Calibration Requirements after Exceeding the Calibration Due Date

A

azamiruddin

#1
Hi all,

I dont know wether this thread already exist or not, cannot find it. :(

I am in charge of company's calibration lab.

I have a problem that required help from Elsmar Cove members. Within our composites manufacturing process, there is a process that need composites to be cured inside a pressurized oven (@300F++). this oven calibration interval is 30days (customer req.). now it is already exceeding due date at about 31days and it is allowable per our internal procedure (10% extension). calibration of this oven will expired on 31st May. actual expired date was 28th May. so, on 30th May 11.59:59pm and beyond, this oven no longer can be used (on paper). we will not allow them to start any cure after 12.00:00am 31st May.

my question (also asked by production team), IF composites cure start at 9.00pm on 30th May, and finish cure at about 3.00am 31st May, is that cure acceptable or not? :confused:

p/s:
1) Oven = Autoclave we called it here.
2) 10% extension = set by company, i dont quite agree with this number, can i revised it to 25% ?
 
Elsmar Forum Sponsor

v9991

Trusted Information Resource
#2
Re: Calibration after exceeding the due date

my question (also asked by production team), IF composites cure start at 9.00pm on 30th May, and finish cure at about 3.00am 31st May, is that cure acceptable or not?
speaking compliance-''wise'', one must not allow the cure to start at first; however,technically it could be done on ''deviation'' that, batch release will only happen after suitable verification measures of oven-performance as identified/commited steps (using additional probe such as ones used during calibration etc.,additional qc steps etc.,) suitable in deviation-approval.

Now same can be extended for processing multiple batches, depending on the equipment reliability, dependability of methods to verify the product&process performance, (some times we do subject such batches to additional performance studies ...) and risk there-off from any negative fall out such as product failure, customer complaints etc.,

but these practices actually dilute the systems, and has tendency to become a practice; + its not always easy to justify the deviations and does not ad value to aditional steps being implemented..

2) 10% extension = set by company, i dont quite agree with this number, can i revised it to 25% ?
You will find other threads on similar topic (just at the end of page... see "similar threads")

However, it depends on the trends of calibration - failures - breakdowns - RCAs traceable to the performance of given equipment ... Which will determine the feasibility of revising the extension of grace period for re-calibration. (or frequency)
 

Jen Kirley

Quality and Auditing Expert
Staff member
Admin
#3
Re: Calibration after exceeding the due date

I agree that knowing the gauge will fall out of compliance during the process, that gauge ought not be used at start of process. However, if your system allows it a deviation might (depends on customer, industry, standard etc.) be allowed so long as you can verify the product is, in fact conforming even though the gauge is out of calibration cycle.
 

BradM

Staff member
Admin
#4
Re: Calibration after Exceeding the Due Date

You state that you have a customer requirement here.
Exactly what is the requirement?
 
A

azamiruddin

#5
Re: Calibration after Exceeding the Due Date

Thank you for the response.

You state that you have a customer requirement here.
Exactly what is the requirement?
Brad, my end-customer is Boeing, they require us to perform 30 days calibration interval on all machines that perform temperature & pressure composites cure cycle. This figure quite tight for us due to manpower and calibrator (comparator) limitation.
 

BradM

Staff member
Admin
#6
Re: Calibration after Exceeding the Due Date

Thank you for the response.



Brad, my end-customer is Boeing, they require us to perform 30 days calibration interval on all machines that perform temperature & pressure composites cure cycle. This figure quite tight for us due to manpower and calibrator (comparator) limitation.
Well, the reason I ask is that you are kind of bound by customer requirements. Unless, you have an exception in the purchase order or something. Otherwise, anything past 30 days is past due.

I would think that given successful calibration events and if you have monitoring/ recording equipment that could be relaxed. Otherwise I would charge them more. :)
 

EEJen

Involved In Discussions
#7
Calibration schedules are a risk management tool. Expanding by 25% may increase your risk but you need to know how much product is measured with the tool. If the tool is used infrequently and low volume of product then your risk is lower.

My experience with Boeing is that compliance to the schedule is important. If calibration schedule conflicts with delivery schedule, I would request a waiver from them explaining the delivery impact.
 

blueicecube2

Involved In Discussions
#8
as usual, perform an experiment and study data to back up whatever decision made i.e extension of calibration interval.

as the owner of the equipment, it is always good practice to convince yourself the equipment is indeed performing what it is designated to do before convincing others.. hehe

2 cents
 

SpinDr99

Involved In Discussions
#9
I agree with Brad. You are compelled to comply with customer requirement or risk their wrath. This could range from replacing the components where curing went past midnight of day 30, resulting in a CAR, to more stringent oversight, to possibly losing business.

As for establishing the calibration interval, when in doubt, I always defaulted to the manufacturer's recommended interval (usually one year).

The risk I run at my place of business (a contract vehicle maintenance and repair facility), is that we have measurement instruments owned by the customer and the mechanics. Only customer owned measurement instruments are calibrated. Some mechanics insist on a calibration service come to the facility to perform calibration. Then there's the cost factor to my employer. I've brought this up to the Project Manager on several occasions, citing safety risks and the matter is regularly avoided. I could bring this to corporate, but that might have career affecting consequences.
 
