Calibration of Hardness Tester

Wesley Richardson

Wes R
Trusted Information Resource
#11
There are several considerations for calibration of Rockwell Hardness testers. Start with ASTM E18-08b Standard Test Methods for Rockwell Hardness of Metallic Materials. ASTM E18 - 19 Standard Test Methods for Rockwell Hardness of Metallic Materials

When the indentor is changed, there is a requirement to "seat" the indentor. More than one reading is taken, often 3 readings, and the values averaged. Each indentation must be at least 2-1/2 indentation diameters from any other indentation, and at least 3 indentation diameters from an edge of the block. Indentations may only be made on one side of the block. The blocks cannot be resurfaced or ground to reveal a new surface. The indentor must be checked to assure that it is not cracked or chipped. The readings taken with a steel ball are now replaced with tungsten carbide balls and the HBS designation changed to HBW. Diamond indentors each give slightly different readings, depending upon the actual diamond.

The uncertainty of the results depend upon the type of indentor and the range being reported.

At a lab where I worked previously, we kept a log and each day that a different indentor was used, we reported the three readings, after seating the indentor. This was a validation and not a calibration. We used an Accredited calibration supplier that also did cleaning, oiling, adjusting of the indentation rate.

NIST has only within the past 10 to 15 years become involved with providing Standard Reference Materials (SRM) for Rockwell hardness testing, and then only for a limited number or hardness ranges, with HRC being the primary one. http :// www. nist.gov/msel/metallurgy/materials_performance/hardness-standardization.cfm - OBSOLETE BROKEN 404 LINK(s) UNLINKED - PLEASE HELP - REPORT POSTS WITH BROKEN LINKS

My recommendation is to select an ISO/IEC 17025 Accredited calibration laboratory, with a Scope of Accreditation which includes you tester, to perform the calibration. Then do the daily, or interval you select, verifications using traceable blocks.

Wes R.
 
Last edited by a moderator:
Elsmar Forum Sponsor
J

JaxQC

#12
Sorin,
# 1 Daily Calibration – Yes, but let’s call it a verification (confirming) rather than calibration (making adjustment) at this point instead, explain in minute. The standard below walks you through each level of verification.
# 2 Yes, outside source (
# 3 Refer to standard ASTM E 18 2007 Test Methods for Rockwell Hardness & ASTM E 140 Rev 2007 Standard Hardness Conversion Tables for Metals as a good starting point. . There are different layers of verification laid out in the standard.

There are (3) levels. Direct, Indirect, Daily.
Direct – When a machine fails an indirect check, looks at the indenter depth & the whole thing in detail etc..
Indirect – What you’d call calibration by an outside, certified company @ 12 month interval.
Daily – monitor performance, use standard blocks
 
S

Sorin

#13
Sorin,
# 1 Daily Calibration – Yes, but let’s call it a verification (confirming) rather than calibration (making adjustment) at this point instead, explain in minute. The standard below walks you through each level of verification.
# 2 Yes, outside source (
#3 Refer to standard ASTM E 18 2007 Test Methods for Rockwell Hardness & ASTM E 140 Rev 2007 Standard Hardness Conversion Tables for Metals as a good starting point. . There are different layers of verification laid out in the standard.

There are (3) levels. Direct, Indirect, Daily.
Direct – When a machine fails an indirect check, looks at the indenter depth & the whole thing in detail etc..
Indirect – What you’d call calibration by an outside, certified company @ 12 month interval.
Daily – monitor performance, use standard blocks
Heh...I know all that...was just asking becasue it was not clear from your original post.

Anyway, this is the form that I am using...see if it's any help.
 

Attachments

Thread starter Similar threads Forum Replies Date
A Calibration of Rockwell Hardness Tester IATF 16949 - Automotive Quality Systems Standard 10
B Additional Calibration needed for C-Frame Rockwell Hardness Tester? General Measurement Device and Calibration Topics 2
N Verifying Calibration of Hardness Gauge vs Actually Calibrating It General Measurement Device and Calibration Topics 2
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 5
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3

Similar threads

Top Bottom