Calibration of Pin Gages - AS9100 Calibration Requirements
J
Jeff Frost
#11
Re: AS9100 Calibration Requirements - Calibration of Pin Gages
JZUBB and BRIZZLLA
Z class pin gages are not the only class of gages in use on shop floor in many companies. Just like gage blocks they come various tolerances such as:
Class XX .00002" (twenty millionths)
Class X .00004" (forty millionths)
Class Y .00007" (seventy millionths)
Class Z .0001" (one tenth)
As I said in my first posting calibration was based on using a super mic or bench mic which may have been applicable for the company I then worked for based on the class of the pin gage they were using. In any case you need to determine the method of calibration of these gages bases on requirements and the basic fundamentals of dimensional metrology.
Elsmar Forum Sponsor
J
jzubb
#12
Re: as9100 and calibration.
The mics are mitutoyo mics model 293-330 right on mitutoyo's web site they state the accuracy is .00005.
Thanks
Jon
J
jzubb
#14
Nope accuracy is what is says on there website.
Jon
#16
Jon
Jon:
I checked it out. Since it's a 'marketing' website, I stay with my comment; maybe they say 'accuracy', when they mean 'resolution'. The micrometers shown resolve down to .00005. If they were really quoting accuracy, they'd have to report uncertainty..........
Brizilla
Quite Involved in Discussions
#17
Their accuracy may be .00005 but since their resolution is also .00005 you'll never know. Their actual resoluton is .0001 since the 5 in the next decimal place is either 5 or 0 designation on these digital models.
#18
...and that could be a problem when calibrating a pin gauge!
D
Dave Dunn
#20
One point to remember about clause 7.6 is just because you say verification in place of calibration does not lessen the requirements contained within this clause. You must have the necessary calibration or verifications masters to detect that a Mic, Caliper, plug gage etc are either within the required accuracy or not when you perform the verification and that this process must be controlled conforming to all requirements of the clause.
ISO 9000:2005 (normative reference per AS9100b Section 2 & 3) defines verification as “confirmation, through the provision of objective evidence, that specified requirements have been fulfilled”. If the organization describes the processes in a documented procedure that reference one of the many calibration standards like MIL-STD-45662, ISO 10012, ISO 17025 or Z540 your calibration/verification process must also meet the requirements contained within the sited reference in addition to AS9100.
In summary your calibration or verification system for M&MD’s must meet the requirements of AS9100, customer imposed contractual requirements and the organizations own internal operational procedures.
PS. Yes I know that MIL-STD-45662 has been canceled but it still shows up as a contractual requirement in many government POs
Just as a point of interest too, MIL-STD-45662 defines calibration, not verification, as "comparison of M&TE or measurement standard of unknown accuracy to a measurement standard of known accuracy in order to detect, correlalate (sic.) report, or eliminate by adjustment any variation in the accuracy of the instrument being compared" So by that standard, what you're doing is in fact calibration. I'm not as familiar with other standards, so take it with a grain of salt. :2cents:
Thread starter Similar threads Forum Replies Date
D Calibration of Pin Gages using a Calibrated Caliper that measures to 4 decimal places General Measurement Device and Calibration Topics 15
N Calibration of Setup Fixtures & Pin Gages General Measurement Device and Calibration Topics 5
D Nonadjustable M&TE - 10 year calibration interval? Pin Gages Calibration Frequency (Interval) 11
M Calibration of Pin Gages - What are my options? General Measurement Device and Calibration Topics 6
C Black Oxide Gage Pin Calibration Requirements General Measurement Device and Calibration Topics 4
W Setting up a Zmike for Pin Gage Calibration General Measurement Device and Calibration Topics 3
C ZZ Pin Gage Tolerances and Calibration General Measurement Device and Calibration Topics 2
S Pin Gage Calibration - Pins seem to be "growing" General Measurement Device and Calibration Topics 9
K Pin Gage Use vs. Calibration Frequency Calibration Frequency (Interval) 13
M Gage pin calibration - Actual calibration readings required on the calibration cert General Measurement Device and Calibration Topics 3
C Gage pin calibration procedure for accurately calibrating gage pins General Measurement Device and Calibration Topics 18
B Pin Gage Calibration Procedure - ISO 17025:2005 accreditation requirements? ISO 17025 related Discussions 8
V Calibration of Pin Gauges General Measurement Device and Calibration Topics 2
C Pin Gauge Calibration using a Micrometer - Accuracy & Calibration Level Question General Measurement Device and Calibration Topics 10
C Internal Diameter Measurement and Pin Gage Calibration General Measurement Device and Calibration Topics 6
L Gage Pin Gauge R&R - Our calibration guy said he doesn't know how to do one Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 21
N Seeking Procedure(s) - Methodology: Calibration of Pin Gauges General Measurement Device and Calibration Topics 9
N Pin Gage Sets - Calibration costs - Just buy new ones? General Measurement Device and Calibration Topics 9
J Calibration, should we.... Go-No-Go, thread, pin, and ring gauges QS-9000 - American Automotive Manufacturers Standard 2
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 4
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
Similar threads
Top Bottom