Calibration of processing equipment - 21 CFR 820.70 Production and process controls

R

RLKESTER

#1
Recently been involved in conversations regarding the need for calibration of processing equipment - specifically injection molding of medical products.

The careful reading of 21 CFR 820 Sec. 820.70 Production and process controls. Specifically Sec. 820.72 "Inspection, measuring, and test equipment" sub section (b) is where calibration is listed.

Specific reading of the section could indicate only items used for acceptance activities require calibration. The extension of the concept to processing equipment therefore results in the interpretation that calibration of process parameters are unnecessary.

Considering Process Validation; ref (Sec. 820.75 Process validation)
The implication is that the process variables must be controlled "....monitoring and control of process parameters for validated processes to ensure that the specified requirements continue to be met. "
creating a product that meets specification requires knowledge of process parameters and associated controls (e.g. time, temperatures, pressures, chemical, etc.)

Can anyone provide more specific citation of the requirements for process controls to be calibrated?

Thanks
 
Elsmar Forum Sponsor

Mark Meer

Trusted Information Resource
#3
Re: Calibration of processing equipment - 21 CFR 820.70 Production and process contro

Specific reading of the section could indicate only items used for acceptance activities require calibration.
Can you elaborate on how you arrived at this interpretation?

820.72 is under Subpart G—Production and Process Controls, and not Subpart H—Acceptance Activities.

Also, 820.72 only says "is suitable for its intended purposes and is capable of producing valid results", which can cover just about any quality system process, not just acceptance activities.

Finally, remember that process validation is intended "where the results of a process cannot be fully verified by subsequent inspection and test" (820.75(a)).

So, in the case of injection molding, is the output product checked? I imagine they are (e.g. dimensions, durability, any other requirements you've specified...), in which case I think it can be argued that the results of the injection molding are verified by subsequent inspection and test, and hence outside the scope of 820.75.
 
R

RLKESTER

#4
Re: Calibration of processing equipment - 21 CFR 820.70 Production and process contro

Process control elements such as time, temperature pressures, contamination (O2, %RH, etc.) can affect the output of a product in ways that cannot be verified. Consider heat treating, sterilization, chemical processes, etc.

Many injection molded products have contact with patient, medication or drugs. Chemical leechables, biocompatibility, toxicology, stability/aging, material physical properties and sterilization processing can all impact the intended use and approved use of the device.

Processing parameters ( eg. Temperature, material dryness, etc) are known to impact the material physical characteristics.
 
Thread starter Similar threads Forum Replies Date
M Polyurethane Equipment Calibration - Thermoset Polyurethane Processing General Measurement Device and Calibration Topics 2
E AS9100 Utilizing Calibration Equipment during Processing AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
Vader22 As-Found FAILED calibration ISO 14001:2015 Specific Discussions 16
M Gravimetric blender for resin in automotive industry - periodical calibration via external lab with certification Reliability Analysis - Predictions, Testing and Standards 0
N Calibration Method for Scales and Flowmeters Reliability Analysis - Predictions, Testing and Standards 7
T Calibration Tracking ISO 13485:2016 - Medical Device Quality Management Systems 4
Q Considering addition of mass calibration to ISO17025 scope ISO 17025 related Discussions 6
Z Resolution in calibration General Measurement Device and Calibration Topics 1
J Run Down Adapters Calibration Processes General Measurement Device and Calibration Topics 4
S Looking for accredited calibration lab - Onsite up to 650A and 1 kV. General Measurement Device and Calibration Topics 3
CaliperJim Screw Thread Micrometer Anvil Calibration General Measurement Device and Calibration Topics 7
Strange Juan HP 85054B Standard Mechanical Calibration Kit Calibration Procedure General Measurement Device and Calibration Topics 1
H Air Pressure Gage Calibration General Measurement Device and Calibration Topics 6
T Calibration on AVL AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
Sam.F Do heat treatment ovens need calibration? General Measurement Device and Calibration Topics 22
J Selecting reference standards for calibration General Measurement Device and Calibration Topics 3
C IATF 16949:2016 Major NC pertaining to 7.1.5.2.1 Calibration/verification records IATF 16949 - Automotive Quality Systems Standard 15
E Calibration Records during AS9100 Audit AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
John Predmore Calibration Determination decision Calibration Frequency (Interval) 15
J ISO 17025 Calibration Other Medical Device Related Standards 2
A Electronic Weighing scale calibration Calibration Frequency (Interval) 4
M Bourdon gauge calibration procedure failure ISO 17025 related Discussions 3
Ed Panek Calibration question Manufacturing and Related Processes 4
S Calibration/Worn gages Manufacturing and Related Processes 9
A Settings rings calibration General Measurement Device and Calibration Topics 4
R 7.1.5.2 -- Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
C CALIBRATION CERTIFICATE REQUIREMENTS General Measurement Device and Calibration Topics 9
S Calibration Label issues General Measurement Device and Calibration Topics 14
H Calibration / Capability in a PCP FMEA and Control Plans 1
R First Time Managing Calibration and Measurement System. Need Help. General Measurement Device and Calibration Topics 19
M CE marked medical device - do they need UKAS certified calibration provider ISO 17025 related Discussions 4
JDJohnson Micrometer & Caliper Calibration (ISO-9001) General Measurement Device and Calibration Topics 26
B Precision Current Loop for Current Probe Calibration Reliability Analysis - Predictions, Testing and Standards 2
Strange Juan Freezers fridges Incubators Calibration Labels General Measurement Device and Calibration Topics 12
D Tolerance and MPEs for Re-calibration Manufacturing and Related Processes 0
D ISO 9001:2015 Calibration Procedures Manufacturing and Related Processes 16
G Calibration of Angle microscope stage micrometer General Measurement Device and Calibration Topics 1
K END CALIBRATION CANNOT BE PERFORMED Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
C Gage Block Grade 1 - Calibration Measurement Uncertainty (MU) 15
JoeRandom11 Question of Calibration Cycles AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
placebo_master What is the difference between "Fitness for Purpose" and "Calibration/Verification"? General Measurement Device and Calibration Topics 6
G AMS 2750 F Instrument Calibration Results and Recording AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Q Gage calibration in disarray ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
R GearBox Test Bench Calibration General Measurement Device and Calibration Topics 2
Q Internal calibration (stopwatch) IATF 16949 - Automotive Quality Systems Standard 3
D Is calibration really required? IATF 16949 - Automotive Quality Systems Standard 6
B AS9100 Calibration Register AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
N Misc Calibration Standard Inquiry Various Other Specifications, Standards, and related Requirements 5
G IATF 7.1.5.2.1 Calibration/verification records :Program/software verification IATF 16949 - Automotive Quality Systems Standard 7

Similar threads

Top Bottom