Hi
Could someone put an end to an on-going debate please?
On our calibration certificates we have basically
Col 1. Controlled test reading
Col 2. Actual readings BEFORE adjustment
Col 3. Errors BEFORE adjustment
Col 4. Actual readings AFTER adjustment
Col 5. Errors AFTER Adjustment
If the unit does not need adjusting and is within tolerance, is it the "done thing" to complete column 4&5 with the readings from col 2&3 or should they be left blank (---) or have n/a in them. Or is it personal preference
Thanks in advance.
BradM
Hello there! Welcome to the Cove!
I guess to your question, my response would be largely opinion. I would not leave it blank. Blank spaces are not good in many industries. Plus it makes the form look incomplete.
I would recommend copying over the numbers, or otherwise marking: same as: as found, for the as-left status.
dgriffith
Does your cert need to have both as found and as left on the same page? Many certs say found/left or only as found, and I agree leaving the as left blank might be confusing.
Ex. Fluke has an as found cert with cert number, an as left cert with new cert number, or a single as found/as left cert if no adjustment is needed.
N
notadog
Hello, I Just received a cert from Keithley.
My question is should the Limits be calculated from the Test Desc (the Nominal Value) or the Expected Value (the value being generated)?
As Rec'd DataTest DescExpected ValueMeasured ValueLower LimitUpper LimitTest Status100.00099.9965100.16699.846100.148 FailShould the Limits be calculated from the Test Desc or the Expected Value?
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
F Calibration certificate content - Average Value and True Value ISO 17025 related Discussions 6
M How to read a calibration certificate for thermo hygrometer General Measurement Device and Calibration Topics 2
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
A Calibration Certificate Results General Measurement Device and Calibration Topics 3
E Calibration Certificate Content Requirements General Measurement Device and Calibration Topics 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate Qualification and Validation (including 21 CFR Part 11) 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate General Measurement Device and Calibration Topics 2
G Is the calibration due date of equipment used to be on calibration certificate ISO 17025 related Discussions 38
G Calibration Certificate, fancy paper or not? General Measurement Device and Calibration Topics 9
Crusader Certificate of calibration from Manufacturer allowed? General Measurement Device and Calibration Topics 14
C Calibration Certificate "As Found" condition - TS16949 Section 7.6.2 General Measurement Device and Calibration Topics 7
M ILAC Regulations - Expression of Units on the Calibration Certificate General Measurement Device and Calibration Topics 6
A Coverage factor in calibration certificate for a multimeter General Measurement Device and Calibration Topics 4
N Client Request for Reference Equipment Calibration Certificate General Measurement Device and Calibration Topics 3
T Microsoft Access Database to create a Simple Calibration Certificate General Measurement Device and Calibration Topics 6
V How to use a Calibration Certificate in Uncertainty Calculation for a Test Measure Measurement Uncertainty (MU) 13
T Uncertainty in Hygrometer Calibration Certificate General Measurement Device and Calibration Topics 4
O How can I use reading from calibration certificate in measurement Calibration Frequency (Interval) 7
K New Gauges calibration certificate shows deviations from the measurement uncertainty IATF 16949 - Automotive Quality Systems Standard 5
Q Can we go by different Calibration Frequency vs. What's on the Certificate ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
R Torque Tester Calibration Traceability and Certificate question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S No Calibration Due Date on a Calibration Certificate unless requested by a customer? Calibration Frequency (Interval) 9
P Rounding Uncertainty to Zero on a Calibration Certificate Measurement Uncertainty (MU) 7
N Calibration Lab unable to issue Accredited Certificate when error is too high? General Measurement Device and Calibration Topics 3
S How to Evaluate a Calibration Report Certificate General Measurement Device and Calibration Topics 8
R LCP 01201P and ICPF2P Calibration Methods Stated in Calibration Certificate General Measurement Device and Calibration Topics 2
M ILAC p14:12/2010 - Accredited Calibration Certificate Uncertainty General Measurement Device and Calibration Topics 2
A Can any organization issue a Calibration Certificate? General Measurement Device and Calibration Topics 5
A No Uncertainty Listed on Calibration Certificate: Is it traceable calibration? General Measurement Device and Calibration Topics 8
M External Calibration Certificate Review ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
N Calibration Certificate Requirements for Pressure Gauges General Measurement Device and Calibration Topics 9
A Fluke Digital MultiMeter Calibration Traceability and Certificate General Measurement Device and Calibration Topics 11
S Reference to original Calibration Certificate - ISO 17025 Section 5.10.9 ISO 17025 related Discussions 4
howste Calibration Certificate Contents: How to report Measurement Uncertainty? Measurement Uncertainty (MU) 18
C Accredited Calibration Certificate Content - ISO 17025 Section 5.10.4.2 Question ISO 17025 related Discussions 7
D Calibration Certificate - 'test report shall not be reproduced except in full' General Measurement Device and Calibration Topics 2
howste Condition of Item Calibrated - ISO 17025 Clause 5.10.2f - Calibration Certificate ISO 17025 related Discussions 13
C Interpreting Outside Laboratory Calibration Certificate Measurement Uncertainty (MU) 3
A Calibration Certificate content for New Equipment General Measurement Device and Calibration Topics 8
R What to do after I receive a calibration certificate General Measurement Device and Calibration Topics 4
E Calibration - Manufacturer's Calibration Certificate Date ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
S Verification of calibration certificate Miscellaneous Environmental Standards and EMS Related Discussions 1
B ISO 17025 logo & ISO 9001 logo on calibration certificate/report ISO 17025 related Discussions 7
P Certificate of Calibration Suppliers for an ISO 9001 company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
S Validity of 7 Year Old Calibration Certificate General Measurement Device and Calibration Topics 19
BradM Certificate for In-House Calibration General Measurement Device and Calibration Topics 25
Wayne Never put an Expiration (Due) Date on a Calibration Certificate Calibration Frequency (Interval) 41
J In-House Calibration Certificate wording - Confidence Level or Level of Confidence? Document Control Systems, Procedures, Forms and Templates 3
A Measurement uncertainity is not mentioned in the calibration certificate Measurement Uncertainty (MU) 4