jane_ackerman - 2010

#1
Our current procedure (for any calibration - in house or outsourced) requires checking all test points before even cleaning the gage.
I question whether this is overkill in our procedure.
Is this really what the standard is implying?
I personally would think that it is implying "before readings" are to represent readings collected before any adjustments are made (not including removal of build up during shift production)??

When we have only a 5 micron tolerance on plug gages, this could make several plugs appear out of tolerance (after only one use)... but when they are wiped off they are fine.
This shift build up, does not affect product quality because it simply slides up the gage when it is inserted into the next part and is wiped off each shift.

Wondering what others consider a "before" reading to be.
If there are any TS auditors out there, please give insight.

Thanks!
 
Elsmar Forum Sponsor

Jim Wynne

Staff member
Admin
#2
jane_ackerman said:
I personally would think that it is implying "before readings" are to represent readings collected before any adjustments are made (not including removal of build up during shift production)??
No, because calibration doesn't necessarily include adjustment. What you want to know is the condition of the gage as received. Cleaning should only be done after the as-received measurements are done, so as to mirror the use of the gage in the condition it was last used in production.

jane_ackerman said:
When we have only a 5 micron tolerance on plug gages, this could make several plugs appear out of tolerance (after only one use)... but when they are wiped off they are fine.
This shift build up, does not affect product quality because it simply slides up the gage when it is inserted into the next part and is wiped off each shift.
If cleanliness of the gage doesn't affect measurement results, then cleaning before measurement is reasonable, but I would expect that you would have data to back up the assertion.
[/quote]
 
J

jane_ackerman - 2010

#3
As to mirror the use of the gage in the condition it was last used in production.
This is how we have always done it - but like I said, if production wipes them off, wouldn't I be mirroring production if I did the same thing?

If cleanliness of the gage doesn't affect measurement results, then cleaning before measurement is reasonable,
The cleanliness does affect the measurement results.

but I would expect that you would have data to back up the assertion.
So.... if I were to have X number of documented instances, where the coolant / shift build up on the gage was the culprit of the OOT condition... and that the investigation of product quality showed zero non-conforming parts as a result, this might suffice (to an auditor) as to allow me to wipe the gage down first?
 

Jim Wynne

Staff member
Admin
#4
jane_ackerman said:
As to mirror the use of the gage in the condition it was last used in production.
This is how we have always done it - but like I said, if production wipes them off, wouldn't I be mirroring production if I did the same thing?
Yes.

jane_ackerman said:
If cleanliness of the gage doesn't affect measurement results, then cleaning before measurement is reasonable,
jane_ackerman said:
The cleanliness does affect the measurement results.
I was referring to production measurement results, not calibration measurement results. Sorry for the confusion.

jane_ackerman said:
but I would expect that you would have data to back up the assertion.
jane_ackerman said:
So.... if I were to have X number of documented instances, where the coolant / shift build up on the gage was the culprit of the OOT condition... and that the investigation of product quality showed zero non-conforming parts as a result, this might suffice (to an auditor) as to allow me to wipe the gage down first?
Yes. But--if you're receiving dirty gages from production, and you've ascertained that dirty gages = potentially erroneous results (even if you don't know of a specific instance in production) then maybe production people need some training in gage handling and maintenance, which might obviate the need for you to be concerned about it for calibration purposes. In the end, I think it's common to wipe down a pin gage prior to calibration, so I don't see this as being an issue with regard to auditors. Just out of curiousity though, if your tolerance is 5 microns, what do you use to calibrate them?
 
J

jane_ackerman - 2010

#5
Jim Wynne said:
Just out of curiousity though, if your tolerance is 5 microns, what do you use to calibrate them?
I use a bench micrometer with an electronic amplifier that reads to .00001 mm (I think this is called 10 nanometers?) in conjunction with Grade 1 JO Blocks.
 
Last edited by a moderator:

Jim Wynne

Staff member
Admin
#6
jane_ackerman said:
I use a bench micrometer with an electronic amplifier that reads to .00001 mm (I think this is called 10 nanometers?) in conjunction with Grade 1 JO Blocks.
Geez--that seems like many orders of magnitude of overkill--what are you measuring with those pins, bacteria?
 