Thread starter Similar threads Forum Replies Date
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
E Calibration Certificate Content Requirements General Measurement Device and Calibration Topics 3
E ISO 9001:2015 - Record requirements for out of calibration tool ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 28
G Calibration: Conforming to ISO9001 Requirements General Measurement Device and Calibration Topics 6
S ISO 17025 Requirements for Mobile On-site Calibration ISO 17025 related Discussions 2
eternal_atlas Section 7.8.4 Specific requirements for calibration certificates ISO 17025 related Discussions 2
T In house Calibration? Measurement equipment traceability requirements - ISO9001 General Measurement Device and Calibration Topics 12
J Calibration Justification - Requirements for Measurement with a Calibrated Device General Measurement Device and Calibration Topics 1
E IATF 16949 Cl. 7.1.5.2.1 - Calibration and Verification Records Requirements IATF 16949 - Automotive Quality Systems Standard 1
B IATF 16949 - 7.1.5.3.1 Internal calibration laboratory requirements IATF 16949 - Automotive Quality Systems Standard 9
S Clarification of threaded ring gage calibration procedure/requirements General Measurement Device and Calibration Topics 2
M Calibration Records Requirements for all the Calibrations and Verifications we do General Measurement Device and Calibration Topics 4
D AMS 2750E - Digital Instrument Calibration Accuracy Requirements General Measurement Device and Calibration Topics 1
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
T ISO 9001 Calibration Requirements General Measurement Device and Calibration Topics 27
Raffy ISO 13485 NIST Traceability requirements for calibration of test equipment ISO 13485:2016 - Medical Device Quality Management Systems 3
R AS9100 Calibration Requirements General Measurement Device and Calibration Topics 4
C Black Oxide Gage Pin Calibration Requirements General Measurement Device and Calibration Topics 4
M ISO17025 Testing Laboratory Equipment Calibration Requirements ISO 17025 related Discussions 4
P ISO 9001 Requirements for ISO 17025 Accredited Calibration Laboratory ISO 17025 related Discussions 5
R TUS Sensors in AMS2750 Calibration Requirements Various Other Specifications, Standards, and related Requirements 12
T Gage Block Calibration Requirements question General Measurement Device and Calibration Topics 19
D Calibration requirements for New Gages General Measurement Device and Calibration Topics 11
S Auditing TS 16949 Cl. 7.6.3.1 - Internal Calibration Laboratory Requirements IATF 16949 - Automotive Quality Systems Standard 8
P Clause 5.10.4.3 - Calibration Report Requirements General Measurement Device and Calibration Topics 1
D Calibration Requirements for AS9100 Rev C Calibration Frequency (Interval) 5
L Flat Surface Granite Table Calibration Requirements Calibration Frequency (Interval) 16
M ISO 13485 Calibration Supplier Evaluation Requirements question Other ISO and International Standards and European Regulations 2
K Calibration Requirements for Thread and Ring Gages AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 6
W Calibration and Measuring Device Labeling Requirements Quality Manager and Management Related Issues 4
K ASME BPVC (Boiler & Pressure Vessel Code) Calibration Requirements General Measurement Device and Calibration Topics 6
N ISO 17025 Requirements for Environmental Conditions for In-House Calibration ISO 17025 related Discussions 5
S Purchase Order requirements Gage Blocks sent out for Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
R Calibration Requirements for Employee Owned Gages General Measurement Device and Calibration Topics 5
B Mecury-in-Glass Thermometer Calibration Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S AMS 2750 - Instrumentation Calibration Requirements Manufacturing and Related Processes 12
S Calibration/Certification requirements for an Incoming Inspection Fixture Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
Y Calibration Frequency Requirements of Liquid in Glass Thermometer General Measurement Device and Calibration Topics 2
C 21 CFR Part 820 - Question About Calibration Procedure Content Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
P Resolution Requirements on Calibration Certificates General Measurement Device and Calibration Topics 8
C Micrometer Parallelism and Flatness Calibration Requirements and Specs General Measurement Device and Calibration Topics 10
I ADR (Automatic Data Recorder) Calibration Requirements Calibration and Metrology Software and Hardware 2
R ISO 17025 M&TE (Measuring & Test Equipment) Calibration Requirements ISO 17025 related Discussions 4
T ISO 17025 Clause 5.4.2 - Calibration and Test Method Requirements ISO 17025 related Discussions 6
L Selection Requirements for Non-Manufacturer Calibration Laboratory General Measurement Device and Calibration Topics 4
A Identification of Calibration Status - ISO 9001 Clause 7.6c Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
B Calibration Requirements in Organizations Providing Engineering Services General Measurement Device and Calibration Topics 5
J AS9100C Calibration System Compliance Requirements General Measurement Device and Calibration Topics 8
E AS9100 Equipment Calibration Recall Methods and Requirements AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 5