Calibration of Glassware - Pipettes & Flask

M

Montse

#1
Hello:
In our last audit of QS 9000, one of the observations were that we need to count on with records of calibration for the pipettes and flask that we use in the laboratory.
I need to recognize that I have not any idea about this request of the 4.11 .
We can not calibrate the plastic instruments labs. Therefore as a corrective action we start to record in a log book the number of serie of the instruments and to solicit that the calibration department record this number in the software system. After one year, the pipettes will need to be replace for a news as our procedure calibration document solicit.

I will appreciate if somebody have had a similar experience with regar to this kind of lab instruments and to know their inputs.
 
Elsmar Forum Sponsor

Marc

Hunkered Down for the Duration
Staff member
Admin
#2
There are many glass items which you can have calibrated yet sometimes it's cheaper to replace them. If you are doing this, that is OK. Since you're identifying them in your calibration database, that fills the requirement. You might be able to get away with a 2 year calibration frequency on glassware like pipettes.
 
R

Ryan Wilde

#3
State bureaus of weights and measures do the lion's share of volumetric calibrations in the U.S. Many of these state labs actually do these services for free.

Note: Buying new glassware may only be a band-aid fix. The certificate of conformance provided with new items is not a calibration. Ask yourself this, "What, exactly, does it conform to?" If you contact your glassware vendor and inquire about calibration with actual data, traceable to National Standards, the price will go up substantially. A savvy auditor with any experience in calibration will know this, and you will be back to square one.

Ryan Wilde
 
Thread starter Similar threads Forum Replies Date
T Volumetric Glassware (pipettes) Calibration Uncertainty Measurement Uncertainty (MU) 8
T Need thoughts on calibration of Class A volumetric glassware General Measurement Device and Calibration Topics 3
D Aircraft Coffee Maker Repair Bench Glassware Calibration General Measurement Device and Calibration Topics 7
A Calibration of Type A Glassware in Testing Laboratory General Measurement Device and Calibration Topics 3
A Glassware calibration - Do we need to calibrate or recalibrate laboratory glassware General Measurement Device and Calibration Topics 3
R Glassware calibration - Must all A-class glassware be checked before use? General Measurement Device and Calibration Topics 2
Marc Calibration of Glassware and Liquid in Glass Thermometers General Measurement Device and Calibration Topics 2
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 4
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
A Calibration of Power Supplies AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 2
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Calibration Equipment Selection General Measurement Device and Calibration Topics 2
D Sample Calibration SOP for ISO 13485 General Measurement Device and Calibration Topics 4
L Calibration of Plug Gauges - Tolerancing General Measurement Device and Calibration Topics 1
L In house calibration - Our CMM's are outside certified can I use them to certify our standards General Measurement Device and Calibration Topics 4
R Standard Filter for Calibration Colorimeter ISO 17025 related Discussions 1
J Documenting a Calibration Interval Extension Calibration Frequency (Interval) 7
F Calibration certificate content - Average Value and True Value ISO 17025 related Discussions 6
B Oscilloscope - AC gain calibration General Measurement Device and Calibration Topics 6
G Is it common to do a repeatability check during calibration of an instrument? General Measurement Device and Calibration Topics 5
A Adjustment of Calibration Intervals (Shortening or Lengthen) Calibration Frequency (Interval) 6
A Calibration required by OEM General Measurement Device and Calibration Topics 2
S ISO17020 - Equipment calibration requirement General Measurement Device and Calibration Topics 3
H Can a calibration lab still use ISO 6789:2003 as guide for torque calibrations Other ISO and International Standards and European Regulations 1
H Calibration of InfraRan Vapor Analyzer - where to get calibration for SF6 General Measurement Device and Calibration Topics 2
M How to read a calibration certificate for thermo hygrometer General Measurement Device and Calibration Topics 2
L Calibration Frequency for Surftest SJ-400 General Measurement Device and Calibration Topics 0
C External Laboratories and OEM Calibration IATF 16949 - Automotive Quality Systems Standard 0
V Ammeter calibration - Measuring head (on pic.) problem General Measurement Device and Calibration Topics 1
F How many signatures do we need on calibration certificates? ISO 17025 related Discussions 8
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
A Calibration Certificate Results General Measurement Device and Calibration Topics 3
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
G Third party auditor mentions no grace period for calibration Calibration Frequency (Interval) 22
E Calibration Certificate Content Requirements General Measurement Device and Calibration Topics 3
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
Nicole Desouza AS9100D - Calibration of Dies & Tooling - Plastic Extrusion AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 4
D Determining Calibration Frequency schedule for items used in production Manufacturing and Related Processes 2
WEAVER Electronic Weighing Scale Calibration Tolerance Manufacturing and Related Processes 1
I What kind of information needs to be in a calibration record? General Measurement Device and Calibration Topics 25
C Micrometer and Vernier Caliper simple calibration process General Measurement Device and Calibration Topics 8
D AS9100 Inspection records - Is there a requirement to have gauge ID and calibration status AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 5
I Determining Calibration Tolerance of a Measurement Device General Measurement Device and Calibration Topics 2
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
Similar threads


















































Top Bottom