Calibration of pH Meter Probe

Crimpshrine13

Involved In Discussions
#1
We have a pH meter that needs to be calibrated every day before use, and my assumption is that it is okay to use the calibration buffer solution that is standardized using a pH meter calibrated by means of standard solutions prepared from NIST standard reference material as long as I have the certificate of analysis, or is my understanding incorrect?

If my understanding is correct, how does everyone keep track of the calibration record for pH meter that needs to be calibrated every day? Should I be writing down the pH meter value for each buffer tested and corresponding buffer lot numbers, then keep the record of certificate of analysis? What is considered acceptable for IATF? pH meter is something new we recently added due to our new process, and we do not have other equipment similar to this to compare.

If anyone can pitch in some ideas for maintaining pH meter calibration records it's much appreciated. Thank you.
 
Elsmar Forum Sponsor

Ron Rompen

Trusted Information Resource
#2
I did something almost identical in a previous factory where we used pH meters. As you suggested, log the value of the buffer solution, lot number of the solution (this may be overkill, but it depends on the criticality of this measurement), the actual value reported by the meter, and of course the date tested and the initials/signature of the person performing the test. Also ensure that you have a documented reaction plan for when the meter and the buffer solution values don't match.
 

Crimpshrine13

Involved In Discussions
#3
I did something almost identical in a previous factory where we used pH meters. As you suggested, log the value of the buffer solution, lot number of the solution (this may be overkill, but it depends on the criticality of this measurement), the actual value reported by the meter, and of course the date tested and the initials/signature of the person performing the test. Also ensure that you have a documented reaction plan for when the meter and the buffer solution values don't match.

Thanks. I will probably do the same. I just wanted to see what everyone else do since this was a new thing in our facility. We'll probably have to keep track of the lot numbers because our calibration must be traceable to national standard as our quality system is IATF 16949 and must follow ISO 17025 standard.
 
#4
Hi....the pH of exceptionally purged water isn't precisely estimated by test units, or by pH meters. There are a few distinct explanations behind this, including the way that exceptionally cleaned water has almost no buffering limit, so its pH is effortlessly changed. Indeed, even the sharpness or basicity of a pH test unit's marker color is sufficient to change unadulterated water's deliberate pH. With respect to pH meters, the tests themselves don't work well in the exceptionally low ionic strength of unadulterated freshwater, and follow pollutions on them can swing the pH around a considerable amount.
 
Thread starter Similar threads Forum Replies Date
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
S Calibration - Isolated current meter - Measuring patient auxiliary leakage current Calibration and Metrology Software and Hardware 4
K Calibration of Dose Area Product (DAP) Meter in our Class IIb X-Ray Medical Device General Measurement Device and Calibration Topics 1
C Flow Meter calibration for mixed gases Calibration and Metrology Software and Hardware 1
D Hand-held Digital Tension Meter Calibration General Measurement Device and Calibration Topics 1
M Calibration of a Power Meter Model Hm8115-2 (Hameg) General Measurement Device and Calibration Topics 8
T Calibration procedure for Bubble Gas Flow Meter General Measurement Device and Calibration Topics 3
M In house Volt Meter Calibration General Measurement Device and Calibration Topics 5
R When does a pH Meter require Calibration? Calibration Frequency (Interval) 7
W Digital Multi-Meter Calibration Procedure for various Meter Brands and Functions General Measurement Device and Calibration Topics 7
S Calibration Frequency for a Measurement Device: Vibration Meter Calibration Frequency (Interval) 4
K Survey Meter Calibration - Accredited Laboratory required? General Measurement Device and Calibration Topics 5
C Flow Meter Calibration - Outside Calibration Service vs. In-House Calibration General Measurement Device and Calibration Topics 3
U Calibration of a Flux Magnet Meter - Any resource in Indonesia? General Measurement Device and Calibration Topics 3
E Calibration Frequency of Reference Magnets for Gauss Meter Calibration General Measurement Device and Calibration Topics 5
R Power Meter Calibration Accuracy Measurement Uncertainty (MU) 4
B In-House pH Meter Calibration General Measurement Device and Calibration Topics 13
V Calibration procedure for Torque Meter Needed General Measurement Device and Calibration Topics 6
R Calibration of on-line pH meter General Measurement Device and Calibration Topics 4
ScottK Soliciting recommendations for a fast turnaround calibration house - Mass Flow Meter Manufacturing and Related Processes 1
V Calibration method for a Neptune 10 Water Meter General Measurement Device and Calibration Topics 5
M Zebedee CL-100 Clarity Meter Calibration Procedure needed General Measurement Device and Calibration Topics 2
A Calibration? Landro meter (Waterbath) - Taking average of a few calibrated beakers General Measurement Device and Calibration Topics 2
S Gas Meter (Turbine, Rotary and Diaphragm) Volume Calibration General Measurement Device and Calibration Topics 4
M Illuminance Meter - Extending Manufacturer's Recommended Calibration Frequency Calibration Frequency (Interval) 3
G Thermocouple Meter Calibration - when authorities collide ... General Measurement Device and Calibration Topics 9
Geoff Cotton Noise Meter Calibration (in the UK) General Measurement Device and Calibration Topics 1
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 2
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 2
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 4
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6

Similar threads

Top Bottom