Calibration Interval - AS9100 - Control of Monitoring and Measuring Equipment

#1
Having an 'animated' discussion with a peer whose opinion I value highly; however on this topic we disagree, and seeking some thoughts from the assembled multitude of QA pros.

As9100 (whatever) tells us that control of monitoring and measuring equipment is to be done at specified intervals, and that we can set them based on whatever criteria we choose/deem appropriate.

The gear in question happens to be some floor fitting interface fixtures, used to 1) locate the monument-to-floor fittings prior to bonding. The locations are derived from the model and the fixtures were machined and then digitally scanned to a magnitude greater than the tolerance of the fittings' locations. And yes the scanning equipment itself was all calibrated.

Here is the point of contention: since these plates are made of 1/2" thick aluminum any damage that affects the fitting interface will be quite visible and because of that, I called for a 'One Time Only (OTO)' cal interval. The quality plan for this project and project unique cal process defines this and also allows it as long as no damage to the fixture is present.

My friend does not accept a OTO; instead he maintains that I should have some interval, even if it is a 5 year interval.

When I ask him why, he just repeats himself. I don't want to put us in a position of missing a 5 year window, and at the same time am confident any damage that would impact the fitting locations would be blatantly obvious.

Your thoughts and comments please.

Thanks.
 
Elsmar Forum Sponsor

outdoorsNW

Involved In Discussions
#2
Please provide some more details about why you think damage will be quite visible.

It seems like you could have damage that is just enough to make the fixture not work correctly but might not be obvious. Being a softer metal, aluminum may wear over time and the fixture may slowly shift away from the desired alignment.
 

Sidney Vianna

Post Responsibly
Staff member
Admin
#3
The gear in question happens to be some floor fitting interface fixtures, used to 1) locate the monument-to-floor fittings prior to bonding. The locations are derived from the model and the fixtures were machined and then digitally scanned to a magnitude greater than the tolerance of the fittings' locations. And yes the scanning equipment itself was all calibrated.

Your thoughts and comments please.

Thanks.
From what you describe, this sounds much more like an assembly fixture, rather than an inspection, measuring, test device. If it is indeed a production tool, then the requirement you are concerned with does not apply. It would have to be validated and maintained, but not "calibrated".
 

howste

Thaumaturge
Super Moderator
#4
I agree with what Sidney said. Calibration is required for measuring equipment. Is there something else that is used to measure the product that will detect a problem if the fixture is damaged or wears? If so, this is what should be periodically calibrated.

As far as maintenance, aluminum fixtures are very prone to wear. If it's being used frequently then I would expect that it would wear over time and eventually won't be able to hold a tolerance. I don't know what kind of tolerances you have, but this should be considered. Periodic verification of the tooling or ongoing measurement of resulting product would detect this.
 
#5
Thanks for the inputs; I see that I wasn't clear ( working from a position of 'insider trading'):

The fixture has bolt holes that the floor fittings are fixed to. Then the entire fixture/fitting assembly is mated with our monument. the monument is constructed of honeycomb core panels and the area where the fittings are inserted is cleared away and after the fittings have been inserted, we then bond them in place. The check fixture remains until the adhesive cures. Since the through holes for the bolts is not an interference fit, any wear to these holes would be insignificant. My tolerance is +/- 0.030" on the monument, and by fixture tolerance was +/- 0.010" for the bolt hole centers.
 

Eredhel

Quality Manager
#6
I'm leaning toward what Sidney said as well. It almost sounds like it's a production check fixture that doesn't need to be handled by quality for acceptance. Is this used for acceptance? If not it sounds like production upkeep.
 
#7
We use it as to verify location, so yes it is an acceptance tool and thus requires calibration. It is called out in not only the manufacturing drawings, but also in the customer's source inspection work order. The floor fittings have no adjustment and MUST fit, so this is our only way to ensure there is no installation issue.
 

howste

Thaumaturge
Super Moderator
#8
We use it as to verify location, so yes it is an acceptance tool and thus requires calibration. It is called out in not only the manufacturing drawings, but also in the customer's source inspection work order. The floor fittings have no adjustment and MUST fit, so this is our only way to ensure there is no installation issue.
I agree with your coworker that if calibration is required, a one-time calibration is insufficient. AS9100 says:
AS9100 Rev D clause 7.1.5.1 said:
The organization shall ensure that the resources provided:

a. are suitable for the specific type of monitoring and measurement activities being undertaken;

b. are maintained to ensure their continuing fitness for their purpose.

