Calibration of Power Supplies

#1
Hi All,

The company I am working for recently went through the AS9100 D audit. We got a minor non-conformance because the power supplies we use for testing are not calibrated. We do use a calibrated DMM to verify the voltage input is within the specified tolerance.

Some of my fellow coworkers believe that it should not be calibrated because we use the calibrated DMM as verification while others believe we should. Should the power supplies be calibrated in addition to using a calibrated DMM as verification?

Thank you in advance!
Tony
 
Elsmar Forum Sponsor

outdoorsNW

Involved In Discussions
#2
We do not calibrate our power supplier and use a calibrated meter to check before use because we have found power supplies don't hold calibration over a year. We are AS9100 and have no problems with either our external auditor or aerospace customers.

Assuming people are actually checking before use, I think the AS9100 auditor went too far. Please post the exact wording of the finding in case the auditor poorly wrote the finding and is actually using the problem as evidence of some other violation.
 
#3
OutdoorsNW, Thank you for the response!

We have come to an agreement. We will not be calibrating the power supplies, the wording has been clarified by the auditor. Thank you again for the response.
 
Thread starter Similar threads Forum Replies Date
M Inhouse Calibration of Regulated Power Supplies (RPS) using calibrated Multimeter General Measurement Device and Calibration Topics 7
G ISO 9001:2000 Calibration Procedures for DMM/Oscilloscopes/Power Supplies General Measurement Device and Calibration Topics 12
J Pneumatic & Electrical Torque Tool - Calibration/Verification of 'Power Tools' General Measurement Device and Calibration Topics 2
M Calibration of a Power Meter Model Hm8115-2 (Hameg) General Measurement Device and Calibration Topics 8
T Calibration Laboratory Temperature/Humidity goes out of tolerance due to Power Loss General Measurement Device and Calibration Topics 9
R Power Meter Calibration Accuracy Measurement Uncertainty (MU) 4
R HP8481A Power Sensor Calibration - What standards are normally required? General Measurement Device and Calibration Topics 5
J Agilent Power Sensor Calibration - Need a Calibration Procedure General Measurement Device and Calibration Topics 13
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 2
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 2
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
Q Quality Technician (calibration / metrology) - Memphis, TN Job Openings, Consulting and Employment Opportunities 1
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 5
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Calibration Equipment Selection General Measurement Device and Calibration Topics 2
D Sample Calibration SOP for ISO 13485 General Measurement Device and Calibration Topics 4
L Calibration of Plug Gauges - Tolerancing General Measurement Device and Calibration Topics 1
L In house calibration - Our CMM's are outside certified can I use them to certify our standards General Measurement Device and Calibration Topics 4
R Standard Filter for Calibration Colorimeter ISO 17025 related Discussions 1
J Documenting a Calibration Interval Extension Calibration Frequency (Interval) 7
F Calibration certificate content - Average Value and True Value ISO 17025 related Discussions 6
B Oscilloscope - AC gain calibration General Measurement Device and Calibration Topics 6
G Is it common to do a repeatability check during calibration of an instrument? General Measurement Device and Calibration Topics 5
A Adjustment of Calibration Intervals (Shortening or Lengthen) Calibration Frequency (Interval) 6
A Calibration required by OEM General Measurement Device and Calibration Topics 2
S ISO17020 - Equipment calibration requirement General Measurement Device and Calibration Topics 3
H Can a calibration lab still use ISO 6789:2003 as guide for torque calibrations Other ISO and International Standards and European Regulations 1
H Calibration of InfraRan Vapor Analyzer - where to get calibration for SF6 General Measurement Device and Calibration Topics 2
M How to read a calibration certificate for thermo hygrometer General Measurement Device and Calibration Topics 2
L Calibration Frequency for Surftest SJ-400 General Measurement Device and Calibration Topics 0
C External Laboratories and OEM Calibration IATF 16949 - Automotive Quality Systems Standard 0
V Ammeter calibration - Measuring head (on pic.) problem General Measurement Device and Calibration Topics 1
F How many signatures do we need on calibration certificates? ISO 17025 related Discussions 8
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
A Calibration Certificate Results General Measurement Device and Calibration Topics 3

Similar threads

Top Bottom