Hello friends,
I got confused with the calibration room in my company. They said to me during audit that they cannot calibrate The spectrometer use to analysis chemical composition of material. What they do is to buy master sample and use it as reference point when running the equipement once a month.

Is this correct concept for calibration? If not, what should they do to comply with the calibration concept?

Thank you for your feed back.

QSMSO
 
Elsmar Forum Sponsor

SteelMaiden

Super Moderator
Super Moderator
#2
That's basically it, you don't really calibrate it, you standardize the readings. Now, that is a fine line to walk, but if you make sure that your records are good, you should be ok. We standardize once per shift.

hint: teach everyone to avoid the "C" word whenever possible.
 
C

Craig H.

#3
qsmso

I am sure that there are more qualified people who will chime in here, but let me take a shot.

One suggestion would be to see if the "owner's manual" is available, either at your plant or from the manufacturer. It should have some advice on working calibration.

That said, we use a similar "master sample" system weekly on our most important QC tests, just to confirm that our testing is OK. It has worked well. The Western Electric handbook has advice on how to use this method with XMR charts to maintain testing systems, and it is a valid approach.

Hope this helps, and good luck.
 
B

Bill Ryan - 2007

#4
I agree with Steelmaiden - we call it "profiling of spectrometer" and we "profile" it once per day (minimum) against certified standards (we alloy our own aluminum).

Just curious - what kinds of Gage R&R numbers has anyone seen/submitted for a spectrometer. Mine have been anywhere in the range from 7.5-145% of tolerance, depending on the element being checked. With respect to anything over 25%, we have been able to "correct" the spectrometer and/or operator methodology to have satisfactory numbers to submit.

Bill
 
C

Cyndy B

#5
I have had 2 different experiences with Spectrophotometers. The first was in Blood Banking. The monthly maintenance required that a known set of standards be read by the spectrophotometer in 3 categories: Repeatability (20 times), Linearity and Drift (performance over a predetermined amount of time). These standards were supplied by the manufacturer and NEVER considered a calibration. They were an indicator of lamp stability and performance. The equipment went through a seperate validation prior to use.

The second is in the Chem lab I manage. We don't have a predetermined, colormetric standard to performance check the spec but we do run 5 standards (purchased material diluted to the appropriate concentration) with each assay performed. The readings are used to calculate a curve. The tested samples, of course, fall within this curve. This is done with each assay type being performed. Again, not a "calibration", just a performance check of how the lamp is reading and how well the material being assayed is extracted by the Chemist.

I don't think you can actually calibrate a spec unless there are some filters you can manipulate.

Hope I don't sound like a complete dolt. Have a good one.

:bonk:
 
Q

qsmso

#6
Thank you for all responses. However, the master sample, that used as reference standard, needs to be calibrated or not? I am considering how to track to national standard or equivalent as 4.11.2.b(QS-9000)
The reason I asked because when I audit the calibration room, they use gage blocks to do inside calibrate of vernier gages. Those gage blocks were sent out to calibrate periodically. They can show certificate result when we need to track. So, I think they should use the same concept with master sample of spectrometer. However, they said very difficult.

PLEASE advice !
QSMSO:confused:
 
S

sgray

#7
I was just wondering what type of spectrometer you are talking about. For analysis of chemical composition of a material it could be several instruments. Is it an Emission spectrometer or XRF spectrometer? We happen to use both of these. We have standards for both instruments that we use to make calibration curves with. Due to the type of alloys we make we could only buy a couple certified standards. We cast the rest of the standards ourselves and then analyzed them on an AA. We ran at least one of the certified standards on the AA at the same time we analyzed our own standards and this gives us tracibility. The instruments are then drift corrected, (Standardized), on a regular basis and a known sample is then run as a QA/Check standard to verify the instrument was drift corrected properly.

Hope this helps.
 

Joe Cruse

Mopar or No Car
#8
I'll echo sgray's and Steel Maiden's comments here. Depending on what kind of spectrometer you're running, the term calibration may not be "useful". We use XRF, and calibration means setting up an analytical program for a material using certified reference materials to make a set of calibration curves. This process is "calibration", and is done one time, unless a major change takes place with the spectrometer. Day to day "drift" is corrected for by another procedure, which is NOT calibration. A check for instrument drift is also performed, but this is not a calibration check, per se, but a drift check. This is done per the manufacturer's recommendations, and we set up tolerances in the software to tell us when a drift correction must be performed. We do run certified reference samples for daily QC checks, along with the daily drift check.

We do something very similar with our ICP. The instrument is "profiled" daily with a standard solution, and each method used is checked with a QC standard and standardized, if outside the specified tolerances.

Don't let an auditor paint you in a corner with the term "calibration" when it comes to spectrometers. Our previous lab manager let the auditor make him set up a calibration check/correction that was totally inappropriate for XRF (dealing with pressed powders), and would actually have done the opposite of the intended end, most of the time. Always talk with your instrument maufacturer on this before setting up a "calibration" procedure for a spectrometer.

Bill, I've never submitted an R&R for the spectrometers. The manufacturer suggested calculating 3 SD after about 50 runs of the drift check sample, and setting that for the limits. Regular drift correction should ensure that the spectrometer stays in those limits, barring a catastrophic failure of a part (which can be replaced and then drift correction comes back in.) We've seen this work pretty well on our main spectrometer after implementing it.

