Calibration - An old textile plant with a gazillion gauges on our machines

D

DAdams

#1
Calibration

We are an old textile plant with a gazillion gauges on our machines. Currently we have the yardage clocks and temperature guages on all machines calibrated. Every 1000 yard roll we produce is tested in the lab for shrinkage, color, weight etc. before it is inspected and shipped to our customers. These test take place while the machine is running. All the test equipment in the lab is calibrated by an outside source. If a product fails a lab test the machine operator is notified and adjustments are made (speed, temperature, etc.) These guages on the machines are used as a guide or "starting point". Do we have to have all the guages in the plant calibrated for QS9000?
 
Elsmar Forum Sponsor

Jerry Eldred

Forum Moderator
Super Moderator
#2
No you don't have to have ALL gauges in the plant calibrated to satisfy QS9000.

***I SAY THAT VERY CAREFULLY***

You MUST, however, either calibrate or account in some way for assuring that any or all of the gauges that could in any affect the product quality or it's ability to satisfy specifications read correctly.

There are a number of variations on how this can be done, and many variations on the most intelligent way of doing, depending of course on the particulars of your processes.

If, for example, there are gauges which are purely for "convenience", such as a pressure gauge for a pneumatic function, which, if they were totally incorrect in their reading would still not have any possible impact on the product, those could be INDICATION ONLY. In semiconductors, there are tens of thousands of gauges every where you look. Only the trained professionals can determine wihch of them are important to the process.

I wouldn't dare speculate which of your gauges could impact product quality. But those engineers, etc. at your factory should be able to make that determine. My caution would be to make sure each decision is intelligently made such that you could convince an auditor. If you use the reading on a gauge to set any parameter in your control plan, it certainly ought to be calibrated. If a gauge measures a specified parameter, it needs to be calibrated. If there is a safety issue involved, it should be calibrated. If you have reading limits for proper operation of a piece of production equipment, it should be calibrated. The pneumatic example above-- Let's say your pneumatic cutting blade operates at 100 psi (I may not have a good number there). If there is a lower limit to where the cutting blade will function (let's say 80 psi) and an upper limit before hoses spring leaks (let's say 150 psi), it should be calibrated.

A way around calibrating a lot of the peripheral gauges is to use S.P.C. on the process. If there are 20 gauges on a production tool that you don't believe have a significant impact on the process output, set up a chart and plot their readings daily, every shift, every week, or whatever works well). Have a calibrated measurement of the output of the process. Plot that output on a control chart. Use those gauge readings as feedback. When the control chart drifts toward an out-of-control condition, check for a similar trend in the gauge readings. You have not only eliminated the calibration requirement for 20 gauges, but used them as a trending tool for improved process control. This can actually be more effective, and less costly than traditional calibration schemes (in specific circumstances). This will not eliminate the need to calibrate those gauges that truly do need it.



------------------
 
Thread starter Similar threads Forum Replies Date
M Old calibration standard methods - Question for ISO 17025 auditors..... ISO 17025 related Discussions 1
S Validity of 7 Year Old Calibration Certificate General Measurement Device and Calibration Topics 19
B UV-A and Illuminance meter Calibration Setup General Measurement Device and Calibration Topics 0
S Calibration Related Posters General Measurement Device and Calibration Topics 2
F Quality Control in Electrical Calibration ISO 17025 related Discussions 0
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1

Similar threads

Top Bottom