Calibration Certificate Result issued by an accredited external laboratory

#1
Good afternoon,

I am writing to ask for help in managing the calibration certificates issued by an accredited external laboratory.
Following an AS9100 audit, a client of mine was asked to better highlight the acceptability of the calibration results.
I don't know if it is convenient to make a table with the criteria for each family of instruments (eg gauges, micrometers, threaded plugs, etc.) or to write a fixed sentence in the calibration report as evidence of acceptance of results. My client doesn't understand what kind of evaluations he has to make.
Thanks if you can help me.
MJ
 
Elsmar Forum Sponsor

dwperron

Trusted Information Resource
#2
Good afternoon,

I am writing to ask for help in managing the calibration certificates issued by an accredited external laboratory.
Following an AS9100 audit, a client of mine was asked to better highlight the acceptability of the calibration results.
I don't know if it is convenient to make a table with the criteria for each family of instruments (eg gauges, micrometers, threaded plugs, etc.) or to write a fixed sentence in the calibration report as evidence of acceptance of results. My client doesn't understand what kind of evaluations he has to make.
Thanks if you can help me.
MJ
AS9100D 7.1.5.1 covers this issue:

The organization shall determine and provide the resources needed to ensure valid and reliable results when monitoring or
measuring is used to verify the conformity of products and services to requirements.

The organization shall ensure that the resources provided:
a. are suitable for the specific type of monitoring and measurement activities being undertaken;
b. are maintained to ensure their continuing fitness for their purpose.

The organization shall retain appropriate documented information as evidence of fitness for purpose of the monitoring and
measurement resources.


So the organization is required to determine the "fitness of purpose" of the equipment being used and to document evidence of the fitness for purpose. How this is performed is up to each organization.
If they have done this then they can state that since the equipment being used has been calibrated to manufacturer specifications, then the results are acceptable to their requirements.
 

Tagin

Trusted Information Resource
#3
Following an AS9100 audit, a client of mine was asked to better highlight the acceptability of the calibration results.
Can you please elaborate or explain this more specifically? What comes to my mind is that they are not noting whether the calibration succeeded without any adjustments required; i.e. that the equipment left "AS FOUND". If some adjustment was required during calibration, that should cause your client to initiate an investigation into potential impact on product produced or measured with that equipment. But if they are not checking for that "AS FOUND" status on the cert, not documenting it, and they don't have it in their procedure that it is something to check for each time an item comes back from calibration, then its not clear that they are doing so, or that they are even aware of calibration adjustments as potentially impacting previously produced product.
 

Sidney Vianna

Post Responsibly
Staff member
Admin
#4
In addition to what Tagin mentioned, some calibration certificates have limitations such as, for example, a 0-22” micrometer was only calibrated in the 0-15” range. A Digital Multimeter that only had some functions calibrated.
 
#5
Hallo, thanks everyone for replying.
I have seen that some companies, in the certificates issued by the calibration laboratory, to show that they have verified the certificates themselves write a sentence in the certificate: "The Management, after analyzing the certificate issued by the calibration service provider, certifies that the instrument of calibrated measurement is suitable for use by the company as the mean square error is included within the manufacturing limit deviations of the instrument".
For example I saw it on a certificate of a smooth ring. Do you think I can write that sentence in every kind of certificate of gauges, micrometers, threaded and smooth pads, etc.?
Alternatively, I thought of a sentence like this: "It is declared that, after analyzing the certificate issued by the calibration service provider, the calibrated measuring instrument is suitable for use in relation to the measurements to be made and the expected tolerances".

I have to declare that certificates are suitable for use.

Thanks for your help.
MJ




AS9100D 7.1.5.1 covers this issue:

The organization shall determine and provide the resources needed to ensure valid and reliable results when monitoring or
measuring is used to verify the conformity of products and services to requirements.


The organization shall ensure that the resources provided:
a. are suitable for the specific type of monitoring and measurement activities being undertaken;
b. are maintained to ensure their continuing fitness for their purpose.


The organization shall retain appropriate documented information as evidence of fitness for purpose of the monitoring and
measurement resources.


