Calibration Lab unable to issue Accredited Certificate when error is too high?
N
neoth
Hi all,
is it true that a calibration lab is unable to issue a ISO 17025 accredited cert if the UUT is found to have high error or does not meet the manufacture's accuracy?
W
woodynz
Hi
No not true
It is possible for a lab to issue a certificate - but a number of things are happening here.
Remember calibration, accreditation and manufactures's specs are all different things.
1. Calibration is about reporting what you find before and after adjustments. Has nothing to do with manufacturers specs.
2. 17025 Accreditation is about having your processes, equipment and people externally audited buy a accreditation body - again nothing to do with manufactures specs.
3. Manufactures specs well they do what they want to do.
So a lab can issue a endorsed certificate even if the results show a fail and those results are outside the manufactures specs - unless the calibration makes a compliance statement to some standard or guide.
It is up to the lab to report the results and then up to the customer to make a decision based on those results. But as I mention above if the calibration you are doing is required to meet a standard say like EN 837-1 and the UUT doesn't meet that standard then you can't say it does but you can still issue a failed 17025 endorsed calibration certificate. Saying it does not comply. But it is still a ISO 71025 endorsed calibration certificate. Pass or fail. Or even no statement about pass or fail.
But always a but - special calibrations may override everything I have said above - depends on the labs terms of registration and what they are able to calibrate and issue certificates on.
Make sense ?
dv8shane
Hi all,
No it is not true, unless a statement of conformance is asked for by the customer or part of the labs procedures. All that is required for results is the measurand and the uncertainty of the measurement system at each point measured including the UUT's contribution.
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
H Can a calibration lab still use ISO 6789:2003 as guide for torque calibrations Other ISO and International Standards and European Regulations 1
S How many of you are using Robotic process automation for calibration lab management? ISO 17025 related Discussions 0
S The procedure for replicate check in calibration lab as per ISO 17025 General Measurement Device and Calibration Topics 2
S Is it mandatory for a Calibration Lab to be ISO 17025 accredited? General Measurement Device and Calibration Topics 7
R (OOT) Measurement Equipment after the calibration at a third Party lab General Measurement Device and Calibration Topics 3
SATHYABALARAMAN Calibration Laboratory vs. Standards Lab Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
K Calibration & Testing Lab here in the States - RA Audit - Wow! General Measurement Device and Calibration Topics 8
S Setting up a Scale Calibration Service lab - Where to start? General Measurement Device and Calibration Topics 7
C Setting up Pressure Calibration In-House Lab General Measurement Device and Calibration Topics 7
S FDA guidelines for Calibration Lab needed US Food and Drug Administration (FDA) 2
C Wabash Press Calibration - Lab-sized Wasbash Presses General Measurement Device and Calibration Topics 4
D ISO/TS16949 Internal Calibration Lab need to include Production Equipments IATF 16949 - Automotive Quality Systems Standard 26
J Industry Standard for turning in gages to the Calibration Lab? General Measurement Device and Calibration Topics 2
S Applying principles of lean to a calibration lab? Lean in Manufacturing and Service Industries 8
R In-house Calibration and still maintain my "we do not have a lab" status? IATF 16949 - Automotive Quality Systems Standard 10
D ISO/IEC 17025 required for a calibration lab to be ISO certified? ISO 17025 related Discussions 6
A Looking for Calibration Lab's on Reynosa, Mexico ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
J Calibration lab reqts for QS9000 (TS also?) QS-9000 - American Automotive Manufacturers Standard 2
J Calibration Lab Requirements for QS-9000 (TS also?) QS-9000 - American Automotive Manufacturers Standard 2
K Dimensional Lab (CMM Room) vs. Physical Test/Calibration Lab - Same required temp/hum General Measurement Device and Calibration Topics 2
C Supplier Lab Requirements/Calibration Services QS-9000 - American Automotive Manufacturers Standard 3
N IATF Calibration Lean in Manufacturing and Service Industries 2
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
W Next Calibration Due Date Calibration Frequency (Interval) 5
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3