J

jane_ackerman - 2010

#7
Jim Wynne said:
Geez--that seems like many orders of magnitude of overkill--what are you measuring with those pins, bacteria?
Ha Ha! Actually, when I started here, all they were using was the benchmic... but it only reads to .002 mm alone. I found the amplifier in a cabinet (never used) and it sufficed to meet our needs.
 

Jim Wynne

Staff member
Admin
#8
jane_ackerman said:
Ha Ha! Actually, when I started here, all they were using was the benchmic... but it only reads to .002 mm alone. I found the amplifier in a cabinet (never used) and it sufficed to meet our needs.
:topic: slightly, I guess, but I once took a job as A quality manager and found that the incoming inspection department had an 80-inch vernier caliper. Naturally, I was curious as to what it was used for. Turned out that it was used to measure sheet metal panels that were about five feet long, toleranced at +/- .010." I queried the responsible engineer who said the close tolerance wasn't really necessary; he just defaulted to the block tolerance for 3-decimal place callouts. The tolerance got changed to +/- 1/16" (which was still liberal) and so far as I know, the giant caliper was never used again.:frust:
 
J

jane_ackerman - 2010

#9
Jim Wynne said:
:topic: slightly, I guess, but I once took a job as A quality manager and found that the incoming inspection department had an 80-inch vernier caliper. Naturally, I was curious as to what it was used for. Turned out that it was used to measure sheet metal panels that were about five feet long, toleranced at +/- .010." I queried the responsible engineer who said the close tolerance wasn't really necessary; he just defaulted to the block tolerance for 3-decimal place callouts. The tolerance got changed to +/- 1/16" (which was still liberal) and so far as I know, the giant caliper was never used again.:frust:
At least your scenario wasn't reversed.
When I came on board, I had operators measuring a product dimension that had a 3 decimal metric tolerance with a 2 decimal caliper.
:bonk:
 
Thread starter Similar threads Forum Replies Date
G Before and After Calibration Data ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
W Complying with ISO 9001 Calibration Requirements - Report Before and After Data General Measurement Device and Calibration Topics 12
A Calibration tolerance for data logger General Measurement Device and Calibration Topics 3
I ADR (Automatic Data Recorder) Calibration Requirements Calibration and Metrology Software and Hardware 2
kedarg6500 Protection of Data in Calibration Software & ISO 9001 Calibration and Metrology Software and Hardware 4
W Calibration Register - Traceability - Data Transfer to a New System General Measurement Device and Calibration Topics 7
B How do I calculate Correction to Caliper Calibration Data General Measurement Device and Calibration Topics 4
M "As Found" (aka "As Received") Calibration Data - ISO 13485 Requirements ISO 13485:2016 - Medical Device Quality Management Systems 8
T Calibration of Anemometer Data Transmitters General Measurement Device and Calibration Topics 3
B No editable (alterable) raw data from outside calibration laboratory? General Measurement Device and Calibration Topics 20
P Plug Gage Calibration Report does not include actual data - ISO 9001:2000 section 7.6 General Measurement Device and Calibration Topics 3
R Calibration Intervals (Frequency) derived from Variables Data Calibration Frequency (Interval) 30
B Calibration Data Packages - CMM - Any way around "As found data?" General Measurement Device and Calibration Topics 17
H METCAL Software (from Fluke) used only to get the Electrical Calibration Data? General Measurement Device and Calibration Topics 8
J Seeking: Gage Calibration Data Collection sheet example General Measurement Device and Calibration Topics 3
B Calibration Certificates - Does 17025 require that all 'As Found' data be recorded? ISO 17025 related Discussions 7
C SPC for calibration check data? Daily gram scales check - What is acceptable? Statistical Analysis Tools, Techniques and SPC 5
R Calibration Procedures - Government Industry Data Exchange Program (GIDEP) General Measurement Device and Calibration Topics 18
M Auditor interpretation of PO data - PO submitted for outside calibration General Auditing Discussions 2
Marc Z540, ISO 17025 and Calibration Data - Requesting Calibration data ISO 17025 related Discussions 2
Marc Customer Audit of Guide 25 Calibration - Full pre and post adjustment data? General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3

Similar threads

Top Bottom