Joe
 
R

Ryan Wilde

#9
qsmso said:

Thank you for all responses. However, the master sample, that used as reference standard, needs to be calibrated or not? I am considering how to track to national standard or equivalent as 4.11.2.b(QS-9000)
The reason I asked because when I audit the calibration room, they use gage blocks to do inside calibrate of vernier gages. Those gage blocks were sent out to calibrate periodically. They can show certificate result when we need to track. So, I think they should use the same concept with master sample of spectrometer. However, they said very difficult.

PLEASE advice !
QSMSO:confused:
Yes, your master sample should be a known sample. They are known as "Certified Reference Materials", and they are generally available from your national lab, as well as several private companies.

Ryan
 

Joe Cruse

Mopar or No Car
#10
Ryan Wilde said:

Yes, your master sample should be a known sample. They are known as "Certified Reference Materials", and they are generally available from your national lab, as well as several private companies.

Ryan
Ryan, for many spectrometer types, that is true, but not for all. For XRF, to check the spectrometer for "drift" (or standardization check), all that is required is a sample with the desired elements in sufficient quanity, in a form that is "permanent". All you're doing is checking instrument count rates for drift from Day Zero, so if the check sample is unchanging, it will do fine without ANY certified #'s associated with it. That being said, it is still a really good idea to check the calibration curves with a CRM of the type analyzed by that set of curves.

Daily/shift/hourly correction for instrument drift is a must, but I think it's a misnomer to call it calibration with a spectrometer. It may or may not involve the use of CRM's; depends on the spectrometer type and manufacturer instructions. Calibration and recalibration (or calibration update or method standardization) DOES involve CRM's.

I just got through this subject with a supplier profile from one of our customers, who is a major Big 3 supplier. They wanted a copy of a certificate of calibration (by external calibration source) on a piece of equipment in our process that affects their product. They wanted a spectrometer calibration certificate. I couldn't give them one, because we don't get an official certificate of calibration from the manufacturer for any of our 2 different spectrometer types (XRF and ICP). When they come in to PM 2/year, they give a PM report, but they have no way to "calibrate" the spectrometer in the way a vendor like Midwest Balance would calibrate an analytical balance for us. The spectrometer vendors can't sit there and PM an instrument and then totally redo each set of calibration curves a customer has, and then run QC checks with a CRM.

Well that's what I told the customer, and our registrar, and they seem to be okay with that :)

qsmso, I'd be interested to hear what type of spectrometer(s) you are having to deal with. We investigated the proper route needed for drift correction/"calibration check", put it in our QMS, and then folowed it, with proper records, and the registrar is more than satisfied with this. That should be all that is necessary.

Joe
 
Thread starter Similar threads Forum Replies Date
R Optical Emission Spectrometer Calibration General Measurement Device and Calibration Topics 0
T Optical Emission Spectrometer Calibration to a National or International Standard Various Other Specifications, Standards, and related Requirements 4
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
Q Quality Technician (calibration / metrology) - Memphis, TN Job Openings, Consulting and Employment Opportunities 1
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 5
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Calibration Equipment Selection General Measurement Device and Calibration Topics 2
D Sample Calibration SOP for ISO 13485 General Measurement Device and Calibration Topics 4
L Calibration of Plug Gauges - Tolerancing General Measurement Device and Calibration Topics 1
L In house calibration - Our CMM's are outside certified can I use them to certify our standards General Measurement Device and Calibration Topics 4
R Standard Filter for Calibration Colorimeter ISO 17025 related Discussions 1
J Documenting a Calibration Interval Extension Calibration Frequency (Interval) 7
F Calibration certificate content - Average Value and True Value ISO 17025 related Discussions 6
B Oscilloscope - AC gain calibration General Measurement Device and Calibration Topics 6
G Is it common to do a repeatability check during calibration of an instrument? General Measurement Device and Calibration Topics 5
A Adjustment of Calibration Intervals (Shortening or Lengthen) Calibration Frequency (Interval) 6
A Calibration required by OEM General Measurement Device and Calibration Topics 2
S ISO17020 - Equipment calibration requirement General Measurement Device and Calibration Topics 3
H Can a calibration lab still use ISO 6789:2003 as guide for torque calibrations Other ISO and International Standards and European Regulations 1
H Calibration of InfraRan Vapor Analyzer - where to get calibration for SF6 General Measurement Device and Calibration Topics 2
M How to read a calibration certificate for thermo hygrometer General Measurement Device and Calibration Topics 2
L Calibration Frequency for Surftest SJ-400 General Measurement Device and Calibration Topics 0
C External Laboratories and OEM Calibration IATF 16949 - Automotive Quality Systems Standard 0
V Ammeter calibration - Measuring head (on pic.) problem General Measurement Device and Calibration Topics 1
F How many signatures do we need on calibration certificates? ISO 17025 related Discussions 8
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
A Calibration Certificate Results General Measurement Device and Calibration Topics 3
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
G Third party auditor mentions no grace period for calibration Calibration Frequency (Interval) 22
E Calibration Certificate Content Requirements General Measurement Device and Calibration Topics 3
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
Nicole Desouza AS9100D - Calibration of Dies & Tooling - Plastic Extrusion AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
D Determining Calibration Frequency schedule for items used in production Manufacturing and Related Processes 2
WEAVER Electronic Weighing Scale Calibration Tolerance Manufacturing and Related Processes 1

Similar threads

Top Bottom