Calibration Of Stop Watch using the NIST Clock

C

Calibrator24

#1
Is it acceptable to use the NIST Clock or a web-synchronized stop watch software as the reference to claibrate stop watch or clocks?

How can we demonstrate the traceability if this is acceptable?
 
Elsmar Forum Sponsor

BradM

Staff member
Admin
#2
Re: Calibration Of Stop Watch

Yes, you can. Here is a N.I.S.T. document that can give you some guidance:

http://tf.boulder.nist.gov/general/pdf/2281.pdf

In essence, it will boil down to what tolerance/ uncertainty you are trying to test against. Notice the uncertainties listed for calling time using a land line. You will need to assure adequate ratios. :)

NOTE: Calling N.I.S.T. times listed in the above document are standards.
 
C

Calibrator24

#3
Re: Calibration Of Stop Watch

Yes, you can. Here is a N.I.S.T. document that can give you some guidance:


In essence, it will boil down to what tolerance/ uncertainty you are trying to test against. Notice the uncertainties listed for calling time using a land line. You will need to assure adequate ratios. :)

NOTE: Calling N.I.S.T. times listed in the above document are standards.
We don't need a highly accurate reference standard for our stopwatch. We even use a ±2 seconds tolerance when we compared them with our reference timers.

Normally, what would be the uncertainty of the NIST sync stop watches in the net?
 

BradM

Staff member
Admin
#4
Re: Calibration Of Stop Watch

We don't need a highly accurate reference standard for our stopwatch. We even use a ±2 seconds tolerance when we compared them with our reference timers.

Normally, what would be the uncertainty of the NIST sync stop watches in the net?
Hello there!:bigwave:

What I would do is just call on the telephone for verification. The best uncertainty N.I.S.T. from their testing is 500ms (listed in the link). So that you would give a 4 to 1 T.U.R., which should be sufficient.
 
B

Betty Murphy

#5
Hello! Is there a relatively quick way to calibrate stopwatches that does not require special equipment? My team of observers will be timing various air traffic controller activities (one timer per controller, unfortunately), and we know about the uncertainty introduced by variations in human response time. For purposes of evaluating new software, we want to time to the nearest second, and not carry our reported times out to decimal places. Any advice would be appreciated. Thanks!
 
Thread starter Similar threads Forum Replies Date
Marc Stop Watch Calibration Reality Check General Measurement Device and Calibration Topics 22
P Regulations concerning Calibration of Stop Watches used for Testing Medical Devices 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 24
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 5
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Calibration Equipment Selection General Measurement Device and Calibration Topics 2
D Sample Calibration SOP for ISO 13485 General Measurement Device and Calibration Topics 4
L Calibration of Plug Gauges - Tolerancing General Measurement Device and Calibration Topics 1
L In house calibration - Our CMM's are outside certified can I use them to certify our standards General Measurement Device and Calibration Topics 4
R Standard Filter for Calibration Colorimeter ISO 17025 related Discussions 1

Similar threads

Top Bottom