So the organization is required to determine the "fitness of purpose" of the equipment being used and to document evidence of the fitness for purpose. How this is performed is up to each organization.
If they have done this then they can state that since the equipment being used has been calibrated to manufacturer specifications, then the results are acceptable to their requirements.
Can you please elaborate or explain this more specifically? What comes to my mind is that they are not noting whether the calibration succeeded without any adjustments required; i.e. that the equipment left "AS FOUND". If some adjustment was required during calibration, that should cause your client to initiate an investigation into potential impact on product produced or measured with that equipment. But if they are not checking for that "AS FOUND" status on the cert, not documenting it, and they don't have it in their procedure that it is something to check for each time an item comes back from calibration, then its not clear that they are doing so, or that they are even aware of calibration adjustments as potentially impacting previously produced product.
In addition to what Tagin mentioned, some calibration certificates have limitations such as, for example, a 0-22” micrometer was only calibrated in the 0-15” range. A Digital Multimeter that only had some functions calibrated.
 

dwperron

Trusted Information Resource
#7
Hallo, thanks everyone for replying.
I have seen that some companies, in the certificates issued by the calibration laboratory, to show that they have verified the certificates themselves write a sentence in the certificate: "The Management, after analyzing the certificate issued by the calibration service provider, certifies that the instrument of calibrated measurement is suitable for use by the company as the mean square error is included within the manufacturing limit deviations of the instrument".
For example I saw it on a certificate of a smooth ring. Do you think I can write that sentence in every kind of certificate of gauges, micrometers, threaded and smooth pads, etc.?
Alternatively, I thought of a sentence like this: "It is declared that, after analyzing the certificate issued by the calibration service provider, the calibrated measuring instrument is suitable for use in relation to the measurements to be made and the expected tolerances".

I have to declare that certificates are suitable for use.

Thanks for your help.
MJ
How in the world is a calibration lab able to know if a tool is "suitable" for your use?
AS9100 requires your organization to make, document, and retain that determination.
AS9100 requires that the tools are suitable, not the certificates.
The boilerplate on the calibration certificate is cute, but worthless.
 

Brakeman

Starting to get Involved
#8
I interpreted the 17025 and 9001 requirements to mean that someone has to be the authority to review the required MSA along with the calibrations and to determine if the gauge is acceptable to use for the measurement. That is the statement that we required in our gauge system, not our certs. But I guess the certs could be signed to show evidence of review, which is a requirement, as the calibrator is not a company authorized representative unless specifically contracted to do so.
 
#9
Thanks for replying.
The company performs work on customer specifications and, based on the measurements and tolerances on the drawing, uses the most suitable tools. During last AS9100 audit, the auditor requested the following:
"The organization has established the acceptance criteria for measurement instrumentation, however it is suggested to improve the register to better formalize the acceptance process".
In another similar situation, to close this OFI, a company wrote this: "The Management, after analyzing the certificate issued by the calibration service provider, certifies that the instrument of calibrated measurement is suitable for use by the company as the mean square error is included within the manufacturing limit deviations of the instrument".
I wonder if this type of sentence can be suitable for all instruments or not.

Thanks for helping.
MJ


If you want to do that, it is your prerogative, but I do not agree the AS9100 standard requires that to be done.
How in the world is a calibration lab able to know if a tool is "suitable" for your use?
AS9100 requires your organization to make, document, and retain that determination.
AS9100 requires that the tools are suitable, not the certificates.
The boilerplate on the calibration certificate is cute, but worthless.
I interpreted the 17025 and 9001 requirements to mean that someone has to be the authority to review the required MSA along with the calibrations and to determine if the gauge is acceptable to use for the measurement. That is the statement that we required in our gauge system, not our certs. But I guess the certs could be signed to show evidence of review, which is a requirement, as the calibrator is not a company authorized representative unless specifically contracted to do so.
 

dwperron

Trusted Information Resource
#10
"The Management, after analyzing the certificate issued by the calibration service provider, certifies that the instrument of calibrated measurement is suitable for use by the company as the mean square error is included within the manufacturing limit deviations of the instrument".

That statement is gobbledygook.
You need a statement that says that the calibration results have been reviewed and that they are adequate for your measurement processes.

