Calibration Room Environmental Condition Requirements

T

Tavares

#1
Hello All

I want to known if is possibel what kind of environmental conditions must have a calibration room.

Thanks

Tavares
 
Elsmar Forum Sponsor
A

Al Dyer

#2
It depends on what you are calibrating. For simple items I would start with temp. and humidity.

I'm sure other will have some other types.

ASD...
 

Jerry Eldred

Forum Moderator
Super Moderator
#5
You need to review the specifications of the items you calibrate, and assure your calibration laboratory (room) area is environmentally controlled to the spec for the unit(s) with the tightest environmental requirement. If you are doing relatively low accuracy electrical calibrations, the requirements may not be that tight. Some examples of tighter specs I have seen are as narrow as 23 Deg C +/- 1 Deg C with no greater than 1 Deg change per hour and %RH 45% RH +/- 5% RH. That is for a higher accuracy environment. I have also seen as loose as 23 Deg C +/-5 Deg C and %RH Greater than 30 %RH and less than 60 %RH. For low accuracy electrical calibrations, your environmental requirements are likely to be closer to the second set of requirements. But you MUST review the specs of the units you actually use. Whatever the manufacturer specifies for those units is the limits you must maintain. There may be units where the manufacturer does not specify environmental limits. Electrical power plant switchboard meters, for example are designed for a very wide range of environmental conditions. The calibrator you use, however, may have much tighter limits. Review all of the items, and spec to meet the tightest one.
 
D

DICKIE

#6
You might get some help by looking at NCSL recommended practice 14 "Guide to selecting Laboratory environments"
 
B

blueicecube

#7
Am afraid I can't get a copy of NCSL guide. Is there any other standard reference that can help me to guide me to find a reference for environmental condition for electrical lab?

It seems the auditor may need a reference for my selected temperature control.
 

Stijloor

Staff member
Super Moderator
#8
Am afraid I can't get a copy of NCSL guide. Is there any other standard reference that can help me to guide me to find a reference for environmental condition for electrical lab?

It seems the auditor may need a reference for my selected temperature control.
Hello blueicecube,

This is by no means an all-inclusive resource, but look here, and scroll through this book. Something is better than nothing. ;)

Stijloor.
 
F

fabQA

#9
Electrical power plant switchboard meters, for example are designed for a very wide range of environmental conditions. The calibrator you use, however, may have much tighter limits. Review all of the items, and spec to meet the tightest one.
So the operating range of the tool does not mean calibration environment range? I reported an NC on our vendor calibrating our tool beyond the relative humidity allowable documented in our internal spec (calibrated at 55+/-10% RH, vs spec of 43+/-5%RH). The vendor, who is the manufacturer of the tool -- power sensor -- said that this should not be a concern as the power sensor can operate up to 95% RH as stated in the tool's technical spec.
Our head of internal audit thinks that the calibration environment limits should follow the max-min limits of the tool's operating range and asked us to change our calibration spec. I disagreed as there is no guarantee from vendor that it can give us the required accuracy throughout the operating range. Even if the vendor is the manufacturer and is the sole calibrator of this equipment, they should answer in terms of recommended calibration environment and not in terms of the operating range of the tool.

Secondly, does the manufacturer of the power sensor recommend the calibration environment or does the user decide the limits based on usage environment? In the case above, the spec of 43+/-5%RH is based on our usage environment. Should we change our calibration spec to suite the tool's operating range? Because this is what our head of internal audit thinks -- we should modify our spec.
 
M

merrick65

#10
I know this is a little off subject but could anyone recommend a thermostat to keep the lab at a constant temperature. Ex. a thermostat that could be programmed to turn on the heat once the temperature lowers to 66 and the air conditioning to kick on when the temperature reaches 70. Right now we have an older thermostat that can only be programmed to do only one function; heat or air. This works out well in either the dead of winter or in the summer, we have problems when the seasons change.
 
Thread starter Similar threads Forum Replies Date
P Clean Room Calibration (Validation) Other ISO and International Standards and European Regulations 4
N Setting up a new Calibration Room in a Machine Shop Calibration and Metrology Software and Hardware 3
S Calibration of Foundry Tool Room Equipment General Measurement Device and Calibration Topics 15
E Building a Calibration Room Test Bench - Finding components, test leads, etc. General Measurement Device and Calibration Topics 8
K Dimensional Lab (CMM Room) vs. Physical Test/Calibration Lab - Same required temp/hum General Measurement Device and Calibration Topics 2
Mikey324 External calibration - Finding in our 3rd party audit General Measurement Device and Calibration Topics 54
L A certificate of conformance used to support a calibration General Measurement Device and Calibration Topics 6
A AS9100 - 7.1.5.2 In-House Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 25
B UV-A and Illuminance meter Calibration Setup General Measurement Device and Calibration Topics 0
S Calibration Related Posters General Measurement Device and Calibration Topics 2
F Quality Control in Electrical Calibration ISO 17025 related Discussions 0
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5

Similar threads

Top Bottom