Calibration of Humidity Instruments

D

D.Hutti

#1
I've recently been tasked with setting up our Laboratory to handle the calibration of Humidity Test Equipment. I ahave a couple of questions concerning this. 1)What are the governing documents for this type of calibration process? 2)What are the basic requiremnts for a Temp/Humidity Chamber?
 
Elsmar Forum Sponsor

Jerry Eldred

Forum Moderator
Super Moderator
#2
I believe there are ASTM specs for RH cals. Although I haven't used them because I have always had GIDEP procedures available, our RH tech should have those documents (he isn't here at the moment).

As for chamber requirements, there a number of options for your equipment. If you have chart recorders that do NOT have a detachable/movable RH probe (i.e.: the old Honeywell style, or similar), you will need a chamber (which I'll get to in a moment). If you have meters and/or chart recorders that have removable/detachable probes (i.e.: chart recorders similar to the Dickson or Omega or other brand name), and/or meters with probes, you may not need a full chamber.

Some of the options for the probe type meters are the HMK11 salt solution blocks made by Vaisala, or Rotronic makes some single use disks. General Eastern has a desiccator/RH generator that uses a combination of dry nitrogen to dry air, and saturation (it is normally externally monitored with a chilled mirror dewpoint meter (be careful on NIST traceability if you use that one). I am trying to be fair by mentioning all the vendors I know that have cal equipment for these. I am not recommending any of them.

If you need a larger chamber, Thunder Scientific is the only vendor I am familiar with that makes a large two pressure chamber.
This is considered by many to be top of the line for doing in-chamber cals of chart recorders.

Depending on your requirements, another option is to build a plexiglass chamber. If you build one of these, it involves taking sheet clear plastic, cutting it into squares and gluing together a cube shaped chamber with a hinged, airtight sealed door. Probably about 24-30 inches on each side, with a shelf about 6 inches below the bottom with many holes drilled through it, and with a fan to circulate the air. You use a pyrex or corning ware type lab quality glass cake pan, and pour salts or Assay grade in the pan, and spray distilled water on them. You then need to qualify the chamber with a high accuracy hygrometer, externally calibrated to verify uniformity. A good quality thermometer needs to be mounted in the chamber to monitor temperature as well. I don't reommend this method if it is a critical part of your process. Some places where this is not critical choose this method to save costs. I recommend the better methods.

Anyway, just a few ideas.

------------------
 
Thread starter Similar threads Forum Replies Date
T Calibration Laboratory Temperature/Humidity goes out of tolerance due to Power Loss General Measurement Device and Calibration Topics 9
D Buying a Temperature and Humidity Device, and Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
R Does Humidity affect Electronic Instrument Calibration? General Measurement Device and Calibration Topics 16
K Relative Humidity Requirement on Calibration Certificates General Measurement Device and Calibration Topics 5
D Temperature and Humidity requirements for a Calibration Laboratory General Measurement Device and Calibration Topics 11
P Temperature and Humidity Device Calibration - Came back as Out Of Calibration General Measurement Device and Calibration Topics 8
L Internal Calibration Laboratory Requirements - Ambient temperature, Humidity, Etc. Statistical Analysis Tools, Techniques and SPC 3
P Environmental Conditions for Calibration - Temperature and Humidity General Measurement Device and Calibration Topics 22
P Calibration temerature and humidity requirements General Measurement Device and Calibration Topics 4
A AS9100 - 7.1.5.2 In-House Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 18
B UV-A and Illuminance meter Calibration Setup General Measurement Device and Calibration Topics 0
S Calibration Related Posters General Measurement Device and Calibration Topics 2
F Quality Control in Electrical Calibration ISO 17025 related Discussions 0
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3

Similar threads

Top Bottom