Calibration of Humidity Instruments

D

D.Hutti

#1
I've recently been tasked with setting up our Laboratory to handle the calibration of Humidity Test Equipment. I ahave a couple of questions concerning this. 1)What are the governing documents for this type of calibration process? 2)What are the basic requiremnts for a Temp/Humidity Chamber?
 
Elsmar Forum Sponsor

Jerry Eldred

Forum Moderator
Super Moderator
#2
I believe there are ASTM specs for RH cals. Although I haven't used them because I have always had GIDEP procedures available, our RH tech should have those documents (he isn't here at the moment).

As for chamber requirements, there a number of options for your equipment. If you have chart recorders that do NOT have a detachable/movable RH probe (i.e.: the old Honeywell style, or similar), you will need a chamber (which I'll get to in a moment). If you have meters and/or chart recorders that have removable/detachable probes (i.e.: chart recorders similar to the Dickson or Omega or other brand name), and/or meters with probes, you may not need a full chamber.

Some of the options for the probe type meters are the HMK11 salt solution blocks made by Vaisala, or Rotronic makes some single use disks. General Eastern has a desiccator/RH generator that uses a combination of dry nitrogen to dry air, and saturation (it is normally externally monitored with a chilled mirror dewpoint meter (be careful on NIST traceability if you use that one). I am trying to be fair by mentioning all the vendors I know that have cal equipment for these. I am not recommending any of them.

If you need a larger chamber, Thunder Scientific is the only vendor I am familiar with that makes a large two pressure chamber.
This is considered by many to be top of the line for doing in-chamber cals of chart recorders.

Depending on your requirements, another option is to build a plexiglass chamber. If you build one of these, it involves taking sheet clear plastic, cutting it into squares and gluing together a cube shaped chamber with a hinged, airtight sealed door. Probably about 24-30 inches on each side, with a shelf about 6 inches below the bottom with many holes drilled through it, and with a fan to circulate the air. You use a pyrex or corning ware type lab quality glass cake pan, and pour salts or Assay grade in the pan, and spray distilled water on them. You then need to qualify the chamber with a high accuracy hygrometer, externally calibrated to verify uniformity. A good quality thermometer needs to be mounted in the chamber to monitor temperature as well. I don't reommend this method if it is a critical part of your process. Some places where this is not critical choose this method to save costs. I recommend the better methods.

Anyway, just a few ideas.

------------------
 
Thread starter Similar threads Forum Replies Date
T Calibration Laboratory Temperature/Humidity goes out of tolerance due to Power Loss General Measurement Device and Calibration Topics 9
D Buying a Temperature and Humidity Device, and Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
R Does Humidity affect Electronic Instrument Calibration? General Measurement Device and Calibration Topics 16
K Relative Humidity Requirement on Calibration Certificates General Measurement Device and Calibration Topics 5
D Temperature and Humidity requirements for a Calibration Laboratory General Measurement Device and Calibration Topics 11
P Temperature and Humidity Device Calibration - Came back as Out Of Calibration General Measurement Device and Calibration Topics 8
L Internal Calibration Laboratory Requirements - Ambient temperature, Humidity, Etc. Statistical Analysis Tools, Techniques and SPC 3
P Environmental Conditions for Calibration - Temperature and Humidity General Measurement Device and Calibration Topics 22
P Calibration temerature and humidity requirements General Measurement Device and Calibration Topics 4
S Netsuite Preventive Maintenance and Calibration Manufacturing and Related Processes 0
D Roughness specimen calibration Calibration and Metrology Software and Hardware 5
dhuot1990 General Calibration SOP Manufacturing and Related Processes 3
B Gage calibration Gonogo (calcul of cgk) General Measurement Device and Calibration Topics 0
Vader22 As-Found FAILED calibration ISO 14001:2015 Specific Discussions 16
M Gravimetric blender for resin in automotive industry - periodical calibration via external lab with certification Reliability Analysis - Predictions, Testing and Standards 0
N Calibration Method for Scales and Flowmeters Reliability Analysis - Predictions, Testing and Standards 7
T Calibration Tracking ISO 13485:2016 - Medical Device Quality Management Systems 4
Q Considering addition of mass calibration to ISO17025 scope ISO 17025 related Discussions 6
Z Resolution in calibration General Measurement Device and Calibration Topics 1
J Run Down Adapters Calibration Processes General Measurement Device and Calibration Topics 4
S Looking for accredited calibration lab - Onsite up to 650A and 1 kV. General Measurement Device and Calibration Topics 3
CaliperJim Screw Thread Micrometer Anvil Calibration General Measurement Device and Calibration Topics 7
Strange Juan HP 85054B Standard Mechanical Calibration Kit Calibration Procedure General Measurement Device and Calibration Topics 1
H Air Pressure Gage Calibration General Measurement Device and Calibration Topics 6
T Calibration on AVL AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
Sam.F Do heat treatment ovens need calibration? General Measurement Device and Calibration Topics 22
J Selecting reference standards for calibration General Measurement Device and Calibration Topics 3
C IATF 16949:2016 Major NC pertaining to 7.1.5.2.1 Calibration/verification records IATF 16949 - Automotive Quality Systems Standard 15
E Calibration Records during AS9100 Audit AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
John Predmore Calibration Determination decision Calibration Frequency (Interval) 15
J ISO 17025 Calibration Other Medical Device Related Standards 2
A Electronic Weighing scale calibration Calibration Frequency (Interval) 4
M Bourdon gauge calibration procedure failure ISO 17025 related Discussions 3
Ed Panek Calibration question Manufacturing and Related Processes 4
S Calibration/Worn gages Manufacturing and Related Processes 9
A Settings rings calibration General Measurement Device and Calibration Topics 4
R 7.1.5.2 -- Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
C CALIBRATION CERTIFICATE REQUIREMENTS General Measurement Device and Calibration Topics 11
S Calibration Label issues General Measurement Device and Calibration Topics 21
H Calibration / Capability in a PCP FMEA and Control Plans 1
R First Time Managing Calibration and Measurement System. Need Help. General Measurement Device and Calibration Topics 19
M CE marked medical device - do they need UKAS certified calibration provider ISO 17025 related Discussions 4
JDJohnson Micrometer & Caliper Calibration (ISO-9001) General Measurement Device and Calibration Topics 26
B Precision Current Loop for Current Probe Calibration Reliability Analysis - Predictions, Testing and Standards 2
Strange Juan Freezers fridges Incubators Calibration Labels General Measurement Device and Calibration Topics 12
D Tolerance and MPEs for Re-calibration Manufacturing and Related Processes 0
D ISO 9001:2015 Calibration Procedures Manufacturing and Related Processes 16
G Calibration of Angle microscope stage micrometer General Measurement Device and Calibration Topics 1
K END CALIBRATION CANNOT BE PERFORMED Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2

Similar threads

Top Bottom