Ryan Wilde said:
Yes, your master sample should be a known sample. They are known as "Certified Reference Materials", and they are generally available from your national lab, as well as several private companies.
Ryan
Ryan, for many spectrometer types, that is true, but not for all. For XRF, to check the spectrometer for "drift" (or standardization check), all that is required is a sample with the desired elements in sufficient quanity, in a form that is "permanent". All you're doing is checking instrument count rates for drift from Day Zero, so if the check sample is unchanging, it will do fine without ANY certified #'s associated with it. That being said, it is still a really good idea to check the calibration curves with a CRM of the type analyzed by that set of curves.
Daily/shift/hourly correction for instrument drift is a must, but I think it's a misnomer to call it calibration with a spectrometer. It may or may not involve the use of CRM's; depends on the spectrometer type and manufacturer instructions. Calibration and recalibration (or calibration update or method standardization) DOES involve CRM's.
I just got through this subject with a supplier profile from one of our customers, who is a major Big 3 supplier. They wanted a copy of a certificate of calibration (by external calibration source) on a piece of equipment in our process that affects their product. They wanted a spectrometer calibration certificate. I couldn't give them one, because we don't get an official certificate of calibration from the manufacturer for any of our 2 different spectrometer types (XRF and ICP). When they come in to PM 2/year, they give a PM report, but they have no way to "calibrate" the spectrometer in the way a vendor like Midwest Balance would calibrate an analytical balance for us. The spectrometer vendors can't sit there and PM an instrument and then totally redo each set of calibration curves a customer has, and then run QC checks with a CRM.
Well that's what I told the customer, and our registrar, and they seem to be okay with that
qsmso, I'd be interested to hear what type of spectrometer(s) you are having to deal with. We investigated the proper route needed for drift correction/"calibration check", put it in our QMS, and then folowed it, with proper records, and the registrar is more than satisfied with this. That should be all that is necessary.
Joe