Hi ,
Can anyone out there advise me on this issue.
We are measuring the input voltage,current,wattage and power factor with the use of 4 digital panel meters that has got no adjustment watsoever.
We read and record the values for quality jugdgement purposes. Since there is no way for us to do calibration for these meters, we verify the voltmeter and ammeter against a calibrated voltmeter and ammeter at the start of every shift to ascertain that they are reading the same ( within the acceptable tolerance limit). Or maybe we do it once a month or so to verify its accuracy. Is this method acceptable? Or do we need to send these meters for calibration with an outside agency?.
Your inputs please.
Thanks
 
Elsmar Forum Sponsor

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#2
What are these used on? Are you referring to panel meters on process equipment?
 

somashekar

Staff member
Super Moderator
#3
Hi ,
Can anyone out there advise me on this issue. <snip>
Digital panel meters are factory calibrated using a calibrated source and calibration software. The basic measurement that you can calibrate yourself as you have detailed (voltage / current / frequency) should be fine.
Other power and energy outputs are derivatives from the basic measurement which the DPM software, embedded in the meter processor will perform and display suitably.
Any outside agency cannot do much, unless you send back to the factory for them to re-calibrate using the source and software that is propitiatory with them.
Note: I understand that you are referring to DPM's in DIN 96 x 96 mm or 96 x 48 mm cutout sizes in panels.
 
M

mooi mooi

#4
Thanks ..for the reply.
I think i'll just do as what i said earlier.
Was also wondering how could they calibrate tons of dpms in big refineries or process that have control rooms with plenty of such devices. I gues the quick and easy way would be just to verify them with calibrated measureing devices.
mooimooi
 

Big Jim

Super Moderator
#5
Thanks ..for the reply.
I think i'll just do as what i said earlier.
Was also wondering how could they calibrate tons of dpms in big refineries or process that have control rooms with plenty of such devices. I gues the quick and easy way would be just to verify them with calibrated measureing devices.
mooimooi
Verification is calibration activity.

"When necessary to ensure valid results, calibration equipment shall be calibrated or verified, or both . . . "

And don't forget the beginning of that sentence "When necessary to ensure valid results . . . ". If it isn't necessary for valid results it doesn't need to be calibrated.
 
Thread starter Similar threads Forum Replies Date
C Is calibration of digital panel important in any semicon company? General Measurement Device and Calibration Topics 3
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
D AMS 2750E - Digital Instrument Calibration Accuracy Requirements General Measurement Device and Calibration Topics 1
A Automated calibration of Digital Multimeter without communication port Calibration and Metrology Software and Hardware 2
D Hand-held Digital Tension Meter Calibration General Measurement Device and Calibration Topics 1
J Digital Scale Re-Calibration Interval Frequency General Measurement Device and Calibration Topics 10
W Digital Multi-Meter Calibration Procedure for various Meter Brands and Functions General Measurement Device and Calibration Topics 7
A Fluke Digital MultiMeter Calibration Traceability and Certificate General Measurement Device and Calibration Topics 11
N Thermocouple Calibration - Digital Display vs. Converting the Millivolt into ?C General Measurement Device and Calibration Topics 8
N Digital Temperature Indicator Calibration (Thermocouple) - Using a reference junction General Measurement Device and Calibration Topics 1
T In-House Calibration - Digital Micrometer found Out of Calibration General Measurement Device and Calibration Topics 10
V Calibration of a Digital Vernier - Need to check/calibrate/validate all 3 functions? General Measurement Device and Calibration Topics 9
B Tolerance for Digital Indicator Calibration - Machinists Indicators General Measurement Device and Calibration Topics 1
T Calibration of a digital usb temprature probe. General Measurement Device and Calibration Topics 3
O Calibration Supplier needed for Digital Temperature/Pressure Controllers General Measurement Device and Calibration Topics 6
O Calibration tolerances needed - Calipers, digital indicators, feeler gages, etc General Measurement Device and Calibration Topics 4
S Safeguarding Calibration on Digital Equipment General Measurement Device and Calibration Topics 25
F Digital Multimeter - Verification vs. Calibration MSA (Measurement System Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J Digital Caliper Calibration vs. Verification - Full Scale vs. Used Range General Measurement Device and Calibration Topics 8
M Calibration of Timers to NIST - How are digital and analog timers calibrated? General Measurement Device and Calibration Topics 15
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3

Similar threads

Top Bottom