Calibration Items - Necessary to include the hand crimp tools into the system?

I

IanHoughton

#1
I am reviewing our calibration system that consists of tape measures, hand crimp tools, continuity testers and multi-meters. My question is this:
Is it really necessary to include the hand crimp tools into the system, as they are neither inspection or test equipment. If they don't need to be calibrated what other procedures (first off tests) needs to be carried out?
 
Elsmar Forum Sponsor

Jerry Eldred

Forum Moderator
Super Moderator
#2
If you are using the crimp tools on product or on anything where you either specify the reliability of the tensile strength of the crimp, or for any other reason care about the quality of the crimp, they need to be periodically verified. I am not an expert in that area, and its been a while, but I believe there are a couple of methods to verify them. I believe one is verification of the pull strength of the crimp, and the other is a measurement of the crimped connection size. Depending on your specifics, as to which one would apply. They are not inspection equipment, but if used on product, they are production equipment, with the capability to impact product quality.

Any measurement instruments you use that can impact product quality (multitesters, etc) need to be verified/calibrated periodically.
 
I

IanHoughton

#3
Thanks Jerry,
The crimp tools at present are calibrated every 6 months but this takes up a lot of my time. I was hoping to pass on the responsibility for checking crimps to the operator who will have to carry out first off pull tests before starting each batch. (These will be recorded by batch number of harness) These tests are carried out at present on a calibrated pull test machine. Will this meet ISO9000 & QS9000 requirements ??
 

Jerry Eldred

Forum Moderator
Super Moderator
#4
Ian,

Bear in mind in my replies that I am a Metrologist (22 years), and not necessarily a QS9000/ISO9000 guru.

I think that if you have a documented procedure for checking a harness with a properly calibrated pull tester at the start of each lot is a good direction. I might suggest that you button up the possibility for bad units to escape by checking one at the end of the lot also. That way you are validating that all connections made by the crimper in between "should" be good. The metrologist and quality engineer in me also thinks that a random sample be externally tested. Especially if you are not going to calibrate the crimpers. Perhaps as a part of your outgoing inspection.

Conversely, it is not that great of a chore to do a verification on the crimpers. Depending on what type of connector you are crimping, it may be something as simple as measuring a crimp diameter with a digital caliper.

And even a third method might be doing that caliper measurement a number of times during each lot (or the pull tester data - determine what pull force it takes to destructively pull a connection apart), and plotting that dimension on an SPC chart (X bar R). When you see a trend past a control
limit, the process is out of control and you make adjustments to bring the process back into control.

My instincts tell me, however, that only the one non-destructive check at the beginning of the lot does not guarantee that the crimp will be in spec for the whole lot. If you have a spec strength of a 10 (just an imaginary number for discussion), and you check to be sure the connector meets a 10, you may actually have a destructive strength of say 10.2. By the end of the lot, if your crimp fixturing is worn and close to the replace poit, perhaps that figure will change by the end of the lot to a 9.9. Thus, you will have had a portion of a lot which was escaped product.

Sorry if I am overly wordy. I would rather give a little too much info than too little.
 
D

David Guffey

#5
The crimper tool itself does not have to be in the calibration cycle. What you check the crimped connections with DOES have to be in the calibration cycle.

That is not to say that having the process tool (crimper) in the cal cycle is not a good idea, just that it's not necessary.
 

edgedirk

Starting to get Involved
#6
I have a question similar in nature. I have tons of hand crimp tools here. An ISO inspector told me that each hand crimper has a life expectancy and that the manufacturer can provide how many crimps the tool can have before it is considered non-useable.

First of all is this true?

Secondly, If these tools have been around a long time how would I know of a tool that is considered not useable by an ISO auditor? Tossing out all current hand tools is not a option to due quantity that I have and the cost of buying new crimping tools is not feasible.

Any suggestions on how to handle hand held crimp tools?
 

Marc

Fully vaccinated are you?
Staff member
Admin
#7
Every tool has a life expectancy, some of which are affected by whether maintenance/repair/refurbishment is possible and cost effective. I really don't know enough about hand crimpers to say other than I would expect the manufacturer of anything to be able to give an expected life. In this case it would probably be the a certain number of crimp "cycles". But - So what? You are the ones who determine when a crimper goes bad and has to be disposed of (I assume they are not repairable). If an auditor told me I had to get rid of a tool based upon the manufacturer's end of life prediction, I would immediately say:

"Where does it say that? Please show me that requirement in the standard."
<snip> how would I know of a tool that is considered not useable by an ISO auditor? <snip>
Your crimpers are fine until the test (see the previous posts in this thread) indicates they are bad. As long as a crimper is not producing nonconforming product, it is OK to use.

My opinion is the auditor is on a fishing trip which has gone beyond his/her scope.
 
#8
As Marc has stated, the crimpers do, just like any hand tool, have a finite life. Typically, it is usual to monitor them by making sample crimps and measuring the crimp height and then performing a pull-off test to see what the effect is on performance. Some crimper manufacturers sell little "gauge blocks" to test the jaws when closed. I'd still go after crimp testing, myself. I've done thousands of the things.

If the auditor actually knew, they might have been helpful enough to tell you this stuff (it's not exactly consulting). And, please don't put them into the calibration system. If anything, they need to be part of your PM procedures.
 

dgriffith

Quite Involved in Discussions
#10
At our facility all crimpers, strippers, and the like are in the recall system. All have a tag affixed to them with a due date. It's been this way since You-Know-Who was a baby.
 
Thread starter Similar threads Forum Replies Date
D Determining Calibration Frequency schedule for items used in production Manufacturing and Related Processes 2
K Is there an overall standard for calibration of items used in TS 16949 Certified Co. IATF 16949 - Automotive Quality Systems Standard 3
G Which items require Calibration? General Measurement Device and Calibration Topics 18
G Understanding AMS 2750D compliant calibration report - items under 3.2.6.2 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
A AS9100 - 7.1.5.2 In-House Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
B UV-A and Illuminance meter Calibration Setup General Measurement Device and Calibration Topics 0
S Calibration Related Posters General Measurement Device and Calibration Topics 2
F Quality Control in Electrical Calibration ISO 17025 related Discussions 0
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6

Similar threads

Top Bottom