Calibration Discrimination - Acceptance criteria is not having the accuracy required?

A

Andrews

A temperature indicator is used to check temperature of a bath that has to be maintained at 35.2 deg celsius to 39.4 deg celsius.This temp. indicator is being calibrated and the acceptance criteria is fixed as +/- 1 deg celsius.Isn't this a non-conformity as the acceptance criteria is not having the accuracy required by the test?Shouldn't the acceptance criteria be +/-0.1 deg celsius?

Comments pls.

Andrews
 

The Taz!

Quite Involved in Discussions
Depends . . . you must have the discrimination to read to 0.1 or 0.05 to see the variation in the process. IMHO, I would like to see the instrument calibrated within the working range and at the resolution required. You are paying for the calibration thus you are the customer and make the requirements.

You might want to check around and find out what is common practice for that type of process before getting into a dispute.
 
T

tomvehoski

There are several questions to ask:

How critical is 35.2 to 39.4? What happens if it is 39.8? Or 42?

If nothing happens, why is the spec. so tight and why are you bothering to calibrate the gage. If it may create bad material, slow down the process, etc - it is probably best to have it calibrated. If the building explodes, it should definately be calibrated. You need to make the call based on the risk and cost/benefit analysis.

What type of device is being used? RTD, Thermocouple, thermometer, etc?

For a range that narrow you may need to look at calibrating both the sensor (RTD/TC) and the readout device. Its been awhile, but I recall you can get a TC certified - for big $$$. They are not very accurate devices. Even if you have resolution to .1 degree, you need to make sure the device is still accurate enough.

Hope this helps,

Tom
 
Thread starter Similar threads Forum Replies Date
C IR thermometer for refrigerated Centrifuge calibration General Measurement Device and Calibration Topics 2
G ISO 17025.2017 Clause 7.8.4.3 Certificate of Calibration Label ISO 17025 related Discussions 5
M CMM Calibration - What, exactly, is the output? General Measurement Device and Calibration Topics 11
J Using MSA instead of calibration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
P Hierarchy of Calibration in Metrology General Measurement Device and Calibration Topics 1
J Calibration frequency for glass scale Calibration Frequency (Interval) 2
S Calibration SOP and external vendors General Measurement Device and Calibration Topics 2
P IATF16949 External Lab Calibration -CIPM MRA IATF 16949 - Automotive Quality Systems Standard 1
CaliperJim Acclimation time prior to calibration General Measurement Device and Calibration Topics 8
S How to report a measurement and relate it to calibration certificate General Measurement Device and Calibration Topics 1
R Do I need to get calibration certificate from ISO 17025 for IATF Audit IATF 16949 - Automotive Quality Systems Standard 8
S Netsuite Preventive Maintenance and Calibration Manufacturing and Related Processes 0
D Roughness specimen calibration Calibration and Metrology Software and Hardware 5
dhuot1990 General Calibration SOP Manufacturing and Related Processes 3
B Gage calibration Gonogo (calcul of cgk) General Measurement Device and Calibration Topics 0
Vader22 As-Found FAILED calibration ISO 14001:2015 Specific Discussions 16
M Gravimetric blender for resin in automotive industry - periodical calibration via external lab with certification Reliability Analysis - Predictions, Testing and Standards 0
N Calibration Method for Scales and Flowmeters Reliability Analysis - Predictions, Testing and Standards 7
T Calibration Tracking ISO 13485:2016 - Medical Device Quality Management Systems 4
Q Considering addition of mass calibration to ISO17025 scope ISO 17025 related Discussions 6
Z Resolution in calibration General Measurement Device and Calibration Topics 1
J Run Down Adapters Calibration Processes General Measurement Device and Calibration Topics 4
S Looking for accredited calibration lab - Onsite up to 650A and 1 kV. General Measurement Device and Calibration Topics 3
CaliperJim Screw Thread Micrometer Anvil Calibration General Measurement Device and Calibration Topics 7
Strange Juan HP 85054B Standard Mechanical Calibration Kit Calibration Procedure General Measurement Device and Calibration Topics 1
H Air Pressure Gage Calibration General Measurement Device and Calibration Topics 6
T Calibration on AVL AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
Sam.F Do heat treatment ovens need calibration? General Measurement Device and Calibration Topics 25
J Selecting reference standards for calibration General Measurement Device and Calibration Topics 3
C IATF 16949:2016 Major NC pertaining to 7.1.5.2.1 Calibration/verification records IATF 16949 - Automotive Quality Systems Standard 15
E Calibration Records during AS9100 Audit AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
John Predmore Calibration Determination decision Calibration Frequency (Interval) 16
J ISO 17025 Calibration Other Medical Device Related Standards 2
A Electronic Weighing scale calibration Calibration Frequency (Interval) 4
M Bourdon gauge calibration procedure failure ISO 17025 related Discussions 3
Ed Panek Calibration question Manufacturing and Related Processes 4
S Calibration/Worn gages Manufacturing and Related Processes 9
A Settings rings calibration General Measurement Device and Calibration Topics 4
R 7.1.5.2 -- Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
C CALIBRATION CERTIFICATE REQUIREMENTS General Measurement Device and Calibration Topics 11
S Calibration Label issues General Measurement Device and Calibration Topics 21
H Calibration / Capability in a PCP FMEA and Control Plans 1
R First Time Managing Calibration and Measurement System. Need Help. General Measurement Device and Calibration Topics 19
M CE marked medical device - do they need UKAS certified calibration provider ISO 17025 related Discussions 4
JDJohnson Micrometer & Caliper Calibration (ISO-9001) General Measurement Device and Calibration Topics 26
B Precision Current Loop for Current Probe Calibration Reliability Analysis - Predictions, Testing and Standards 2
Strange Juan Freezers fridges Incubators Calibration Labels General Measurement Device and Calibration Topics 12
D Tolerance and MPEs for Re-calibration Manufacturing and Related Processes 0
D ISO 9001:2015 Calibration Procedures Manufacturing and Related Processes 16

Similar threads

Top Bottom