Calibration Interval - AS9100 - Control of Monitoring and Measuring Equipment

A

A6Intruder

#1
Having an 'animated' discussion with a peer whose opinion I value highly; however on this topic we disagree, and seeking some thoughts from the assembled multitude of QA pros.

As9100 (whatever) tells us that control of monitoring and measuring equipment is to be done at specified intervals, and that we can set them based on whatever criteria we choose/deem appropriate.

The gear in question happens to be some floor fitting interface fixtures, used to 1) locate the monument-to-floor fittings prior to bonding. The locations are derived from the model and the fixtures were machined and then digitally scanned to a magnitude greater than the tolerance of the fittings' locations. And yes the scanning equipment itself was all calibrated.

Here is the point of contention: since these plates are made of 1/2" thick aluminum any damage that affects the fitting interface will be quite visible and because of that, I called for a 'One Time Only (OTO)' cal interval. The quality plan for this project and project unique cal process defines this and also allows it as long as no damage to the fixture is present.

My friend does not accept a OTO; instead he maintains that I should have some interval, even if it is a 5 year interval.

When I ask him why, he just repeats himself. I don't want to put us in a position of missing a 5 year window, and at the same time am confident any damage that would impact the fitting locations would be blatantly obvious.

Your thoughts and comments please.

Thanks.
 
Elsmar Forum Sponsor

outdoorsNW

Quite Involved in Discussions
#2
Please provide some more details about why you think damage will be quite visible.

It seems like you could have damage that is just enough to make the fixture not work correctly but might not be obvious. Being a softer metal, aluminum may wear over time and the fixture may slowly shift away from the desired alignment.
 

Sidney Vianna

Post Responsibly
Staff member
Admin
#3
The gear in question happens to be some floor fitting interface fixtures, used to 1) locate the monument-to-floor fittings prior to bonding. The locations are derived from the model and the fixtures were machined and then digitally scanned to a magnitude greater than the tolerance of the fittings' locations. And yes the scanning equipment itself was all calibrated.

Your thoughts and comments please.

Thanks.
From what you describe, this sounds much more like an assembly fixture, rather than an inspection, measuring, test device. If it is indeed a production tool, then the requirement you are concerned with does not apply. It would have to be validated and maintained, but not "calibrated".
 

howste

Thaumaturge
Super Moderator
#4
I agree with what Sidney said. Calibration is required for measuring equipment. Is there something else that is used to measure the product that will detect a problem if the fixture is damaged or wears? If so, this is what should be periodically calibrated.

As far as maintenance, aluminum fixtures are very prone to wear. If it's being used frequently then I would expect that it would wear over time and eventually won't be able to hold a tolerance. I don't know what kind of tolerances you have, but this should be considered. Periodic verification of the tooling or ongoing measurement of resulting product would detect this.
 
A

A6Intruder

#5
Thanks for the inputs; I see that I wasn't clear ( working from a position of 'insider trading'):

The fixture has bolt holes that the floor fittings are fixed to. Then the entire fixture/fitting assembly is mated with our monument. the monument is constructed of honeycomb core panels and the area where the fittings are inserted is cleared away and after the fittings have been inserted, we then bond them in place. The check fixture remains until the adhesive cures. Since the through holes for the bolts is not an interference fit, any wear to these holes would be insignificant. My tolerance is +/- 0.030" on the monument, and by fixture tolerance was +/- 0.010" for the bolt hole centers.
 

Eredhel

Quality Manager
#6
I'm leaning toward what Sidney said as well. It almost sounds like it's a production check fixture that doesn't need to be handled by quality for acceptance. Is this used for acceptance? If not it sounds like production upkeep.
 
A

A6Intruder

#7
We use it as to verify location, so yes it is an acceptance tool and thus requires calibration. It is called out in not only the manufacturing drawings, but also in the customer's source inspection work order. The floor fittings have no adjustment and MUST fit, so this is our only way to ensure there is no installation issue.
 

howste

Thaumaturge
Super Moderator
#8
We use it as to verify location, so yes it is an acceptance tool and thus requires calibration. It is called out in not only the manufacturing drawings, but also in the customer's source inspection work order. The floor fittings have no adjustment and MUST fit, so this is our only way to ensure there is no installation issue.
I agree with your coworker that if calibration is required, a one-time calibration is insufficient. AS9100 says:
AS9100 Rev D clause 7.1.5.1 said:
The organization shall ensure that the resources provided:

a. are suitable for the specific type of monitoring and measurement activities being undertaken;

b. are maintained to ensure their continuing fitness for their purpose.

The organization shall retain appropriate documented information as evidence of fitness for purpose of the monitoring and measurement resources.
Bullet point "a" would be covered by initial calibration. Bullet point "b" would be covered by periodic verification/calibration. The last sentence requires records as evidence that it continues to be fit for use. Also there is the part of 7.1.5.2 that you mentioned including "specified intervals" which would not be met with a one-time calibration.
 

dgriffith

Quite Involved in Discussions
#9
There are many programs that have calibrate (or verify) before use categories. There is no date-driven interval. A mechanical fixture with dimensional requirements seems like a good candidate.
 
Thread starter Similar threads Forum Replies Date
J Documenting a Calibration Interval Extension Calibration Frequency (Interval) 7
D Torque Wrench Calibration Interval Extension with Non-Use? General Measurement Device and Calibration Topics 1
Marc Calibration Frequency (Interval) Discussions Sub-Forum General Measurement Device and Calibration Topics 0
K Calibration Interval based on Usage Calibration Frequency (Interval) 14
J Digital Scale Re-Calibration Interval Frequency General Measurement Device and Calibration Topics 10
M A Guide or Standard for Calibration Interval (Frequency) Calibration Frequency (Interval) 2
D "As Found" Condition and Calibration Interval Changes Calibration Frequency (Interval) 5
B Calibration Interval of Munsell Color Charts General Measurement Device and Calibration Topics 3
C Furnace Classes in relation to AMS 2750 to determine Calibration Interval Calibration Frequency (Interval) 2
A How to Determine Machine Tool Calibration Interval? General Measurement Device and Calibration Topics 4
Z CMM (Coordinate Measuring Machine) Calibration Interval (Frequency) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
MarilynJ6354 Torque Wrench Calibration Frequency (Calibration Interval) Calibration Frequency (Interval) 38
D Nonadjustable M&TE - 10 year calibration interval? Pin Gages Calibration Frequency (Interval) 11
somashekar Calibration interval (Clause 7.6 ISO 13485:2003) - Biomedical test instruments ISO 13485:2016 - Medical Device Quality Management Systems 4
D Thermometer Calibration and Calibration Interval ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J What's your interpretation of Clause 7.6 - Calibration 'Interval'? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 16
J Recommendations on calibration interval (frequency) ISO/IEC 17025:2005 para 5.10.4.4 ISO 17025 related Discussions 5
A Calibration Interval (Frequency) Adjustment using Hilliard type 2 algorithm Calibration Frequency (Interval) 11
L Projecting Calibration interval for Resistors Calibration Frequency (Interval) 2
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 0
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 4
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 5
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32

Similar threads

Top Bottom