What should a valid calibration certificate for a new piece of equipment look like? I'm particularly interested in a Temperature/Humidity monitor & logger.
 
Elsmar Forum Sponsor

Helmut Jilling

Auditor / Consultant
#2
Re: Calibration Certificate for New Equipment

What should a valid calibration certificate for a new piece of equipment look like? I'm particularly interested in a Temperature/Humidity monitor & logger.
The same information that a calibration cert for a used piece of equipment should contain. There is no real difference. It is a calibration of an instrument that determines the gage can make reliable readings.
 

BradM

Staff member
Admin
#3
Re: Calibration Certificate for New Equipment

The same information that a calibration cert for a used piece of equipment should contain. There is no real difference. It is a calibration of an instrument that determines the gage can make reliable readings.
Agreed. Just roughly:

1. Customer information; company information
2. Date/ time, recall information (if applicable).
3. Standards used information
4. Procedure/ guideline listed
5. Instrument under test specific information
6. Data test points, values observed, tolerances
7. Comments, observations
8. Date, signature, review, etc.

As far as the monitor/logger, it depends. A factory calibration may check voltages, board readings, cold junction compensators, etc. Another may check one channel with a simulator; yet another may check all input channels. So the data could be a paragraph, or three pages long.

Al, is any of that helpful?
 

Jerry Eldred

Forum Moderator
Super Moderator
#4
Re: Calibration Certificate for New Equipment

I won't reiterate what the others have already said. On a certificate for a temp/humidity logger, I would look for three humidity point (low/mid/high), typically somewhere around 20%RH, 50%RH, 80%RH or so. I would look for at least two (preferrably three) temperature data points with ambient being somewhere around the middle.

If there is just a single point cal (particularly on humidity), the accuracy is suspect, as humidity sensors are notorious for non-linearity.
 
D

Daniel Walker - 2011

#5
MEASUREMENT UNCERTAINTY!!!!! the 17025 forum and nobody mentions the need for an uncertainty value? :mg:
 
J

JAltmann

#7
I would also look for whether or not the calibration was accredited or not, meaning performed to ISO-17025. Not that non-accredited may or may not be good, it adds a level of comfort that the calibration was done properly.
 

Jim Wynne

Staff member
Admin
#8
hi, what if a new equipment with calibration certificate is not put in use but kept in storage for 3 months. after that it is used, When will be the date of next Calibration if the periodicity is One year. Is it one year from the date from which it was used or is it one year from the date of manufacture.?
It's your choice, but the device should be calibrated before you release it for production use, regardless of the calibration certificate that it came with. You have no way of knowing its calibration status otherwise.
 

Tagin

Trusted Information Resource
#9
As regards this, what if the Tape was not put to use say for 3 months after being bought along with Calibration certificate.Now when it should be calibrated? Is it one year from the Start date of Actual Usage or one year from the date of Calibration as mentioned in the certificate. Assume the tape was preserved well while in Storage. I have this issue in my organisation. how to answer th auditor reg this.thank u
I think calibration intervals have to be determined by the organization using at least these criteria:
  • Expiration interval of calibration certificate
  • Frequency of use of calibrated device
  • Likelihood of damage and/or wear&tear to calibrated device
  • Risk of producing product with out-of-calibration device
  • Risks (financial, safety, reputation, etc. risks) from delivering product produced with out-of-calibration device
So just going by the expiration on the calibration cert is not really adequate. If you can answer the above criteria for the auditor, I think they will be satisfied.
 
