Calibration Labels - Original Calibrating Date or New Date?

Raffy

Quite Involved in Discussions
#1
Hi Everyone,
Last week we were been audited by a customer and he suggest that during extension of calibration of a certain equipment, we must still use the original date in which it tells the expiring date, e.g. Calibration Date: 06/01/01, New Calibration date: 07/01/01. Instead of the date of new calibration date to be used, we must used the original date. Does it reflected on the new standard (ISO9001:2000)? Is there any clauses wherein we may deviate? Please advice.
Any feedback would be appreciated.
Thanks in advance,
Raffy
 
Elsmar Forum Sponsor
D

D.Scott

#2
Raffy - Not quite sure where your customer is headed with this but if we are talking about the calibration label on the equipment, IMHO it should show the date of last calibration and the due date of next calibration along with the initials of the technician. There must also be a unique identifier for the equipment to trace back to calibration records.
That is where there is a requirement that original date of purchase, first calibration, repairs, etc. should be kept.
I would try to get further clarification from your customer on what he is looking for.

Dave
 

Jerry Eldred

Forum Moderator
Super Moderator
#3
Extension of calibration interval in my context is used to temporarily give permission under extraordinary circumstances to use the equipment past it's assigned calibration interval. For example, a unit with a 12 month interval is calibrated on June 15th 2000, and due for re-calibration on June 15th 2001. The user contacts the lab and says it is in the middle of automated testing that cannot be interrupted, and which will complete on July 13th. I review results of a few previous calibrations , and find that the unit has stayed in tolerance for many years. Based on a high confidence level that it will remain in tolerance for the additional one month extension, I add a thirty day extension to the interval.

It is important to still reflect the original calibration date, and perhaps use a special label identifying that this is an extended interval. In no circumstance should the actual calibration date be changed. The label must reflect when the unit was last actually calibrated; that is not negotiable.
 
J

JerryStem

#4
We sometimes run into the same problem here. Our tech's and customers may not be able to schedule together in a certain month (we do site cal's also).

We have a special procedure where the customer runs a "stability loop" on the instrument (fairly simple procedure) and faxes the results to us. We then issue a decal to place next to the original cert decal that has the original end date and new end date (+30 days). We also issue a certificate that details what we/they did, with dates.

Interestingly, we changed our cert/recal dates to a month/year format & that has reduced the problem considerably. (Gives the customer & our tech a whole month to get together).

Jerry
 
A

Al Dyer

#5
Raffy,

The customer is always right! :rolleyes:

Don't sweat the small stuff. Write a clause into your procedure or instruction that covers calibration date extensions (and follow it).

ASD...
 
T

Trakman

#6
Jerry,

Good point on the calibration extension. For Raffy, keep in mind that the responsible department defines the interval, ISO does not. The important thing is that an interval is determined. Some equipment may be extended to two years (depending on many factors of course!) but this can still be done. (corrections/beatings/yelling will be tolerated if I am out to lunch...)We recently ran into the catch-22 situation; the customer owned the test gear, but required it to be within calibration. The calib date neared and passed, with numerous requests for approval to calibrate, but was not received. (they are a massive company...) Nevertheless, we still were contractually bound to deliver product using their test equip! The entire situation was completely documented. (very important!) We ended up being forced to utilize the "out of calib" equipment. Yuk!
Trakman

------------------
 
G

Graeme

#7
Originally posted by Jerry Eldred:
... I review results of a few previous calibrations , and find that the unit has stayed in tolerance for many years. ...
Another thing you may want to consider if you have "many years" of data is to actually increase the calibration interval. If you have the data to back it up and use a recognized method to calculate the new interval there should not be a problem. A side benefit to the customer is that they save money by not having to send the instrument for calibration as often.

You must have enough data to make a statistically significant estimate of the new calibration interval. That can be a problem if you don't have many of a particular model or type. See NCSL RP-1 for very detailed information on this topic.

As for documenting, what we did a a Naval Shipyard (now closed) that I worked in was document the temporary extension in the database, and apply a new label with the work "extended" and the lab manager's initials in the original cal date space, and the new due date in the ... due date space. This label was placed NEXT TO the original one. (The organization I am working with now prohibits extending the due date at all, for several very good reasons that don't apply in most industries.)


------------------
Graeme C. Payne
ASQ Certified Quality Engineer
[email protected]
 
A

Al Dyer

#8
Upon reflection I would like to ammend my previous statement.

-Are there back-up gages available? If no, why not?

-Should this be addressed during FMEA/contract review? Yes

-Is there a robust contingency plan? Should be.

I'm leaning towards the position that extensions should not be "way out" but that pre-planning is the methodology of a company that is pro-active and does the job right the first time.

Considering the variables inherent to floor and lab gages I don't believe that past history is a good reason to let a gage pass calibration dates.

-How many times has a caliper been dropped?
-How many times has the variation in air pressure affected an air gage?
-How many times has the temperature in a controlled lab deviated?
-How many times have gages been stolen and sold at the local flea market? (yes it happens and what is the contingency?)

Just a renewed opinion,

ASD...
 

Raffy

Quite Involved in Discussions
#9
Hi Al,
I agree with you that "Customer Is Always Right". However, when we try to expound him the idea of using the original date, we realize that as we use the original date its like fooling ourselves in extending the calibration interval.
quote:
---------------------------------------------
Originally posted by Al Dyer:
I'm leaning towards the position that extensions should not be "way out" but that pre-planning is the methodology of a company that is pro-active and does the job right the first time.
I think this is the best thing we must do to have a pre-planning.
Thank you very much and for all who participate in this discussion.
Raffy
[email protected]
 
Thread starter Similar threads Forum Replies Date
Tagin Recommendations for calibration labels for ESD floor mats General Measurement Device and Calibration Topics 9
S Calibration Labels Per Piece or is a Listing Acceptable? General Measurement Device and Calibration Topics 4
D Calibration labels - Does all measuring equipment require a label to indicate status? General Measurement Device and Calibration Topics 8
D Calibration Labels - Is it a requirement that you spell out the word calibration? General Measurement Device and Calibration Topics 2
D Gage Labels - Is the gage number on the calibration label enough? General Measurement Device and Calibration Topics 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 5
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 3
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2

Similar threads

Top Bottom