Calibration Source not ISO 17025 accredited

Crusader

Trusted Information Resource
#1
I have a Colorimeter that is sent back to the manufacturer for calibration. The manufacturer states that they "satisfy the requirements of MIL-I-45208A, MIL-STD-45662A, ISO-IED Guide 25, and ANSI/NCSL-Z540-1." They do mention traceability to NIST, though.
The manufacturer has their own laboratory and advertises the calibration services, but no mention of ISO 17025.

It is a specific color meter instrument...I suppose there might be someone else who can check it. Arghhh.

Per ISO 9001 requirements, this isn't good, huh?

What do you recommend?
 
Elsmar Forum Sponsor

Michael_M

Trusted Information Resource
#2
As far as I am aware, calibration only need to be traceable to international or national measurement standards; where no such standards exist, the basis used for calibration or verification shall be recorded. Requiring ISO 17025 would be an internal procedure requirement.

Most gage manufactures that I have seen (not that many) are only compatible to a standard but do not have any certification to the standard. The internal procedures I have to follow do specify "No contracted metrology laboratory is used until it has been qualified to ISO/IEC 17025. Certification will qualify any potential laboratory" but again, this is from an internal procedure.
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#3
There are very specific fields in calibration for which accreditation is generally not sought. This may be one. If you are given the traceability path, then you may likely be stuck. Understand, if the traceability includes the so-called NIST number, you have the full right to request a copy of the report for that number, and match it to the cert provided to you.

Hope this helps.
 

dwperron

Trusted Information Resource
#4
Per ISO 9001 requirements you must have it calibrated traceable to "national measurement standards". If the manufacturer claims "traceability to NIST" then the manufacturer must be able to supply their traceability path to you for their calibrations, ask them to do so. I have had to do that for many customers over the years, it is part of the business of calibration.

You mentioned 17025 - do you have any requirements for 17025 accredited measurements that you use this colorimeter for? That would change everything.
 

1010011010

Starting to get Involved
#5
Hi! We've just had our recertification audit last week and had the same topic, with a different gauge. As the others already pointed out, the calibration only needs to be retraceable to national or international standards, that's totally sufficient for ISO 9001. 17025 is a requirement from ISO/TS 16949, however if there is no accredited calibration lab you "just" need to get your customers approval.

So if none of your customers or any other institution requires this, you should be fine ISO 9001 wise.
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#6
The traceability is the key for the calibration. The traceability path must be supported.

If they give the "NIST Number", then it must be matched to the specific calibration by NIST of THEIR standard. If they do not have the report of NIST calibrating THEIR standard, then there is a broken link and there is no traceability. That is because the "NIST Number" is a work order number, and must be matched to a specific calibration at a specific time in order to be valid for traceability.

Also, they should not include ISO Guide 25 in their verbiage, that standard was officially retired in 1999. That should automatically raise questions to force evidence of traceability.

Hope this helps.
 
#7
Also, they should not include ISO Guide 25 in their verbiage, that standard was officially retired in 1999. That should automatically raise questions to force evidence of traceability.
As should the MIL-I- and MIL-Q standards, since they are obsolete now.
 
Thread starter Similar threads Forum Replies Date
K Ideas for an Open Source Calibration Database Software Calibration and Metrology Software and Hardware 3
E Faro Laser Tracker Calibration Source Needed General Measurement Device and Calibration Topics 5
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
R Calibration of a frequency measurement device/frequency source General Measurement Device and Calibration Topics 2
D Halogen Leak Detector - Calibration source needed General Measurement Device and Calibration Topics 1
R Calibration of Noise Source Instrument General Measurement Device and Calibration Topics 4
S Disappointed with Calibration Source - Different versions of certificates General Measurement Device and Calibration Topics 5
M Internal tool calibration program - Best source to become "knowledgeable"? General Measurement Device and Calibration Topics 5
M Source code and explanation for computer simulation in camera calibration, HELP Calibration and Metrology Software and Hardware 2
J Source for calibration stickers - Narrow "For Reference Use Only" wrap-a-round General Measurement Device and Calibration Topics 4
L A certificate of conformance used to support a calibration General Measurement Device and Calibration Topics 6
A AS9100 - 7.1.5.2 In-House Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 25
B UV-A and Illuminance meter Calibration Setup General Measurement Device and Calibration Topics 0
S Calibration Related Posters General Measurement Device and Calibration Topics 2
F Quality Control in Electrical Calibration ISO 17025 related Discussions 0
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32

Similar threads

Top Bottom