Calibration Requirements for Production Equipment
My company has been formally registered to IS9001 for 25+years. We are reviewing our maintenance departments calibration program and have found many inconsistencies with their interpretation of the standard. I would appreciate feedback on the following: when is calibration or verification required for machine process settings? For instance, if we have a coating machine that has line speed and oven temperatures that are used to produce Product A and we know these are critical characteristics, do we need to perform verification or calibration on these machine settings? We do end product testing on Product A to ensure that it is made correctly and meets the end customer's specs of thickness and hardness and calibrate the test equipment (ex: micrometer), so is it necessary to perform calibration on the machine's line speed and oven temp or is it up to us based on our risk analysis.
Ninja
Looking for Reality
For production line...it is up to your risk analysis.
For measurements of "does this meet spec?", calibration is required...typically by the customer if nothing else...
I use micrometers for controlling coating thickness in the production line, and I do not calibrate the mic heads...it would be pointless.
I choose to calibrate the line speed...I choose not to calibrate the heaters or the air flow meter.
That said, the gages I use to verify final (dry, cured, ready to ship) coating thickness and other properties are calibrated without question.
ScottK
Not out of the crisis
Hi NatanyaD - welcome to the Cove.
Since you are basing acceptance of product on testing you can make a risk based argument that you don't need to calibrate your line speed and temp indicators.
You should consider how much product is tested in that risk analysis... if I was a customer I would want to see a decent sample size throughout the run tested.
I worked in a place where we did dipped powder coating. Parts were essentially 100% checked coming off the line and QC sampled per a c=0 plan with a tight AQL. We did calibrate the temperature controllers for our own needs as there was a fine line between melting and burning. We did not calibrate the line speed or dwell timers. We did get that questioned by our ISO auditors on occasion but defended it successfully.
Ninja
Looking for Reality
FWIW...I had a real gem as an internal auditor (hated him...but he did a great job and I appreciated it...or appreciated the results, anyway)...
We wrote directly into the procedure document what did and did not require calibration just to shut him up...not required, but easier than fighting with him...once we did that, we were never questioned again by ISO, IATF or customer audits...so it was worthwhile.
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
E Calibration Certificate Content Requirements General Measurement Device and Calibration Topics 3
E ISO 9001:2015 - Record requirements for out of calibration tool ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 28
G Calibration: Conforming to ISO9001 Requirements General Measurement Device and Calibration Topics 6
S ISO 17025 Requirements for Mobile On-site Calibration ISO 17025 related Discussions 2
eternal_atlas Section 7.8.4 Specific requirements for calibration certificates ISO 17025 related Discussions 2
T In house Calibration? Measurement equipment traceability requirements - ISO9001 General Measurement Device and Calibration Topics 12
J Calibration Justification - Requirements for Measurement with a Calibrated Device General Measurement Device and Calibration Topics 1
E IATF 16949 Cl. 7.1.5.2.1 - Calibration and Verification Records Requirements IATF 16949 - Automotive Quality Systems Standard 1
B IATF 16949 - 7.1.5.3.1 Internal calibration laboratory requirements IATF 16949 - Automotive Quality Systems Standard 9
S Clarification of threaded ring gage calibration procedure/requirements General Measurement Device and Calibration Topics 2
M Calibration Records Requirements for all the Calibrations and Verifications we do General Measurement Device and Calibration Topics 4
D AMS 2750E - Digital Instrument Calibration Accuracy Requirements General Measurement Device and Calibration Topics 1
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
T ISO 9001 Calibration Requirements General Measurement Device and Calibration Topics 27
Raffy ISO 13485 NIST Traceability requirements for calibration of test equipment ISO 13485:2016 - Medical Device Quality Management Systems 3
R AS9100 Calibration Requirements General Measurement Device and Calibration Topics 4
C Black Oxide Gage Pin Calibration Requirements General Measurement Device and Calibration Topics 4
M ISO17025 Testing Laboratory Equipment Calibration Requirements ISO 17025 related Discussions 4
P ISO 9001 Requirements for ISO 17025 Accredited Calibration Laboratory ISO 17025 related Discussions 5
R TUS Sensors in AMS2750 Calibration Requirements Various Other Specifications, Standards, and related Requirements 12
T Gage Block Calibration Requirements question General Measurement Device and Calibration Topics 19
D Calibration requirements for New Gages General Measurement Device and Calibration Topics 11
S Auditing TS 16949 Cl. 7.6.3.1 - Internal Calibration Laboratory Requirements IATF 16949 - Automotive Quality Systems Standard 8
P Clause 5.10.4.3 - Calibration Report Requirements General Measurement Device and Calibration Topics 1
D Calibration Requirements for AS9100 Rev C Calibration Frequency (Interval) 5
L Flat Surface Granite Table Calibration Requirements Calibration Frequency (Interval) 16
M ISO 13485 Calibration Supplier Evaluation Requirements question Other ISO and International Standards and European Regulations 2
K Calibration Requirements for Thread and Ring Gages AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
W Calibration and Measuring Device Labeling Requirements Quality Manager and Management Related Issues 4
K ASME BPVC (Boiler & Pressure Vessel Code) Calibration Requirements General Measurement Device and Calibration Topics 6
N ISO 17025 Requirements for Environmental Conditions for In-House Calibration ISO 17025 related Discussions 5
S Purchase Order requirements Gage Blocks sent out for Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
R Calibration Requirements for Employee Owned Gages General Measurement Device and Calibration Topics 5
B Mecury-in-Glass Thermometer Calibration Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S AMS 2750 - Instrumentation Calibration Requirements Manufacturing and Related Processes 12
S Calibration/Certification requirements for an Incoming Inspection Fixture Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
Y Calibration Frequency Requirements of Liquid in Glass Thermometer General Measurement Device and Calibration Topics 2
C 21 CFR Part 820 - Question About Calibration Procedure Content Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
P Resolution Requirements on Calibration Certificates General Measurement Device and Calibration Topics 8
A Calibration Requirements after Exceeding the Calibration Due Date General Measurement Device and Calibration Topics 8
C Micrometer Parallelism and Flatness Calibration Requirements and Specs General Measurement Device and Calibration Topics 10
I ADR (Automatic Data Recorder) Calibration Requirements Calibration and Metrology Software and Hardware 2
R ISO 17025 M&TE (Measuring & Test Equipment) Calibration Requirements ISO 17025 related Discussions 4
T ISO 17025 Clause 5.4.2 - Calibration and Test Method Requirements ISO 17025 related Discussions 6
L Selection Requirements for Non-Manufacturer Calibration Laboratory General Measurement Device and Calibration Topics 4