding machines?

This is the means for the welder to verify his set up (a requirement). Calibration of weld equipment will only enable your AWS certified welders to adjust the equipment within the specified ranges to achieve a better weld. If your equipment is not calibrated your ranges for your settings are not a true value.
 
R

Randy Stewart

#5
Yes and no,
I agree with the comments that they provide reference information to the welders. However, depending on the type of weld it could be a serious problem.

Before anyone jumps my case, let me say that I'm not suggesting calibration. We were experiencing cold welds from our spot welders, since you can't pull test spot welds on aluminium we ended up sending bad parts to our customer. Who, by the way, checked our welds by doing pull tests!
We installed a device that gives a print out of the current and voltage at the time of weld and will stop the weld if it is under a set amount.
 
E

energy

#6
How now....

Christi Day said:
Hi Folks! I have worshiped you all from afar for awhile now, but just recently joined and this is my first post.

My question is this: is there any way to justify not calibrating the meters and thingamabobs on welding machines?

I've told everyone we are maintaining the status quo and continuing calibration until I can feret out a solution, if any. Any advice and/or suggestions are welcome.
Welcome to the Cove, Christi. :agree: How are you calibrating the machines now? What types of machines. Electric, Gas, etc.? I also agree with those that say that most welders turn their machines on and go. However, there are applications where the integrity of the welds are required to be verified, other than a visual inspection. With a little more information we may be able to solicit input from the Welding "Experts". ;)
 
C

CHESHIRE STEVE

#7
We use a Validator machine which puts a measured load on to a welding set,giving a Voltage/Amperage reading on the validator. This is compared to the readout on the dials/gauges of the welding set. We use this internally to test the welding machines that belong to our hire fleet.Our Validator is controlled within our normal calibration records, and is subject to annual external calibration. Any discrepancies are usually rectified in-house by our service team.
This ensures that the machines we hire out perform as they say they do. We have not been asked for proof of calibration yet, and we have been hiring equipment out for around 15 years.
 
Last edited by a moderator:

CarolX

Super Moderator
Super Moderator
#8
Hi Christi and Welcome to the Cove:bigwave: ,


I will just repeat most of what has been said here. It depends on what your doing. I work for a sheet metal job shop. All our work is commercial, we build mostly cabinets and enclosures. We do not calibrate our MIG/TIG welding machines. We investigated the process at one time, and found it would not be of any value added, as our welds are not structural.

So to answer your question...what are you welding?

Carol
 

howste

Thaumaturge
Super Moderator
#9
I know that you posted this in the 17025 forum, but I think the key here is what 7.6 in ISO 9001 says: "where necessary to ensure valid results, measuring equipment shall be calibrated or verified..."

Assuming you're talking about TIG, MIG, or some type of stick welding, I would argue that it's possible to ensure that the welds are good without having calibrated equipment. I'm currently working with a client that manufactures overhead cranes, where the CMAA requires AWS certified welders for all structural welds. In an audit last week, we reviewed many of the requirements of AWS D1.1. Nowhere in that document did I read anything about calibrated equipment - the requirements are focused on the results of the welding performed.

If the welders themselves are really qualified, they should be able tell while they're welding if they're getting a good weld. If the equipment went haywire, they should be able to tell because they wouldn't be able to get proper penetration, etc. Your periodic destructive testing should confirm that the welds are good. Have you ever had destructive testing failures?

The above doesn't apply if you're doing resistance welding...
 
G

gburns

#10
Since the welder doesn't use the meters as process control, and since the meters aren't part of your product acceptance criteria, I see no reason why they'd need to be calibrated.
 
