Calibration of Pin Gages using a Calibrated Caliper that measures to 4 decimal places

DAHSR

Registered Visitor
#1
I have 30 Pin Gage sets to "calibrate". They are all 3 decimal place (X.XXX) and either a + or - accuracy. My plan is to use a calibrated caliper, that measures to 4 decimal places, that we maintain as a standard, not used for production, with documentation, a certificate of calibration and variable data traceable to NIST. Is there a standard in the world of calibration such as ISO, ANSI, etc. that would specifically not permit me to use calipers to verify the accuracy of each pin. The pins are not use for final inspection, that is done by QC, at the end of the manufacturing process. Our product dimensional tolerance is measured to a maximum of 3 decimal places and the tightest tolerance is +/- .005".
 
Elsmar Forum Sponsor
A

alspread

#2
Re: Calibration of Pin Gages

I don’t have a specific answer to your question, but I had a similar problem and found a solution that was acceptable to my auditor and customers without putting too much burden on the organization.
Basically, the fundamental problem I had was that is was nearly impossible to identify individual pin gages which would be necessary to perform any kind of meaningful calibration. Add to this the problem that the pins would invariably “float” around the shop from set to set with little ability to control.
I had multiple sets of gage pins located in many different areas on the shop floor. These pins were in addition to Go/No Go pins installed in the typical red and green handles. The pins mounted in the handles were used for acceptance or rejection and could be easily identified (labeled) which may calibration easy within the calibration system.
I wrote my internal procedure to specify that loose pins were not to be used for product acceptance and that they must be calibrated prior to use for all applications.
In order to be sure that everyone knew the procedure for pin calibration, I laminated and posted a copy of the loose pin calibration procedure in the cover of all of the pin set cases. My procedure specified that a calibrated micrometer equipped with a tenths barrel was required and that the pin was to be inspected (i.e. calibrated) 2 places 90 degrees apart at both ends and the middle. All results had to be within +/-.0004 of the stated pin size.
I couldn’t tell you that everyone who used a pin performed this process. But I could tell you that they knew what the requirement was and they understood the importance.

Good Luck
 

Jerry Eldred

Forum Moderator
Super Moderator
#3
Re: Calibration of Pin Gages

I'm not much of a dimensional person (more RF and precision electrical standards), but my only thought would be the uncertainty ratio. Take the various uncertainty factors for the calipers (their tolerance, 1/2 the resolution, repeatability, temperature coefficient, and anything else), and algebraically add them using Root Sum Square method, and multiply by two for K=2 confidence; and compare that with the tolerance of the pin gages. The K=2 combined and expanded uncertainty of the measurement needs to be at least 4 times greater than the tolerance of the pin gage. If it is not, it is not generally acceptable as a method.

However, if you don't have to comply with any ISO standard for these, it is a matter of assessing the risk if you don't have an adequate uncertainty ratio. If it is 1.5 to 1, or 2 to 1, your quality people (or who ever has that authority) would need to make that judgment as to if it is acceptable.

I'm certain there are procedures out there for calibrating pin gages, and I would guess a laser mic or something equivalent would normally be used. Theoretically (and I'll probably stir the pot up on this one), you can calibrate anything any way you want. However, any method you use is only as appropriate as its compliance with what ever your compliance is, what ever your quality system dictates, the end product quality or reliability can tolerate or what ever is acceptable to your customer.

So there are right and wrong ways to calibrate, but the definition of right and wrong is determined by factors above.

I was supervisor in a lab in an FDA regulated environment where every instrument had to be calibrated by the 100% full OEM procedure only, no exceptions. That is not a bad practice. But (and maybe one of my old bosses there may read this - and they know me) there are times when the full OEM procedure is inferior to other methods. They also required full before and after data on every calibration (and I mean EVERY calibration). That is not a bad thing. But there are a lot of instruments where there is no value added by the data (I'll probably stir up the pot on that one as well). Then there is the whole issue of ISO17025 Accredited Cals. Some companies get only accredited cals when they don't need them. I used to do nuclear compliant cals where we had to hand write the mfr, model, serial number cal date and due date on every standard used on a separate form (in addition to what was printed on the certificate. And in hand written before and after data, we had to write in the units (i.e.: mV, VDC, KHz, etc.). I got a call from a nuclear customer and he had to give me a deviation because I forgot to write "mV" after a value, even though it was clearly identified as such elsewhere on that data line.

