Calibration in the Hard Tool Industry

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#1
This is an interesting take on calibration, I must say! Comments any one?

-------snippo------

Subject: Re: Q: Cal of Design Instruments/Woodley/Kleinpaste/Perdue
Date: Wed, 30 Sep 1998 11:27:42 -0600
From: ISO Standards Discussion

From: Jon Perdue
Subject: RE: Q: Cal of Design Instruments/Woodley/Kleinpaste/Perdue

I respectfully disagree. The company I am the Management Rep. for builds "Hard Tools" for metal stamping. Our customer gives us a blueprint for the part they want the tool to produce. They tell us they want it to be able to produce "X" amount of parts per hour, and they expect the product life to be "X" amount of parts. We design the tool based on that input. Hard Tool designs can take into account many variables, and with minor changes from one lot to another in material hardness, tensile strength, and the "chaos" factor, it is an inexact science (tool designers may disagree, but for the sake of argument, let's continue) . The First Article report generated from measurements taken from an actual part that has been produced by the hard tool. The First Article measurements are made using calibrated equipment. The acceptance of our hard tool is based on the First Article report.

So, the measuring and test equipment of the hard tool designers and the people making the hard tool is not calibrated. The tools (4.11.1) "used by the supplier to demonstrate the conformance of product to the specified requirements" are calibrated.

Obviously, one can question if it makes good business sense to not calibrate the toolmakers measuring equipment, but since they don't use them for final approval of the hard tool...

This application of this interpretation has passed ISO9001 scrutiny (so far).

OK, take aim and let me have it.
Jon Perdue
_ _ _ _ _ _ _ _ _ _ _ _ _ _ _ _ _

>From: "Kleinpaste, Marcia E"

Yes! We are principally a design firm with no manufacturing at this point. Underwriter's Laboratories has required calibration records, i.e. certificates, recall data, etc. on all instruments, including a tape measure. Our technicians and design engineers must keep accurate records of calibration.

> M Kleinpaste
> battelle.org
>- - - - - - - - - - - -
>> From: Craig Woodley - menasha.com
>
>> Do instruments (gen. calipers, rulers, thickness gauges) that are used by
>> Design Engineers in the development of a design need to be a part of the
>> calibration program?
>>
>> 4.11.1 says "used by the supplier to demonstrate the conformance of
>> product to the specified requirements"
>>
>> The design function is at a different site than the manufacturing
>> function.
>> The equipment would be used to measure the customer supplied part to
>> determine the design for packaging around the customer's product. The
>> designs are then done in a CAD system and a die built from that data.
>>
>> Our output is the actual packaging, which is verified at manufacture. The
>> customer signs off on the design and pre-production sample.
>>
>> Thanks in advance for your response.
>>
>> C.S. Woodley
>> menasha.corp
 
Elsmar Forum Sponsor

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#2
Randall is pretty good. Here's his response:

-------snippo--------

Subject: Re: Q: Calibration of Design Instruments/Woodley/Randall
Date: Wed, 30 Sep 1998 11:32:00 -0600
From: ISO Standards Discussion

From: RCR9000
Subject: Re: Q: Calibration of Design Instruments/Woodley/Randall

> From: Craig Woodley
>
> Do instruments (gen. calipers, rulers, thickness gauges) that are used by
> Design Engineers in the development of a design need to be a part of the
> calibration program?

During the ISO 9001 audits I perform, I expect to see all instrumentation used for Design verification and Design validation under calibration control. Remember that ISO 9001 addresses design, NOT research and development.

> 4.11.1 says "used by the supplier to demonstrate the conformance of
> product to the specified requirements"
>
> The design function is at a different site than the manufacturing function.

If this function is included in your scope of registration, then all of ISO 9001 requirements apply to that site regardless of its location.

> The equipment would be used to measure the customer supplied part to
> determine the design for packaging around the customer's product. The
> designs are then done in a CAD system and a die built from that data.

I would expect to see instruments used to "verify" customer supplied product (parts) under calibration control. On a personal note, it seems to me that since you are collecting design input from your own measurements that you would want to be certain that those measurements were accurate regardless of any ISO 9001 requirements.

> Our output is the actual packaging, which is verified at manufacture. The
> customer signs off on the design and pre-production sample.

If the packaging (the final output) is not verified (validated) until manufacture, doesn't that defeat (or at least impede) the purpose of design verification and design validation? (Assuming that that the product (packaging) you develop is mass produced) Calibration, in this instance, appears to be a means of preventive action. If design input data is not accurate, causing production of incorrect packaging, couldn't that potentially create a tremendous unnecessary expense?

I understand that one argument in this area is that there may have never been any defective product manufactured - even though "uncalibrated" instrumentation has always been used. This is because many instruments calibrated by the manufacturer retain their accuracy for several years (even though a yearly calibration cycle may be recommended by the manufacturer). This argument is a great basis for extending calibration cycles, not for debating whether a calibration system is needed or not.

I hope that this helps,

Richard C. Randall
 
Thread starter Similar threads Forum Replies Date
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 5
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 3
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1

Similar threads

Top Bottom