When someone in a future AS9100D audit asks you to show your instruments are adequate for the job you will need to be able to present evidence that you comply with this:

7.1.5.1
"The organization shall ensure that the resources provided:
a. are suitable for the specific type of monitoring and measurement activities being undertaken;
 
Thread starter Similar threads Forum Replies Date
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
F Calibration certificate content - Average Value and True Value ISO 17025 related Discussions 6
M How to read a calibration certificate for thermo hygrometer General Measurement Device and Calibration Topics 2
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
A Calibration Certificate Results General Measurement Device and Calibration Topics 3
E Calibration Certificate Content Requirements General Measurement Device and Calibration Topics 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate Qualification and Validation (including 21 CFR Part 11) 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate General Measurement Device and Calibration Topics 2
G Is the calibration due date of equipment used to be on calibration certificate ISO 17025 related Discussions 38
G Calibration Certificate, fancy paper or not? General Measurement Device and Calibration Topics 9
Crusader Certificate of calibration from Manufacturer allowed? General Measurement Device and Calibration Topics 14
T Calibration Certificate Readings After Adjustment General Measurement Device and Calibration Topics 5
C Calibration Certificate "As Found" condition - TS16949 Section 7.6.2 General Measurement Device and Calibration Topics 7
M ILAC Regulations - Expression of Units on the Calibration Certificate General Measurement Device and Calibration Topics 6
A Coverage factor in calibration certificate for a multimeter General Measurement Device and Calibration Topics 4
N Client Request for Reference Equipment Calibration Certificate General Measurement Device and Calibration Topics 3
T Microsoft Access Database to create a Simple Calibration Certificate General Measurement Device and Calibration Topics 6
V How to use a Calibration Certificate in Uncertainty Calculation for a Test Measure Measurement Uncertainty (MU) 13
T Uncertainty in Hygrometer Calibration Certificate General Measurement Device and Calibration Topics 4
O How can I use reading from calibration certificate in measurement Calibration Frequency (Interval) 7
K New Gauges calibration certificate shows deviations from the measurement uncertainty IATF 16949 - Automotive Quality Systems Standard 5
Q Can we go by different Calibration Frequency vs. What's on the Certificate ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
R Torque Tester Calibration Traceability and Certificate question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S No Calibration Due Date on a Calibration Certificate unless requested by a customer? Calibration Frequency (Interval) 9
P Rounding Uncertainty to Zero on a Calibration Certificate Measurement Uncertainty (MU) 7
N Calibration Lab unable to issue Accredited Certificate when error is too high? General Measurement Device and Calibration Topics 3
S How to Evaluate a Calibration Report Certificate General Measurement Device and Calibration Topics 8
R LCP 01201P and ICPF2P Calibration Methods Stated in Calibration Certificate General Measurement Device and Calibration Topics 2
M ILAC p14:12/2010 - Accredited Calibration Certificate Uncertainty General Measurement Device and Calibration Topics 2
A Can any organization issue a Calibration Certificate? General Measurement Device and Calibration Topics 5
A No Uncertainty Listed on Calibration Certificate: Is it traceable calibration? General Measurement Device and Calibration Topics 8
M External Calibration Certificate Review ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
N Calibration Certificate Requirements for Pressure Gauges General Measurement Device and Calibration Topics 9
A Fluke Digital MultiMeter Calibration Traceability and Certificate General Measurement Device and Calibration Topics 11
S Reference to original Calibration Certificate - ISO 17025 Section 5.10.9 ISO 17025 related Discussions 4
howste Calibration Certificate Contents: How to report Measurement Uncertainty? Measurement Uncertainty (MU) 18
C Accredited Calibration Certificate Content - ISO 17025 Section 5.10.4.2 Question ISO 17025 related Discussions 7
D Calibration Certificate - 'test report shall not be reproduced except in full' General Measurement Device and Calibration Topics 2
howste Condition of Item Calibrated - ISO 17025 Clause 5.10.2f - Calibration Certificate ISO 17025 related Discussions 13
C Interpreting Outside Laboratory Calibration Certificate Measurement Uncertainty (MU) 3
A Calibration Certificate content for New Equipment General Measurement Device and Calibration Topics 8
R What to do after I receive a calibration certificate General Measurement Device and Calibration Topics 4
E Calibration - Manufacturer's Calibration Certificate Date ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
S Verification of calibration certificate Miscellaneous Environmental Standards and EMS Related Discussions 1
B ISO 17025 logo & ISO 9001 logo on calibration certificate/report ISO 17025 related Discussions 7
P Certificate of Calibration Suppliers for an ISO 9001 company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
S Validity of 7 Year Old Calibration Certificate General Measurement Device and Calibration Topics 19
BradM Certificate for In-House Calibration General Measurement Device and Calibration Topics 25
Wayne Never put an Expiration (Due) Date on a Calibration Certificate Calibration Frequency (Interval) 41
J In-House Calibration Certificate wording - Confidence Level or Level of Confidence? Document Control Systems, Procedures, Forms and Templates 3