does anyone know how to calibrate torque wrenches and screwdrivers using Torque mate 200 and what other accesories are needed?
Jerry Eldred
I don't know that particular model. You have to have the correct fitting adapters for the specific torque wrench or screwdriver, and if there is a spanner handle extension, you have to use the correct one, as hanbdle length impacts accuracy.
Calibrating using a digital calibrator is quite straightforward. You have to have the calibrator rigidly mounted on a good work bench. Set each unit being cal'd to desired settings and slowly and evenly check it clockwise and counter-clockwise. Set the calibrator on peak hold and reset after each reading. Repeat each reading a number of times in each direction to assure repeatability. Some people like to use loaders which apply an even smooth pressure to the units until they click over.
That's the short answer. Let me know if I can help with any further level of detail on this.
Al Dyer
Jerry,
I'd appreciate it if you would go into further detail. We use a fixture that utilizes a torque wrench to hold the part stable for further measurements and have noticed alot of variation related to the torque. Anywhere from 3-15% variation on a simple R&R.
Thanks,
ASD...
There's an ANSI standard B107.14M-1994 that may be helpful. You can get it from ANSI or you might try Global.
J Calibration Merry Go Round - Torque Wrenches and Testers Calibration Frequency (Interval) 13
H Can a calibration lab still use ISO 6789:2003 as guide for torque calibrations Other ISO and International Standards and European Regulations 1
J Pneumatic & Electrical Torque Tool - Calibration/Verification of 'Power Tools' General Measurement Device and Calibration Topics 2
D Torque Wrench Calibration Interval Extension with Non-Use? General Measurement Device and Calibration Topics 1
W Torque Wrench Calibration question Service Industry Specific Topics 9
A Torque Wrench Calibration/Verification General Measurement Device and Calibration Topics 24
R Torque Tester Calibration Traceability and Certificate question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
L Pneumatic Torque Wrench Calibration Procedure Calibration and Metrology Software and Hardware 1
MarilynJ6354 Torque Wrench Calibration Frequency (Calibration Interval) Calibration Frequency (Interval) 38
V Calibration procedure for Torque Meter Needed General Measurement Device and Calibration Topics 6
C Norm for Acceptance Calibration Results Torque Gauge and /or Dynamometers General Measurement Device and Calibration Topics 5
Douglas E. Purdy Torque Calibration Accuracy Criteria General Measurement Device and Calibration Topics 7
J Torque Wrench Calibration - Difficult time with repeatable measurements General Measurement Device and Calibration Topics 12
H Calibration uncertainty with torque transducer Measurement Uncertainty (MU) 6
K Calibration method for Torque Wrench & AC gas leakage Tester General Measurement Device and Calibration Topics 4
A TYPE B Uncertainity - Torque standard calibration - How do you get the factors? General Measurement Device and Calibration Topics 1
B Dynamic Torque Transducers: Calibration or Verification? General Measurement Device and Calibration Topics 1
M Calibration System - Finding Product Made with Out of Tolerance Torque Drivers General Measurement Device and Calibration Topics 7
P Calibration of Pneumatic Torque Guns General Measurement Device and Calibration Topics 2
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
W Next Calibration Due Date Calibration Frequency (Interval) 5
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3