Thread starter Similar threads Forum Replies Date
F Calibrating Taper Sleev/Plug Gauge General Measurement Device and Calibration Topics 0
F Calibrating Gagemaker Products General Measurement Device and Calibration Topics 1
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
R Feeler Gauge Calibration - What tolerance is used for calibrating feeler gauges? General Measurement Device and Calibration Topics 4
I Tool maintenance and calibration procedure - Calibrating tools/equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Crusader Interval for Self-Calibrating Measurement Device? Calibration Frequency (Interval) 15
A Calibrating a PRT sensor - Temperature Sensitivity Coefficient General Measurement Device and Calibration Topics 2
M Calibrating Digital Pressure Gages In-House General Measurement Device and Calibration Topics 10
F OEM has no procedure for calibrating their device General Measurement Device and Calibration Topics 4
N Verifying Calibration of Hardness Gauge vs Actually Calibrating It General Measurement Device and Calibration Topics 2
B Problem Calibrating a Krautkramer DMS 2 Ultrasonic Thickness Meter Calibration Frequency (Interval) 2
J Calibrating an 80,000 psi Fluid Pressure Gauge General Measurement Device and Calibration Topics 6
S Calibrating Thermometer with Simulated Output Temperature from Fluke 5520A General Measurement Device and Calibration Topics 6
hogheavenfarm Is there a relatively inexpensive method of Calibrating our Torque Wrenches? General Measurement Device and Calibration Topics 7
L Calibrating Tool Makers Personal Tools General Measurement Device and Calibration Topics 9
P Exempt from Calibrating our Tape Measures and Micrometers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 32
L Calibrating a Template Gage General Measurement Device and Calibration Topics 7
C Best Method for Calibrating Multimeters General Measurement Device and Calibration Topics 8
D How much Deviation is Acceptable when Calibrating a Hygrometer? General Measurement Device and Calibration Topics 3
K Calibrating Plain Tapered Bores at the Small End Diameter General Measurement Device and Calibration Topics 7
M Calibrating a Weight Scale in a Laboratory Calibration Frequency (Interval) 4
S Calibrating Panel Mounted Pressure Gauges located on Hydraulic Test Stands General Measurement Device and Calibration Topics 3
D Calibrating a Thermometer for use in Cold Rooms - ISO 9001 Clause 7.6 General Measurement Device and Calibration Topics 8
T Hands-On training on calibrating equipment General Measurement Device and Calibration Topics 2
L Calibrating Press Brakes, How to do? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
H Calibrating a Matheson Flowmeter-help appreciated General Measurement Device and Calibration Topics 2
Jerry Eldred Calibrating Mitutoyo SJ-201P Surface Roughness Testers General Measurement Device and Calibration Topics 4
O Calibrating a Gage Block In-house with Calibrated Digital Calipers General Measurement Device and Calibration Topics 7
L Smallest Gage block - What is the smallest gage block for calibrating micrometers? General Measurement Device and Calibration Topics 10
S Calibrating a 1/8-27 NPT thread gage General Measurement Device and Calibration Topics 4
J Calibrating Electronic Equipment Tester Software - ISO 9001 Clause 7.6 Benchmarking 1
M Calibrating 1/2-20 UNF 2B STI Threaded Plug Gage General Measurement Device and Calibration Topics 3
A Calibrating 4" and larger Micrometers General Measurement Device and Calibration Topics 6
R Checking (Calibrating) Truncated Thread Set Plugs General Measurement Device and Calibration Topics 11
G Calibrating a reflectometer - Scale uncertainty with reading? Measurement Uncertainty (MU) 1
D Calibrating Lufkin ID Micrometer ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
Z Calibrating Sunnen PG800 precision bore gage & PG700 setting fixtures General Measurement Device and Calibration Topics 1
C Gage pin calibration procedure for accurately calibrating gage pins General Measurement Device and Calibration Topics 18
Crusader Self-Calibrating Furnace does not require calibration General Measurement Device and Calibration Topics 15
B Using & Calibrating Ultrasonic Micrometer General Measurement Device and Calibration Topics 4
M Calibrating Angle Gage Blocks General Measurement Device and Calibration Topics 2
R Validation of A Procedure for Calibrating Dial Calipers using Traceable Gauge Blocks General Measurement Device and Calibration Topics 6
J Measurement Uncertainty for Calibrating Custom Check Fixtures with a Portable CMM Measurement Uncertainty (MU) 3
S Calibrating Biomed equipment in an ISO 17025 accredited calibration laboratory ISO 17025 related Discussions 5
T Calibrating or Verifying Metal or Steel Tape Measures General Measurement Device and Calibration Topics 191
T Calibrating the calibrator (torque test) - Electronic torque tester CDI 4001-0-ETT General Measurement Device and Calibration Topics 3
R Calibrating Leak Testers - Old fashioned water tank type tester General Measurement Device and Calibration Topics 2
P Calibrating Density Cups & Viscosity Cups In-House - Seeking Procedure / Help General Measurement Device and Calibration Topics 2
R Micrometer resolution for calibrating thread plug gages General Measurement Device and Calibration Topics 7
J Thread Wire Chart to use when Calibrating a particular size Thread Gage General Measurement Device and Calibration Topics 16

Similar threads

Top Bottom