Calibration of pressure/air gages

D
#1
We have gas regulators on our furnaces and pressuree gages. Do they need to be calibrated? If so, how?
2. What do we do with air gauges used for testing. Do we need to have them calibrated, and how?
3. If customers calibrate their gages and return them to us, what type of documentation do we need to have on file for QS? They will claibrate them themselves with a special gage.
 
Elsmar Forum Sponsor

Marc

Fully vaccinated are you?
Staff member
Admin
#2
We have gas regulators on our furnaces and pressuree gages. Do they need to be calibrated? If so, how?

Probably don't need to be calibrated. They provide an operating range but don't control the process.

2. What do we do with air gauges used for testing. Do we need to have them calibrated, and how?

Yup - gotta be calibrated. Send them out or back to the manufacturer. Who ever sold you the instrument should be able to guide you to a company which can provide calibration of the instrument. Where is the first cal cert (which comes with it when it was shipped from the manufacturer) from?

3. If customers calibrate their gages and return them to us, what type of documentation do we need to have on file for QS? They will claibrate them themselves with a special gage.

I'd just ask for a statement or something from your customer whereby it is stated that the appartus is customer supplied and calibrated. It should have a cal sticker. If your auditor questions it, s/he can go to your customer and check their files, but they won't.
 
Thread starter Similar threads Forum Replies Date
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
S Pressure Gauge Calibration Uncertainty Calibration and Metrology Software and Hardware 5
R Looking for calibration specification for Fluke 7250 pressure controller General Measurement Device and Calibration Topics 5
D Calibration and Tolerance Ranges of Pressure Gages General Measurement Device and Calibration Topics 8
K ASME BPVC (Boiler & Pressure Vessel Code) Calibration Requirements General Measurement Device and Calibration Topics 6
W Pressure Calibration to psiA (psi Absolute) General Measurement Device and Calibration Topics 8
W Pressure Transducer Calibration Standards General Measurement Device and Calibration Topics 6
N Calibration Certificate Requirements for Pressure Gauges General Measurement Device and Calibration Topics 9
M Manufacturing Equipment Pressure Gages Calibration Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 30
V Analog Pressure Gauge Calibration - smallest resolution practice General Measurement Device and Calibration Topics 12
G Calibration of Temperature, Pressure Displays of Molding Machines General Measurement Device and Calibration Topics 13
C Setting up Pressure Calibration In-House Lab General Measurement Device and Calibration Topics 7
J Pressure Gauge Tolerances after Calibration General Measurement Device and Calibration Topics 5
O Calibration Supplier needed for Digital Temperature/Pressure Controllers General Measurement Device and Calibration Topics 6
C Pressure Transducer - Number of Calibration Test Points General Measurement Device and Calibration Topics 12
B Injection Molding: Pressure, Time, Temperature and Calibration General Measurement Device and Calibration Topics 7
W Pressure Gauge Calibration Procedure needed General Measurement Device and Calibration Topics 3
B Published standards dealing with calibration of pressure and vacuum gauges? General Measurement Device and Calibration Topics 5
V Calibration of Temperature and Pressure Controller General Measurement Device and Calibration Topics 1
L A certificate of conformance used to support a calibration General Measurement Device and Calibration Topics 6
A AS9100 - 7.1.5.2 In-House Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 25
B UV-A and Illuminance meter Calibration Setup General Measurement Device and Calibration Topics 0
S Calibration Related Posters General Measurement Device and Calibration Topics 2
F Quality Control in Electrical Calibration ISO 17025 related Discussions 0
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6

Similar threads

Top Bottom