Calibration for Solid Length Gauges

V

Vadamalai

#1
? For measuring length below head in bolt blanks in as forged soft condition , we use thousnads of solid length gauges.These gauges are duly heat-treated to 45/48 HRC.The tolerance for the length in bolt blanks is 0.8 mm. These gauges are easy for verifying length for operators and there is no adjustment involved. Gauges are of ground finish and flatness is ensured.Conisdering the tolerance of the blank length,robustness of the solid gauges,extremely negligible chances of wear is periodic calibration is necessary for these?The Cpk in SPC of the bolt length is >2.
************
Edited html error.

[This message has been edited by Marc Smith (edited 04 December 2000).]
 
Elsmar Forum Sponsor
A

Al Dyer

#2
Is the gage used to qualify the part to customer specifications? If so, it needs to be calibrated at an appropriate frequency.

This can be determined based on the various statistical tools available. I.E. ongoing Cpk, PPM, Stability etc...

Looking forward to additional input!

ASD...
 
A

Atul Khandekar

#3
From what you describe, it appears that these are Attribute gages (GO-NO GO). As such R&R should be done. If chances of wear as negligible, you may choose a suitable frequency for studies, like once a year or so.
 
Thread starter Similar threads Forum Replies Date
D ISO/Metric Solid Thread Ring Calibration General Measurement Device and Calibration Topics 1
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 5
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 3
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6

Similar threads

Top Bottom