Recently been involved in conversations regarding the need for calibration of processing equipment - specifically injection molding of medical products.

The careful reading of 21 CFR 820 Sec. 820.70 Production and process controls. Specifically Sec. 820.72 "Inspection, measuring, and test equipment" sub section (b) is where calibration is listed.

Specific reading of the section could indicate only items used for acceptance activities require calibration. The extension of the concept to processing equipment therefore results in the interpretation that calibration of process parameters are unnecessary.

Considering Process Validation; ref (Sec. 820.75 Process validation)
The implication is that the process variables must be controlled "....monitoring and control of process parameters for validated processes to ensure that the specified requirements continue to be met. "
creating a product that meets specification requires knowledge of process parameters and associated controls (e.g. time, temperatures, pressures, chemical, etc.)

Can anyone provide more specific citation of the requirements for process controls to be calibrated?

Thanks
 
Elsmar Forum Sponsor

Mark Meer

Trusted Information Resource
#3
Re: Calibration of processing equipment - 21 CFR 820.70 Production and process contro

Specific reading of the section could indicate only items used for acceptance activities require calibration.
Can you elaborate on how you arrived at this interpretation?

820.72 is under Subpart G—Production and Process Controls, and not Subpart H—Acceptance Activities.

Also, 820.72 only says "is suitable for its intended purposes and is capable of producing valid results", which can cover just about any quality system process, not just acceptance activities.

Finally, remember that process validation is intended "where the results of a process cannot be fully verified by subsequent inspection and test" (820.75(a)).

So, in the case of injection molding, is the output product checked? I imagine they are (e.g. dimensions, durability, any other requirements you've specified...), in which case I think it can be argued that the results of the injection molding are verified by subsequent inspection and test, and hence outside the scope of 820.75.
 
R

RLKESTER

#4
Re: Calibration of processing equipment - 21 CFR 820.70 Production and process contro

Process control elements such as time, temperature pressures, contamination (O2, %RH, etc.) can affect the output of a product in ways that cannot be verified. Consider heat treating, sterilization, chemical processes, etc.

Many injection molded products have contact with patient, medication or drugs. Chemical leechables, biocompatibility, toxicology, stability/aging, material physical properties and sterilization processing can all impact the intended use and approved use of the device.

Processing parameters ( eg. Temperature, material dryness, etc) are known to impact the material physical characteristics.
 
Thread starter Similar threads Forum Replies Date
M Polyurethane Equipment Calibration - Thermoset Polyurethane Processing General Measurement Device and Calibration Topics 2
E AS9100 Utilizing Calibration Equipment during Processing AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 24
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 5
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Calibration Equipment Selection General Measurement Device and Calibration Topics 2
D Sample Calibration SOP for ISO 13485 General Measurement Device and Calibration Topics 4
L Calibration of Plug Gauges - Tolerancing General Measurement Device and Calibration Topics 1
L In house calibration - Our CMM's are outside certified can I use them to certify our standards General Measurement Device and Calibration Topics 4
R Standard Filter for Calibration Colorimeter ISO 17025 related Discussions 1

Similar threads

Top Bottom