The organization shall retain appropriate documented information as evidence of fitness for purpose of the monitoring and measurement resources.
Bullet point "a" would be covered by initial calibration. Bullet point "b" would be covered by periodic verification/calibration. The last sentence requires records as evidence that it continues to be fit for use. Also there is the part of 7.1.5.2 that you mentioned including "specified intervals" which would not be met with a one-time calibration.
 

dgriffith

Quite Involved in Discussions
#9
There are many programs that have calibrate (or verify) before use categories. There is no date-driven interval. A mechanical fixture with dimensional requirements seems like a good candidate.
 
Thread starter Similar threads Forum Replies Date
J Documenting a Calibration Interval Extension Calibration Frequency (Interval) 7
D Torque Wrench Calibration Interval Extension with Non-Use? General Measurement Device and Calibration Topics 1
Marc Calibration Frequency (Interval) Discussions Sub-Forum General Measurement Device and Calibration Topics 0
K Calibration Interval based on Usage Calibration Frequency (Interval) 14
J Digital Scale Re-Calibration Interval Frequency General Measurement Device and Calibration Topics 10
M A Guide or Standard for Calibration Interval (Frequency) Calibration Frequency (Interval) 2
D "As Found" Condition and Calibration Interval Changes Calibration Frequency (Interval) 5
B Calibration Interval of Munsell Color Charts General Measurement Device and Calibration Topics 3
C Furnace Classes in relation to AMS 2750 to determine Calibration Interval Calibration Frequency (Interval) 2
A How to Determine Machine Tool Calibration Interval? General Measurement Device and Calibration Topics 4
Z CMM (Coordinate Measuring Machine) Calibration Interval (Frequency) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
MarilynJ6354 Torque Wrench Calibration Frequency (Calibration Interval) Calibration Frequency (Interval) 38
D Nonadjustable M&TE - 10 year calibration interval? Pin Gages Calibration Frequency (Interval) 11
somashekar Calibration interval (Clause 7.6 ISO 13485:2003) - Biomedical test instruments ISO 13485:2016 - Medical Device Quality Management Systems 4
D Thermometer Calibration and Calibration Interval ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J What's your interpretation of Clause 7.6 - Calibration 'Interval'? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 16
J Recommendations on calibration interval (frequency) ISO/IEC 17025:2005 para 5.10.4.4 ISO 17025 related Discussions 5
A Calibration Interval (Frequency) Adjustment using Hilliard type 2 algorithm Calibration Frequency (Interval) 11
L Projecting Calibration interval for Resistors Calibration Frequency (Interval) 2
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 4
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
A Calibration of Power Supplies AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 2
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Calibration Equipment Selection General Measurement Device and Calibration Topics 2
D Sample Calibration SOP for ISO 13485 General Measurement Device and Calibration Topics 4
L Calibration of Plug Gauges - Tolerancing General Measurement Device and Calibration Topics 1
L In house calibration - Our CMM's are outside certified can I use them to certify our standards General Measurement Device and Calibration Topics 4
R Standard Filter for Calibration Colorimeter ISO 17025 related Discussions 1
F Calibration certificate content - Average Value and True Value ISO 17025 related Discussions 6
B Oscilloscope - AC gain calibration General Measurement Device and Calibration Topics 6
G Is it common to do a repeatability check during calibration of an instrument? General Measurement Device and Calibration Topics 5
A Adjustment of Calibration Intervals (Shortening or Lengthen) Calibration Frequency (Interval) 6
A Calibration required by OEM General Measurement Device and Calibration Topics 2
S ISO17020 - Equipment calibration requirement General Measurement Device and Calibration Topics 3
H Can a calibration lab still use ISO 6789:2003 as guide for torque calibrations Other ISO and International Standards and European Regulations 1
H Calibration of InfraRan Vapor Analyzer - where to get calibration for SF6 General Measurement Device and Calibration Topics 2
M How to read a calibration certificate for thermo hygrometer General Measurement Device and Calibration Topics 2
L Calibration Frequency for Surftest SJ-400 General Measurement Device and Calibration Topics 0
C External Laboratories and OEM Calibration IATF 16949 - Automotive Quality Systems Standard 0
V Ammeter calibration - Measuring head (on pic.) problem General Measurement Device and Calibration Topics 1
F How many signatures do we need on calibration certificates? ISO 17025 related Discussions 8
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
A Calibration Certificate Results General Measurement Device and Calibration Topics 3
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
Similar threads


















































Top Bottom