Calibration required by OEM
I was curious about user manuals for some assets state “Recalibrate only at OEM”.
If my calibration facility has capability to calibrate an asset but the asset manual requests the calibration goes to OEM, should it go only to the OEM? The asset is a class 2 medical device; if my calibration facility has standard accreditation, can they perform the calibration, or should it always go to OEM?
John Broomfield
This note may be the OEM's attempt to gather data on the performance of their devices. Or they may one of the few organizations that has access to the traceable standards.
Or it may be the OEM trying to keep you up to date of the latest versions of this device. Some OEMs are interested in your continued success.
But from a technical point of view your cal lab my be competent to calibrate this device but you would be expected to provide evidence of this.
dwperron
There are OEM's out there who require calibrations to be sent back to them simply as a money generating tool. Many of them do not make adjustment information available to owners. Unfortunately, there are a number of OEM's who are not good at calibration..... even on their own instruments.
If you are just verifying that your equipment is in tolerance your cal lab may be able to perform that task. Have them supply a copy of typical calibration results so you can compare them to the OEM.
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
M Are model# and Serial # required on a calibration cert? General Measurement Device and Calibration Topics 4
H In-house device calibration record - Actual values obtained required? General Measurement Device and Calibration Topics 9
B Changing out Thermocouple - Is new calibration required? General Measurement Device and Calibration Topics 1
C ISO17025 for In-House Calibration Labs: Is it required? ISO 17025 related Discussions 30
Manix Calibration of Scales to Traceable Standards (Clause 7.6) - Required or Not Calibration Frequency (Interval) 6
A Is NIST traceability required if a Refractometer calibration is done in-house? Calibration Frequency (Interval) 6
S Is calibration of a weigh scale required? Calibration Frequency (Interval) 8
L Is calibration required for weighing scales used for shipping purposes ? Misc. Quality Assurance and Business Systems Related Topics 10
D Functional Test Fixtures - Calibration Required? Quality Manager and Management Related Issues 8
L Are Outside Calibration Companies required to NADCAP Certified? General Measurement Device and Calibration Topics 6
Q Engineering purchased a new Caliper - 7.6 - Initial Calibration Required? General Measurement Device and Calibration Topics 6
H When is Calibration not required? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
K Survey Meter Calibration - Accredited Laboratory required? General Measurement Device and Calibration Topics 5
S Required Complete calibration SOP, calibration Plan and master list template! IATF 16949 - Automotive Quality Systems Standard 3
B CMM Stylus Calibration - Required to record the As Found condition AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
M ISO 17025 - Recommended or Required and other Calibration questions ISO 17025 related Discussions 5
J Equipment required for a Dimensional Calibration Laboratory Document Control Systems, Procedures, Forms and Templates 1
A ISO 17020 Audit Findings - Calibration from Accredited Laboratory required Other ISO and International Standards and European Regulations 8
W Is an Oven I/O/Q Validation Required or is Calibration Sufficient? Qualification and Validation (including 21 CFR Part 11) 7
B Can I Claim That Periodic Calibration is Not Required? ISO 13485:2016 - Medical Device Quality Management Systems 6
R HP8481A Power Sensor Calibration - What standards are normally required? General Measurement Device and Calibration Topics 5
D TS 16949 In-House Calibration Requirements - Certification Required? IATF 16949 - Automotive Quality Systems Standard 26
L Is a Supplier Audit required for an ISO certified calibration company? Process Audits and Layered Process Audits 12
M Gage pin calibration - Actual calibration readings required on the calibration cert General Measurement Device and Calibration Topics 3
A "Calibration Not Required" label requirements ISO 13485:2016 - Medical Device Quality Management Systems 36
P Are we required to register to REACH? Small CMM service and calibration company RoHS, REACH, ELV, IMDS and Restricted Substances 3
E Recalculation of MU for Test Equipment after every Calibration required? Measurement Uncertainty (MU) 11
V Gage Calibration Stickers in harsh environments - Are Calibration Stickers Required General Measurement Device and Calibration Topics 14
A Is Calibration Required for Equipment/Devices Used for Measuring Support Services General Measurement Device and Calibration Topics 6
F What are the specifications required in the preparation of a calibration laboratory? General Measurement Device and Calibration Topics 4
S Calibration required? We buy part of our machined parts - Supplier gage calibration? General Measurement Device and Calibration Topics 6
C What type of Calibration is Required for TS16949? IATF 16949 - Automotive Quality Systems Standard 5
C Thermohygrometer Calibration - Number of Required Calibration Points General Measurement Device and Calibration Topics 13
A Are ISO 9001 certified companies required to use ISO17025 facilities for calibration? ISO 17025 related Discussions 13
S Is official MSA training required person performing calibration? IATF 16949 - Automotive Quality Systems Standard 6
B Weight class required for scale calibration General Measurement Device and Calibration Topics 2
D ISO/IEC 17025 required for a calibration lab to be ISO certified? ISO 17025 related Discussions 6
D Interesting Discussion Using "Not A Calibrated Device" or "No Calibration Required" stickers General Measurement Device and Calibration Topics 63
A Calibration Discrimination - Acceptance criteria is not having the accuracy required? General Measurement Device and Calibration Topics 2
K Should QS/TS certified internal test/calibration labs be required to be ISO17025 ISO 17025 related Discussions 16
K Dimensional Lab (CMM Room) vs. Physical Test/Calibration Lab - Same required temp/hum General Measurement Device and Calibration Topics 2
I Calibration Stickers on instruments are NOT required! General Measurement Device and Calibration Topics 28
Marc Required Calibration Certificate Contents General Measurement Device and Calibration Topics 2
D Level of accuracy required for the calibration of glass mercury thermometers General Measurement Device and Calibration Topics 6
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 4