#1
We have gas regulators on our furnaces and pressuree gages. Do they need to be calibrated? If so, how?
2. What do we do with air gauges used for testing. Do we need to have them calibrated, and how?
3. If customers calibrate their gages and return them to us, what type of documentation do we need to have on file for QS? They will claibrate them themselves with a special gage.
 
Elsmar Forum Sponsor

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#2
We have gas regulators on our furnaces and pressuree gages. Do they need to be calibrated? If so, how?

Probably don't need to be calibrated. They provide an operating range but don't control the process.

2. What do we do with air gauges used for testing. Do we need to have them calibrated, and how?

Yup - gotta be calibrated. Send them out or back to the manufacturer. Who ever sold you the instrument should be able to guide you to a company which can provide calibration of the instrument. Where is the first cal cert (which comes with it when it was shipped from the manufacturer) from?

3. If customers calibrate their gages and return them to us, what type of documentation do we need to have on file for QS? They will claibrate them themselves with a special gage.

I'd just ask for a statement or something from your customer whereby it is stated that the appartus is customer supplied and calibrated. It should have a cal sticker. If your auditor questions it, s/he can go to your customer and check their files, but they won't.
 
Thread starter Similar threads Forum Replies Date
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
S Pressure Gauge Calibration Uncertainty Calibration and Metrology Software and Hardware 5
R Looking for calibration specification for Fluke 7250 pressure controller General Measurement Device and Calibration Topics 5
D Calibration and Tolerance Ranges of Pressure Gages General Measurement Device and Calibration Topics 8
K ASME BPVC (Boiler & Pressure Vessel Code) Calibration Requirements General Measurement Device and Calibration Topics 6
W Pressure Calibration to psiA (psi Absolute) General Measurement Device and Calibration Topics 8
W Pressure Transducer Calibration Standards General Measurement Device and Calibration Topics 6
N Calibration Certificate Requirements for Pressure Gauges General Measurement Device and Calibration Topics 9
M Manufacturing Equipment Pressure Gages Calibration Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 30
V Analog Pressure Gauge Calibration - smallest resolution practice General Measurement Device and Calibration Topics 12
G Calibration of Temperature, Pressure Displays of Molding Machines General Measurement Device and Calibration Topics 13
C Setting up Pressure Calibration In-House Lab General Measurement Device and Calibration Topics 7
J Pressure Gauge Tolerances after Calibration General Measurement Device and Calibration Topics 5
O Calibration Supplier needed for Digital Temperature/Pressure Controllers General Measurement Device and Calibration Topics 6
C Pressure Transducer - Number of Calibration Test Points General Measurement Device and Calibration Topics 12
B Injection Molding: Pressure, Time, Temperature and Calibration General Measurement Device and Calibration Topics 7
W Pressure Gauge Calibration Procedure needed General Measurement Device and Calibration Topics 3
B Published standards dealing with calibration of pressure and vacuum gauges? General Measurement Device and Calibration Topics 5
V Calibration of Temperature and Pressure Controller General Measurement Device and Calibration Topics 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 24
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3

Similar threads

Top Bottom