Calibration Environment - Micrometers - NCSL RP 7 - Laboratory Design

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#1
From: "James R. Lloyd"
Subject: RE: Calibration Environment/Lemieux/Lloyd

Couple of questions first Kevin .... What do you mean by micrometer standards? I take it you mean the end user micrometers, etc... usually when I hear the term "standard" Im thinking of Calibration Laboratory level equipment such as gage blocks, parallel flats, and optical parallels used to calibrate outside micrometers .... As far as environmental effects on "floor" usage micrometers, it depends upon your resolution and accuracy requirements that you need on the floor. If I remember correctly expansion coefficients for steel are something like 1 micro inch (.000001) per degree F per inch, this is typically miniscule in most machining type environments. Considering that your typical micrometer is usually calibrated to an accuracy of +/- .001 inches, if your measurements are only in this range it would take some SERIOUS and I mean SERIOUS changes in temperature before the measurement accuracy of your floor micrometers would be affected.

The main concern for Micrometers on a floor is mishandling, and abuse that can affect the measurement accuracy rather than the environment.

For the calibration of Micrometers a bit more environmental control is required, not so much for the micrometer, but for the standards used to calibrate it. Referencing an USAF Metrology Procedure for micrometers .... There are three primary attributes that are calibrated ... measurement accuracy/linearity, flatness of the anvil and the spindle, and parallelism between the anvil and spindle. An acceptable environment to calibrate you typical +/- .001 0-1 inch micrometer would be even 75-79 degrees as the gage blocks would certainly not expand to a degree that would be even measurable ... but the larger the micrometer measurement the more likely that expansion WILL affect its calibration. Usually this threshold is around 4-5 inch micrometers, where you would need an environment controlled to about 71 +/- 4 degrees, and even this becomes excessive with micrometers that measure 10-11 inches where the optimum dimensional calibration environment of 68+/- 1 is required.

I hope I didn't inundate you with information, but metrology, and calibration programs are my specialty.

Best Regards
James Lloyd
QA Manager
Wesport Steel and Supply Inc.

One Comment, I hope all the other mics that are not calibrated are still
identified in your M&TE inventory and identified/labeled as "Not
Calibrated", regardless of use they are still M&TE and must be identified
and controlled.

-----Original Message-----

> From: Kevin Lemieux
Require information regarding consensus on what environmental parameters should calibration be carried out in? What do others do with their micrometer standards so they are protected as well as available for shop floor usage? We currently have calibration tractability on two sets of standards which are available for shop personnel usage. This eliminates the requirement to calibrate all personal standards. The question has came up on the environment in relation to temperature and location of the generally used standards. Help?
> Kevin - QA Manager
 
Elsmar Forum Sponsor
S

SCOTT SNYDER

#2
KEVIN,
YOU MUST HAVE A SEPERATE AND SEGRAGATED SET OF STANDARDS FOR CALIBRATION. YOU CAN NOT USE A SINGLE SET FOR CALIBRATION AND FOR SHOP USE. YOUR CALIBRATION STANDARDS MUST BE PROTECTED AN ISOLATED FORM THE REST OF YOUR STANDARDS. THEY ALSO SHOULD ONLY BE ACCESIBLE BY THE PERSON IN CHARGE OF YOUR CALIBRATION SYSTEM. I DON'T KNOW IF THAT ANSWERS YOUR QUESTION.

------------------
 
D
#3
After two hours of sitting in the QC department to be the same temperature as the calibration equipment, will the micrometer change when it is back out on the plant floor?
Supposing you adjust the micromter in QC, take it out on the floor and it gets warm again. Does this change the accuracy of the micrometer?
 
D
#4
Question # 2:
Why can't you use the same gauge blocks for calibration of micrometers that are used on the shop floor as long as they are certified and checked once a year.
 

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#5
Ummm, well,

After two hours of sitting in the QC department to be the same temperature as the calibration equipment, will the micrometer change when it is back out on the plant floor?
Supposing you adjust the micromter in QC, take it out on the floor and it gets warm again. Does this change the accuracy of the micrometer?
The issue here is sensitivity to temperature and differences between the standard and that being measured. Although some examples are self evident (e.g.: measures of length such as with a micrometers) such as coefficients of expansion, others (e.g.: electrical resistance properties) are not so easy to see and understand.

De Key:
One calibrates a micrometer in a 'lab' at a certain temperature because the standard is certified at that temperature (remember, every substance changes as temperature changes but different substances each temperature change differently with any given change). Bi-metal 'thermometers' are a good example of different coefficients of expansion.

The question always being asked is 'Well, we use it on the floor where it's hot so why not calibrate it there?' or something similar. Typically the items we measure and where we measure them (temperature, humidity, etc.) and with what with respect to required tolerances is not a big issue - BUT - it can be. This is in part the reason for QS9000's drive for Measurement Systems Analysis. It's not just calibration, it's the understanding of what you're doing. What I mean here is that all relative factors are considered from the design stage where tolerances and specs are determined, to use of the FMEA and thence the control plan from whence critical characteristics (as well as required Measurement and Test equipment) are determined, to the production environment. Like we have talked about in a couple of other forums, this is a holistic concept.

So - for question 1, the answer is: Typically it is not a problem (use of instrument on the shop floor after calibration in the lab) because differences in coefficients of expansion (or other temperature related effect{s}) with respect to the temperature difference (delta), the required 'sensitivity' (i.e.: precision, I think, is the word), and tolerances (to a small degree) are negligible.

Question # 2:
Why can't you use the same gauge blocks for calibration of micrometers that are used on the shop floor as long as they are certified and checked once a year.
You can if you want to. It depends upon how your system is designed, if you will. There is no requirement for a single standard to be kept in a lab for calibration. As long as you can show and explain your calibration cycle and show evidence that the yearly verification of calibration is sufficient and (if you will) effective, you may do this. Don't confuse reality with 'teachings'. We are taucht there is a master in the lab and then shopw floor 'standards' are calibrated against this 'golden' standard. The issue is frequency of use (wear) of mechanical/physical M&TE, possible damage on the floor and such factors. However, in reality not that many companies are all that big and complex to neccessitate a 'golden' standard.

[This message has been edited by Marc Smith (edited 12-31-98).]
 

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#6
Date: Wed, 13 Oct 1999 07:17:09 -0500
From: "SMITH, RON D. (JSC-EM)" jsc.nasa.gov
To: 'Greg Gogates'
Subject: RE: Environmental Conditions

There are a couple of documents we use as standards or guides for Calibration Laboratory environmental conditions.

NCSL Recommended Practice (RP) 7, "Laboratory Design" TO 00-20-14, "Air Force Metrology and Calibration Program"

Ron Smith
Johnson Space Center
 
Thread starter Similar threads Forum Replies Date
D Calibration in Warehouse Environment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
C Balance Calibration SOP for a Pharmaceutical Environment General Measurement Device and Calibration Topics 4
V Calibration Laboratory Environment and Instrument Tolerances General Measurement Device and Calibration Topics 15
D Illegible Calibration Stickers - Gages used in oily, dirty environment General Measurement Device and Calibration Topics 5
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 0
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 5
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 3
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3

Similar threads

Top Bottom