Calibration cycle for monitoring & measuring tools used in medical device manufacturing

#1
Hello all,
What would be the appropriate calibration interval for monitoring & measuring tools used in medical device manufacturing? For instance, the instruments which are being used frequently say daily or 3 days in a week have been set 1 year as calibration interval. But what does not make sense to me is even the calibration cycle for the instruments & tools used less frequent say 1 or 3 times a month is also set 1 year as calibration interval. This takes huge toll on the economic condition of the company. So can the calibration interval can be made once a 2 years or 3 years based on the usability or is it mandatory to be calibrated every year to be compliant with ISO 13485?

Looking for your valuable insights on this. TIA.
 
Elsmar Forum Sponsor

Jen Kirley

Quality and Auditing Expert
Staff member
Admin
#2
Hello all,
What would be the appropriate calibration interval for monitoring & measuring tools used in medical device manufacturing? For instance, the instruments which are being used frequently say daily or 3 days in a week have been set 1 year as calibration interval. But what does not make sense to me is even the calibration cycle for the instruments & tools used less frequent say 1 or 3 times a month is also set 1 year as calibration interval. This takes huge toll on the economic condition of the company. So can the calibration interval can be made once a 2 years or 3 years based on the usability or is it mandatory to be calibrated every year to be compliant with ISO 13485?

Looking for your valuable insights on this. TIA.
Hello Jake27, welcome to The Cove!

As is the case with other ISO standards, ISO 13485:2016 does not specify calibration intervals. The standard calls for calibration/verification at specified intervals (it does not say who specifies) or prior to use... please see clause 7.6.

As always, industry good practices are to set calibration intervals based on specification tolerance, the instrument's age/sensitivity, amount of use, environment of use, handling/storage impacts on the instrument's physical condition and expected reliable accuracy, and the cost of getting it wrong.

I hope this helps.
 

Tagin

Trusted Information Resource
#3
and the cost of getting it wrong
@Jake27, when you say 'medical device manufacturing', my thought immediately go to risk:
  • what safety impacts are there to the lives and health of the patients if your M&M are out of calibration, and that condition goes undetected for as long 1-3 years?
  • what reputation impacts are there to your company name if a recall is required for 1-3 years worth of product produced with out of calibration M&M?
  • what financial impacts are there to your company name if a recall is required for 1-3 years worth of product produced with out of calibration M&M?
  • what financial impacts are there to your company name if a lawsuit results from having 1-3 years worth of product produced with out of calibration M&M?
I wonder if 'huge toll on the economic condition of the company' is perhaps not so onerous in light of these risks.
 
#4
Hello all,
What would be the appropriate calibration interval for monitoring & measuring tools used in medical device manufacturing? For instance, the instruments which are being used frequently say daily or 3 days in a week have been set 1 year as calibration interval. But what does not make sense to me is even the calibration cycle for the instruments & tools used less frequent say 1 or 3 times a month is also set 1 year as calibration interval. This takes huge toll on the economic condition of the company. So can the calibration interval can be made once a 2 years or 3 years based on the usability or is it mandatory to be calibrated every year to be compliant with ISO 13485?

Looking for your valuable insights on this. TIA.
Appreciate your concerns and in this case you can take the help of ISO/IEC 17025 as well as ISO 10012 standards
1) it is the knowledgeable person that decides the calibration interval not any calibration service provider - once again i reaffirm that "it is the knowledgeable person that decides the calibration interval"
2) first of all, you need to check up whether there is bias or drift in the measurement results
3) to make it easy - you shall be having equipment that are likely to have drift due to usage and equipment that don't encounter drift due to usage.
4) example for the above is patient monitor Vs dialysis machine
5) you can refer to Ilac publication : ILAC g-24
 
Last edited by a moderator:
#5
Hello all,
Thank you for shedding the lights. It helps All the above suggestions were considered but my concern was to be in compliant with the ISO 13485 as I did not find enough information about the calibration intervals in the standard. Once again, Thank you all :)
 

yodon

Staff member
Super Moderator
#6
@Tagin nailed it. It should definitely be risk-based. Auditors may well ask you what your justification for the frequency is. Sounds like the current approach was just arbitrarily set at 1 year (which, in my experience, most auditors don't question)?
 
Thread starter Similar threads Forum Replies Date
M Calibration Cycle for Standard Iron & Steel Plate Gage General Measurement Device and Calibration Topics 17
O Scale Calibration Frequency - Low Volume - Current Cycle is 6 Months Calibration Frequency (Interval) 6
G When to start calibration cycle - Electronic test equipment rental company General Measurement Device and Calibration Topics 3
R Calibration Cycle: Calibrate When Issued OR Calibrate by Due Date? General Measurement Device and Calibration Topics 7
Marc Brute Force Validation - Difficult to show 10-year calibration cycle is sufficient General Measurement Device and Calibration Topics 0
N IATF Calibration Lean in Manufacturing and Service Industries 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 5
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3

Similar threads

Top Bottom