Calibration Source not ISO 17025 accredited
Crusader
I have a Colorimeter that is sent back to the manufacturer for calibration. The manufacturer states that they "satisfy the requirements of MIL-I-45208A, MIL-STD-45662A, ISO-IED Guide 25, and ANSI/NCSL-Z540-1." They do mention traceability to NIST, though.
The manufacturer has their own laboratory and advertises the calibration services, but no mention of ISO 17025.
It is a specific color meter instrument...I suppose there might be someone else who can check it. Arghhh.
Per ISO 9001 requirements, this isn't good, huh?
What do you recommend?
Michael_M
As far as I am aware, calibration only need to be traceable to international or national measurement standards; where no such standards exist, the basis used for calibration or verification shall be recorded. Requiring ISO 17025 would be an internal procedure requirement.
Most gage manufactures that I have seen (not that many) are only compatible to a standard but do not have any certification to the standard. The internal procedures I have to follow do specify "No contracted metrology laboratory is used until it has been qualified to ISO/IEC 17025. Certification will qualify any potential laboratory" but again, this is from an internal procedure.
Hershal
Metrologist-Auditor
There are very specific fields in calibration for which accreditation is generally not sought. This may be one. If you are given the traceability path, then you may likely be stuck. Understand, if the traceability includes the so-called NIST number, you have the full right to request a copy of the report for that number, and match it to the cert provided to you.
Hope this helps.
dwperron
Per ISO 9001 requirements you must have it calibrated traceable to "national measurement standards". If the manufacturer claims "traceability to NIST" then the manufacturer must be able to supply their traceability path to you for their calibrations, ask them to do so. I have had to do that for many customers over the years, it is part of the business of calibration.
You mentioned 17025 - do you have any requirements for 17025 accredited measurements that you use this colorimeter for? That would change everything.
1010011010
Hi! We've just had our recertification audit last week and had the same topic, with a different gauge. As the others already pointed out, the calibration only needs to be retraceable to national or international standards, that's totally sufficient for ISO 9001. 17025 is a requirement from ISO/TS 16949, however if there is no accredited calibration lab you "just" need to get your customers approval.
So if none of your customers or any other institution requires this, you should be fine ISO 9001 wise.
Hershal
Metrologist-Auditor
The traceability is the key for the calibration. The traceability path must be supported.
If they give the "NIST Number", then it must be matched to the specific calibration by NIST of THEIR standard. If they do not have the report of NIST calibrating THEIR standard, then there is a broken link and there is no traceability. That is because the "NIST Number" is a work order number, and must be matched to a specific calibration at a specific time in order to be valid for traceability.
Also, they should not include ISO Guide 25 in their verbiage, that standard was officially retired in 1999. That should automatically raise questions to force evidence of traceability.
Hope this helps.
As should the MIL-I- and MIL-Q standards, since they are obsolete now.
K Ideas for an Open Source Calibration Database Software Calibration and Metrology Software and Hardware 3
E Faro Laser Tracker Calibration Source Needed General Measurement Device and Calibration Topics 5
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
R Calibration of a frequency measurement device/frequency source General Measurement Device and Calibration Topics 2
D Halogen Leak Detector - Calibration source needed General Measurement Device and Calibration Topics 1
R Calibration of Noise Source Instrument General Measurement Device and Calibration Topics 4
S Disappointed with Calibration Source - Different versions of certificates General Measurement Device and Calibration Topics 5
M Internal tool calibration program - Best source to become "knowledgeable"? General Measurement Device and Calibration Topics 5
M Source code and explanation for computer simulation in camera calibration, HELP Calibration and Metrology Software and Hardware 2
J Source for calibration stickers - Narrow "For Reference Use Only" wrap-a-round General Measurement Device and Calibration Topics 4
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 4
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
W Next Calibration Due Date Calibration Frequency (Interval) 5
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0