I could go on literally for hours with stories like those. Bottom line is there are so many details in calibration that your company or your regulatory or compliance body decides for you, that its a moving target.

Is it okay to use calipers to calibrate pin gages? Yes and No. I won't elaborate as it would be a rerun of the above.

I've been very busy at work and I'm a little behind on my posts here. I'll try to do better.
 

JAltmann

Quite Involved in Discussions
#4
Re: Calibration of Pin Gages using a Calibrated Caliper that measures to 4 decimal pl

I would echo Jerry's thoughts on the increased uncertianty ratio. The caliper itself really has an absolute best case uncertianty of .0005". This would be if the uncertianty of the gage blocks and environment used to calibrate it were perfect and that same perfect environment was used to calibrate/ceritfy the gage pins.

Now we shift to reality which puts that caliper at closer to .001" uncertianty of measure. Well that alone puts you at a 10:1 ratio of the final inspection citeria of the product your using these pins to measure.

Typically gage pins are checked with a .0001" or better micrometer. The use of 30 sets of gage pins seems rather high, does the product line require a full set of gage pins? or could reduced size sets be used?

Depending on your industry/company certifications you may be able to use the calipers in your procedures to calibrate these gage pins, but your comapnies risk of have non-conforming product go out the door will be increased. You may also get some pushback from an auditor who does like this tool for calibrating the gage pins.
 
J

Jeff Frost

#5
Re: Calibration of Pin Gages

Caliper cannot be used for calibration of pin gages. Accuracy of the caliper is typical only .001 inch for both digital or dial models unless otherwise specified by manufacturer. You will need to maintain 4:1 accuracy ratio to the pin gages (accuracy of .0002 to .00002 based on class) which will mandate use of bench mic at a minimum.
 
S

Sturmkind

#6
Re: Calibration of Pin Gages using a Calibrated Caliper that measures to 4 decimal pl

Hi, DAHSR!

Sounds like an acceptable plan to me. Just a caution is that if the calipers are digital the 4th decimal place will either be a zero or a 5 though that may not be significant considering the tolerance range you are using the pins for. My personal preference would be a 0.0001" OD micrometer with friction thimble which I believe would be more likely to pick up any burrs and may more accurately pick up an undersize condition at the pin tips.

Also, make sure there is no wear on the jaws and light can't be seen through them when fully closed, but I suspect you already check that as standard practice.

Best of luck!
 
J
#7
Re: Calibration of Pin Gages using a Calibrated Caliper that measures to 4 decimal pl

I don't know of any specific requirement that would prevent you from doing as you propose, but personally I can't think of a circumstance where I would be able to justify it to myself, let alone an auditor or customer.

I echo the others here that a caliper is absolutely the wrong instrument to use. At the very least you should be using a standard micrometer that measures to 4 decimal places. My personal preference is an old fashoned analog readout - nothing electronic, but that's just me...
I like alspread's plan requiring a floor calibration check before use, particularly since you have so many sets to check. If you have circumstances that allow this, it's the way I would go.

Then you can call in each set on a rotating basis and do a followup check for wear and accuracy, replacing individual pins as needed.

Peace
James
 
Last edited by a moderator:
J

Jeff Frost

#8
Re: Calibration of Pin Gages using a Calibrated Caliper that measures to 4 decimal pl

Hi, DAHSR!

Sounds like an acceptable plan to me. Just a caution is that if the calipers are digital the 4th decimal place will either be a zero or a 5 though that may not be significant considering the tolerance range you are using the pins for. My personal preference would be a 0.0001" OD micrometer
Best of luck!
The lowest tolerance pin gage (Class ZZ) is .0002. If you plan to use a OD Mic with accuracy of .0001 you will need to reclassify tolerance of the pins to .001 (10:1) OR .0004 (4:1). This is why you should be using a bench mic with its higher accuracy ratio.
 