Thread starter Similar threads Forum Replies Date
F Calibration certificate content - Average Value and True Value ISO 17025 related Discussions 6
E Calibration Certificate Content Requirements General Measurement Device and Calibration Topics 3
C Accredited Calibration Certificate Content - ISO 17025 Section 5.10.4.2 Question ISO 17025 related Discussions 7
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
M How to read a calibration certificate for thermo hygrometer General Measurement Device and Calibration Topics 2
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
A Calibration Certificate Results General Measurement Device and Calibration Topics 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate Qualification and Validation (including 21 CFR Part 11) 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate General Measurement Device and Calibration Topics 2
G Is the calibration due date of equipment used to be on calibration certificate ISO 17025 related Discussions 38
G Calibration Certificate, fancy paper or not? General Measurement Device and Calibration Topics 9
Crusader Certificate of calibration from Manufacturer allowed? General Measurement Device and Calibration Topics 14
T Calibration Certificate Readings After Adjustment General Measurement Device and Calibration Topics 5
C Calibration Certificate "As Found" condition - TS16949 Section 7.6.2 General Measurement Device and Calibration Topics 7
M ILAC Regulations - Expression of Units on the Calibration Certificate General Measurement Device and Calibration Topics 6
A Coverage factor in calibration certificate for a multimeter General Measurement Device and Calibration Topics 4
N Client Request for Reference Equipment Calibration Certificate General Measurement Device and Calibration Topics 3
T Microsoft Access Database to create a Simple Calibration Certificate General Measurement Device and Calibration Topics 6
V How to use a Calibration Certificate in Uncertainty Calculation for a Test Measure Measurement Uncertainty (MU) 13
T Uncertainty in Hygrometer Calibration Certificate General Measurement Device and Calibration Topics 4
O How can I use reading from calibration certificate in measurement Calibration Frequency (Interval) 7
K New Gauges calibration certificate shows deviations from the measurement uncertainty IATF 16949 - Automotive Quality Systems Standard 5
Q Can we go by different Calibration Frequency vs. What's on the Certificate ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
R Torque Tester Calibration Traceability and Certificate question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S No Calibration Due Date on a Calibration Certificate unless requested by a customer? Calibration Frequency (Interval) 9
P Rounding Uncertainty to Zero on a Calibration Certificate Measurement Uncertainty (MU) 7
N Calibration Lab unable to issue Accredited Certificate when error is too high? General Measurement Device and Calibration Topics 3
S How to Evaluate a Calibration Report Certificate General Measurement Device and Calibration Topics 8
R LCP 01201P and ICPF2P Calibration Methods Stated in Calibration Certificate General Measurement Device and Calibration Topics 2
M ILAC p14:12/2010 - Accredited Calibration Certificate Uncertainty General Measurement Device and Calibration Topics 2
A Can any organization issue a Calibration Certificate? General Measurement Device and Calibration Topics 5
A No Uncertainty Listed on Calibration Certificate: Is it traceable calibration? General Measurement Device and Calibration Topics 8
M External Calibration Certificate Review ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
N Calibration Certificate Requirements for Pressure Gauges General Measurement Device and Calibration Topics 9
A Fluke Digital MultiMeter Calibration Traceability and Certificate General Measurement Device and Calibration Topics 11
S Reference to original Calibration Certificate - ISO 17025 Section 5.10.9 ISO 17025 related Discussions 4
howste Calibration Certificate Contents: How to report Measurement Uncertainty? Measurement Uncertainty (MU) 18
D Calibration Certificate - 'test report shall not be reproduced except in full' General Measurement Device and Calibration Topics 2
howste Condition of Item Calibrated - ISO 17025 Clause 5.10.2f - Calibration Certificate ISO 17025 related Discussions 13
C Interpreting Outside Laboratory Calibration Certificate Measurement Uncertainty (MU) 3
R What to do after I receive a calibration certificate General Measurement Device and Calibration Topics 4
E Calibration - Manufacturer's Calibration Certificate Date ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
S Verification of calibration certificate Miscellaneous Environmental Standards and EMS Related Discussions 1
B ISO 17025 logo & ISO 9001 logo on calibration certificate/report ISO 17025 related Discussions 7
P Certificate of Calibration Suppliers for an ISO 9001 company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
S Validity of 7 Year Old Calibration Certificate General Measurement Device and Calibration Topics 19
BradM Certificate for In-House Calibration General Measurement Device and Calibration Topics 25
Wayne Never put an Expiration (Due) Date on a Calibration Certificate Calibration Frequency (Interval) 41
J In-House Calibration Certificate wording - Confidence Level or Level of Confidence? Document Control Systems, Procedures, Forms and Templates 3
A Measurement uncertainity is not mentioned in the calibration certificate Measurement Uncertainty (MU) 4

Similar threads

Top Bottom