Calibration Method of Dial Thermometer Gauges
N
nurhakim
Hi All,
I am looking for calibration method for dial thermometer gauge, please if someone could help.
The dials are in range from 0-200 deg C and 0-400 deg C.
Thanks.
Regards,
Nur
harry
Are you talking about Calibration or Verification (sometimes erroneously termed in-house calibration)?
N
nurhakim
I might used incorrect term. Suppose to call it verification. I have 20 units to verify against master device which is sent out to be calibrated by external lab.
Regards,
Nur
BradM
Do you have any calibration equipment like a recirculating oil bath or a dry block calibrator?
What kind of accuracy are you seeking to measure?
N
nurhakim
I do not have any. I read somewhere that the device should be verified against master device,in a hot oil (I am not sure what kind of oil) at 3 different temperatures: 85 C, 175 C and 225 C. I am not sure whether this procedure is correct.
Nur
dgriffith
Nur
Oil at those temperatures, or any fluid for that matter, is nothing to fool around with, especially if you are not experienced. Better to send it out for 2- or 3-point calibration. Get it with data (more expensive) or without (less).
A stirred bath and all that goes with the process (fume hood, etc.) will cost you thousands upon thousands.
Let thermometry professionals handle it.
dgriffith
Hi All,
Thanks.
Regards,
Nur
May I ask what type of dial thermometer? Some folks say you could stick your finger into tepid water and better guess the temperature. Will they meet your needs?
harry
Nur
You can use the attached for information and as a reference. You have to answer the question on the kind of accuracy you need and the usual temperature range that you use for further decisions..
  • TechNote.ProceduresForVerifyingTemperatureCalibration.20091028.pdf
N Calibration Method for Scales and Flowmeters Reliability Analysis - Predictions, Testing and Standards 7
B Dosimeter Calibration Method General Measurement Device and Calibration Topics 5
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
C One Inch Micrometer Calibration Method General Measurement Device and Calibration Topics 2
M Fixed Point Temperature Cells Calibration Method General Measurement Device and Calibration Topics 3
K Validation of an Analytical Method that relies on Multi-Variate Calibration General Measurement Device and Calibration Topics 7
A Calibration Method and Measurement Uncertainty General Measurement Device and Calibration Topics 5
D Calibration Process - What do they mean by "Check Method"? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
T ISO 17025 Clause 5.4.2 - Calibration and Test Method Requirements ISO 17025 related Discussions 6
D Calibration Method for Small Surface Plates (6" x 6") General Measurement Device and Calibration Topics 8
M Sorvall 100S Centrifuge Calibration - SOP (Method and Procedure) Needed General Measurement Device and Calibration Topics 5
Q Does this calibration method of a specific fixture make sense? General Measurement Device and Calibration Topics 7
M Calibration of Load Cells in devices by a comparison method General Measurement Device and Calibration Topics 13
S NPT Thread Calibration Method - NPT Taper Thread Plug Gages General Measurement Device and Calibration Topics 9
B Method for Calibration of Thread Plug Gauges General Measurement Device and Calibration Topics 13
V Calibration method for a Neptune 10 Water Meter General Measurement Device and Calibration Topics 5
B Planekator Calibration Method Details - Calibration of a Surface Plate General Measurement Device and Calibration Topics 6
V Internal Scale and Balance Calibration Method Tolerances and Limits General Measurement Device and Calibration Topics 4
T Basis for Calibration Method and Intervals (Frequency) - FAA approved repair station Federal Aviation Administration (FAA) Standards and Requirements 10
Icy Mountain ISO/TS 16949 Compliant Calibration Method - Laboratory Calibration Levels IATF 16949 - Automotive Quality Systems Standard 13
K Calibration method for Torque Wrench & AC gas leakage Tester General Measurement Device and Calibration Topics 4
M Calibration Method According to ISO Guide 25 General Measurement Device and Calibration Topics 5
T Anyone have an example of a calibration method validation procedure? General Measurement Device and Calibration Topics 4
C IR thermometer for refrigerated Centrifuge calibration General Measurement Device and Calibration Topics 2
G ISO 17025.2017 Clause 7.8.4.3 Certificate of Calibration Label ISO 17025 related Discussions 5
M CMM Calibration - What, exactly, is the output? General Measurement Device and Calibration Topics 11
J Using MSA instead of calibration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
P Hierarchy of Calibration in Metrology General Measurement Device and Calibration Topics 1
J Calibration frequency for glass scale Calibration Frequency (Interval) 2
S Calibration SOP and external vendors General Measurement Device and Calibration Topics 2
P IATF16949 External Lab Calibration -CIPM MRA IATF 16949 - Automotive Quality Systems Standard 1
CaliperJim Acclimation time prior to calibration General Measurement Device and Calibration Topics 8
S How to report a measurement and relate it to calibration certificate General Measurement Device and Calibration Topics 1
R Do I need to get calibration certificate from ISO 17025 for IATF Audit IATF 16949 - Automotive Quality Systems Standard 8
S Netsuite Preventive Maintenance and Calibration Manufacturing and Related Processes 0
D Roughness specimen calibration Calibration and Metrology Software and Hardware 5
dhuot1990 General Calibration SOP Manufacturing and Related Processes 3
B Gage calibration Gonogo (calcul of cgk) General Measurement Device and Calibration Topics 0
Vader22 As-Found FAILED calibration ISO 14001:2015 Specific Discussions 16
M Gravimetric blender for resin in automotive industry - periodical calibration via external lab with certification Reliability Analysis - Predictions, Testing and Standards 0
T Calibration Tracking ISO 13485:2016 - Medical Device Quality Management Systems 4
Q Considering addition of mass calibration to ISO17025 scope ISO 17025 related Discussions 6
Z Resolution in calibration General Measurement Device and Calibration Topics 1
J Run Down Adapters Calibration Processes General Measurement Device and Calibration Topics 4
S Looking for accredited calibration lab - Onsite up to 650A and 1 kV. General Measurement Device and Calibration Topics 3
CaliperJim Screw Thread Micrometer Anvil Calibration General Measurement Device and Calibration Topics 7
Strange Juan HP 85054B Standard Mechanical Calibration Kit Calibration Procedure General Measurement Device and Calibration Topics 1
H Air Pressure Gage Calibration General Measurement Device and Calibration Topics 6