Calibration Source not ISO 17025 accredited

Crusader

Trusted Information Resource
#1
I have a Colorimeter that is sent back to the manufacturer for calibration. The manufacturer states that they "satisfy the requirements of MIL-I-45208A, MIL-STD-45662A, ISO-IED Guide 25, and ANSI/NCSL-Z540-1." They do mention traceability to NIST, though.
The manufacturer has their own laboratory and advertises the calibration services, but no mention of ISO 17025.

It is a specific color meter instrument...I suppose there might be someone else who can check it. Arghhh.

Per ISO 9001 requirements, this isn't good, huh?

What do you recommend?
 
Elsmar Forum Sponsor

Michael_M

Trusted Information Resource
#2
As far as I am aware, calibration only need to be traceable to international or national measurement standards; where no such standards exist, the basis used for calibration or verification shall be recorded. Requiring ISO 17025 would be an internal procedure requirement.

Most gage manufactures that I have seen (not that many) are only compatible to a standard but do not have any certification to the standard. The internal procedures I have to follow do specify "No contracted metrology laboratory is used until it has been qualified to ISO/IEC 17025. Certification will qualify any potential laboratory" but again, this is from an internal procedure.
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#3
There are very specific fields in calibration for which accreditation is generally not sought. This may be one. If you are given the traceability path, then you may likely be stuck. Understand, if the traceability includes the so-called NIST number, you have the full right to request a copy of the report for that number, and match it to the cert provided to you.

Hope this helps.
 

dwperron

Trusted Information Resource
#4
Per ISO 9001 requirements you must have it calibrated traceable to "national measurement standards". If the manufacturer claims "traceability to NIST" then the manufacturer must be able to supply their traceability path to you for their calibrations, ask them to do so. I have had to do that for many customers over the years, it is part of the business of calibration.

You mentioned 17025 - do you have any requirements for 17025 accredited measurements that you use this colorimeter for? That would change everything.
 

1010011010

Starting to get Involved
#5
Hi! We've just had our recertification audit last week and had the same topic, with a different gauge. As the others already pointed out, the calibration only needs to be retraceable to national or international standards, that's totally sufficient for ISO 9001. 17025 is a requirement from ISO/TS 16949, however if there is no accredited calibration lab you "just" need to get your customers approval.

So if none of your customers or any other institution requires this, you should be fine ISO 9001 wise.
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#6
The traceability is the key for the calibration. The traceability path must be supported.

If they give the "NIST Number", then it must be matched to the specific calibration by NIST of THEIR standard. If they do not have the report of NIST calibrating THEIR standard, then there is a broken link and there is no traceability. That is because the "NIST Number" is a work order number, and must be matched to a specific calibration at a specific time in order to be valid for traceability.

Also, they should not include ISO Guide 25 in their verbiage, that standard was officially retired in 1999. That should automatically raise questions to force evidence of traceability.

Hope this helps.
 
#7
Also, they should not include ISO Guide 25 in their verbiage, that standard was officially retired in 1999. That should automatically raise questions to force evidence of traceability.
As should the MIL-I- and MIL-Q standards, since they are obsolete now.
 
Thread starter Similar threads Forum Replies Date
K Ideas for an Open Source Calibration Database Software Calibration and Metrology Software and Hardware 3
E Faro Laser Tracker Calibration Source Needed General Measurement Device and Calibration Topics 5
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
R Calibration of a frequency measurement device/frequency source General Measurement Device and Calibration Topics 2
D Halogen Leak Detector - Calibration source needed General Measurement Device and Calibration Topics 1
R Calibration of Noise Source Instrument General Measurement Device and Calibration Topics 4
S Disappointed with Calibration Source - Different versions of certificates General Measurement Device and Calibration Topics 5
M Internal tool calibration program - Best source to become "knowledgeable"? General Measurement Device and Calibration Topics 5
M Source code and explanation for computer simulation in camera calibration, HELP Calibration and Metrology Software and Hardware 2
J Source for calibration stickers - Narrow "For Reference Use Only" wrap-a-round General Measurement Device and Calibration Topics 4
A Is calibration of test weight required General Measurement Device and Calibration Topics 1
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Next Calibration Due Date Calibration Frequency (Interval) 5
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
R Calibration lab environmental monitoring General Measurement Device and Calibration Topics 4
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Calibration of pH Meter Probe Calibration and Metrology Software and Hardware 3
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
G Is repeatability required for equipment calibration? General Measurement Device and Calibration Topics 10
was named killer AS 9100D - Calibration Instructions - Controlled documents? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
G Glass scale calibration at 74 degrees F? General Measurement Device and Calibration Topics 4
A Digital Timer Calibration Requirements General Measurement Device and Calibration Topics 3
Ron Rompen Barcode Verifier Calibration and Use Calibration and Metrology Software and Hardware 5
S Is Optical Parallel Set required calibration? ISO 17025 related Discussions 1
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
N Calibration Requirements for Production Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Calibration Equipment Selection General Measurement Device and Calibration Topics 2
D Sample Calibration SOP for ISO 13485 General Measurement Device and Calibration Topics 4
L Calibration of Plug Gauges - Tolerancing General Measurement Device and Calibration Topics 1

Similar threads

Top Bottom