Calibration of equipment
#1
Dears,
To check the light conditions on the visual inspection stations (acc VW req. min 1000 lx) we use standard luxometer. Of course, the luxometer needs to be calibrated, but is there nessesary to do that in accredited labolatory?
Elsmar Forum Sponsor
#2
Dears,
How confident in the result do you want to be?
malasuerte
Involved In Discussions
#4
Dears,
To check the light conditions on the visual inspection stations (acc VW req. min 1000 lx) we use standard luxometer. Of course, the luxometer needs to be calibrated, but is there necessary to do that in accredited labolatory?
Answer: It depends. Are you calibrating it in-house or sending it out??
  • If sending it out, per IATF, the answer is a definitive, YES.
  • If it is done internally, the answer is no, but you have to show that you are performing the calibration in accordance with a standard or you would need to define the standard.
Thread starter Similar threads Forum Replies Date
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Calibration Equipment Selection General Measurement Device and Calibration Topics 2
S ISO17020 - Equipment calibration requirement General Measurement Device and Calibration Topics 3
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
B Will you please help me with my academic project? Repair & calibration service for test equipment ISO 17025 related Discussions 0
G Is the calibration due date of equipment used to be on calibration certificate ISO 17025 related Discussions 38
D How to set equipment calibration/verification criteria General Measurement Device and Calibration Topics 1
T In house Calibration? Measurement equipment traceability requirements - ISO9001 General Measurement Device and Calibration Topics 12
C AS9100D 7.1.5.2 / Calibration of Employee Owned Measuring Equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
I Tool maintenance and calibration procedure - Calibrating tools/equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Q AS9100/AS1180-1 - Handling of Equipment and Calibration Records AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 12
M Measurement Equipment - Identification of Calibration Status General Measurement Device and Calibration Topics 25
W Equipment for Oven Calibration, Uniformity and System Accuracy Tests AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
A Calibration Interval - AS9100 - Control of Monitoring and Measuring Equipment Calibration Frequency (Interval) 8
M CO2 Incubators - Equipment Qualifications in addition to Calibration Qualification and Validation (including 21 CFR Part 11) 5
Raffy ISO 13485 NIST Traceability requirements for calibration of test equipment ISO 13485:2016 - Medical Device Quality Management Systems 3
S Calibration of Laboratory Equipment in accordance with the manufacturer's procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
R Calibration of processing equipment - 21 CFR 820.70 Production and process controls 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M ISO17025 Testing Laboratory Equipment Calibration Requirements ISO 17025 related Discussions 4
P Calibration of Equipment Timers General Measurement Device and Calibration Topics 4
I Calibration Void Seals on Internal Equipment General Measurement Device and Calibration Topics 3
R (OOT) Measurement Equipment after the calibration at a third Party lab General Measurement Device and Calibration Topics 3
Gman2 Calibration of Equipment for Field Installers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 21
jkuil Calibration before and after relocation of GMP manufacturing equipment Calibration Frequency (Interval) 7
C How to state that Calibration Services & Equipment are Traceable to NIST General Measurement Device and Calibration Topics 5
L How to set up a basic electronic test equipment calibration laboratory General Measurement Device and Calibration Topics 3
V For Analytical Laboratory Equipment, is PQ = Calibration Criteria? Qualification and Validation (including 21 CFR Part 11) 2
D How Working Standard Calibration result use in Verification of Equipment General Measurement Device and Calibration Topics 7
L How to manage prolongation of Calibration Period on Equipment General Measurement Device and Calibration Topics 9
M New Calibration Laboratory Set Up Measurement Parameters and Equipment General Measurement Device and Calibration Topics 4
N Client Request for Reference Equipment Calibration Certificate General Measurement Device and Calibration Topics 3
S Calibration of Equipment used for Development General Measurement Device and Calibration Topics 3
M How to include Gas Measurement Equipment Servicing & Calibration Scope to ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D OEM Calibration of their Measurement Equipment IATF 16949 - Automotive Quality Systems Standard 4
B Audit of equipment after certified external calibration... ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Casana 7.6 How to document review of out-of-calibration equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Calibration of measurement equipment when product End Of Life is near General Measurement Device and Calibration Topics 2
H Measurement System in Control if the Equipment requires Daily Calibration? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S Control of the Content in the Master Calibration Equipment form Document Control Systems, Procedures, Forms and Templates 2
Uriel Alejandro How to Record Evidence for Verification/Calibration of Equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
R ISO 17025 M&TE (Measuring & Test Equipment) Calibration Requirements ISO 17025 related Discussions 4
R Incorporating New Electronic Test Equipment into our Calibration System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
H Determining Calibration Tolerance for new Leak and Occlusion Equipment Measurement Uncertainty (MU) 2
Similar threads
Top Bottom