D

D.MOTA

#9
Re: Calibration of Pin Gages

Very interesting... my self i have the same problem around the shop we have pins everywhere, are you willing to share that procedure?
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#10
Re: Calibration of Pin Gages using a Calibrated Caliper that measures to 4 decimal pl

There have been many good replies here so I will add just a little.

If the caliper is used, who did the calibration? What is the uncertainty of that calibration from the calibration certificate. That is absolutely critical for calculating the uncertainty for your pins.

The calibration of each size pin also must have the uncertainty calculated, and influences have been listed a bit before my reply. Uncertainty calculations must have both Type A and Type B.

Only once the uncertainty has been calculated for each size pin can the Test Uncertainty Ratio (TUR) be calculated. To achieve a 4:1 as has been mentioned, twice the expanded uncertainty of the calibration of the pin cannot is compared to the tolerance of the pin. This is using two-sided of course.

The liklihood of even a digital caliper being able to achieve that is really small.

Best solution is to negotiate with an accredited calibration provider to cal your pins. Yes it will cost, but the application of any science, done correctly, is not cheap.
 
Thread starter Similar threads Forum Replies Date
N Calibration of Setup Fixtures & Pin Gages General Measurement Device and Calibration Topics 5
D Nonadjustable M&TE - 10 year calibration interval? Pin Gages Calibration Frequency (Interval) 11
J Calibration of Pin Gages - AS9100 Calibration Requirements General Measurement Device and Calibration Topics 23
M Calibration of Pin Gages - What are my options? General Measurement Device and Calibration Topics 6
C Black Oxide Gage Pin Calibration Requirements General Measurement Device and Calibration Topics 4
W Setting up a Zmike for Pin Gage Calibration General Measurement Device and Calibration Topics 3
C ZZ Pin Gage Tolerances and Calibration General Measurement Device and Calibration Topics 2
S Pin Gage Calibration - Pins seem to be "growing" General Measurement Device and Calibration Topics 9
K Pin Gage Use vs. Calibration Frequency Calibration Frequency (Interval) 13
M Gage pin calibration - Actual calibration readings required on the calibration cert General Measurement Device and Calibration Topics 3
C Gage pin calibration procedure for accurately calibrating gage pins General Measurement Device and Calibration Topics 18
B Pin Gage Calibration Procedure - ISO 17025:2005 accreditation requirements? ISO 17025 related Discussions 8
V Calibration of Pin Gauges General Measurement Device and Calibration Topics 2
C Pin Gauge Calibration using a Micrometer - Accuracy & Calibration Level Question General Measurement Device and Calibration Topics 10
C Internal Diameter Measurement and Pin Gage Calibration General Measurement Device and Calibration Topics 6
L Gage Pin Gauge R&R - Our calibration guy said he doesn't know how to do one Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 21
N Seeking Procedure(s) - Methodology: Calibration of Pin Gauges General Measurement Device and Calibration Topics 9
N Pin Gage Sets - Calibration costs - Just buy new ones? General Measurement Device and Calibration Topics 9
J Calibration, should we.... Go-No-Go, thread, pin, and ring gauges QS-9000 - American Automotive Manufacturers Standard 2
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 2
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
Q Quality Technician (calibration / metrology) - Memphis, TN Job Openings, Consulting and Employment Opportunities 1
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 5
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Calibration Equipment Selection General Measurement Device and Calibration Topics 2
D Sample Calibration SOP for ISO 13485 General Measurement Device and Calibration Topics 4
L Calibration of Plug Gauges - Tolerancing General Measurement Device and Calibration Topics 1
L In house calibration - Our CMM's are outside certified can I use them to certify our standards General Measurement Device and Calibration Topics 4
R Standard Filter for Calibration Colorimeter ISO 17025 related Discussions 1
J Documenting a Calibration Interval Extension Calibration Frequency (Interval) 7
F Calibration certificate content - Average Value and True Value ISO 17025 related Discussions 6