Calibration system must meet ISO 10012, ISO 17025 or ANSI/NCSL Z540?
A
AshleyE
#1
Howdy All,
Need a little clarification on this one.
A customer just changed their supplier requirements and requesting that Calibration systems meet ISO 10012, ISO 17025 or ANSI/NCSL Z4540-1
I'm not familiar with these standards or what the requirements might entail. Can anyone offer guidance of what our customer is looking for, or where I should start? We're an ISO9001:2008 company that performs most measuring equipment verification in house.
Thanks,
Ashley
Elsmar Forum Sponsor
Hershal
Metrologist-Auditor
Staff member
Super Moderator
#2
Howdy All,
Thanks,
Ashley
ISO 10012 is a guidance document that is a nice reference and provides no real direction. No one in the calibration lab world places any actual reliance on it.
ANSI/NCSL Z540-1-1994 is an American National Standard. The Z540-1 is officially retired in 2006, however, it is the typical document met by almost every calibration lab in the US. ANSI/NCSL Z540.3-2006 is the official replacement for Z540-1, however, its acceptance is currently still limited, and required only by a few organizations such as NASA. That will change over time, but it is unclear how long that will take. A lab can self-certify to either version of Z540.
ISO/IEC 17025:2005 is the internationally-recognized accreditation Standard, and must be granted by an accreditation body (AB). Accreditation under ISO/IEC 17025:2005 is considered fully compliant with all applicable requirements of ISO9001:2008 and addresses all requirements of Z540-1 Part 1. As with your ISO9001 registration, an accreditation to ISO/IEC 17025 includes on-site assessment. However, additional steps are required during the assessment, including demonstration of technical competence, and such demonstration requires a metrology professional proficient in the discipline(s) for the scope. Also, the Calibration and Measurement Capability (CMC) must be calculated and listed in the scope.
Hope this helps.
Mikishots
Trusted Information Resource
#3
Howdy All,
Thanks,
Ashley
That's a tall order - it is essentially requiring you to have a certified calibration lab in-house.
It's a big investment; is this customer a significant part of your business? We have also had this situation come up (we're an aircraft manufacturer), but we're simply not in the position to go to those lengths in-house. We'll use ISO 17025 labs when required.
Kales Veggie
People: The Vital Few
#4
I did not read "certified" only "meet", which means self-certify to me.
Kales Veggie
#5
Howdy All,
Thanks,
Ashley
I would read carefully what your customer requires. There is a big difference between the mentioned standards. Also determine is certification by a 3rd party is required or if you can "self-certify" and be compliant.
10012 is the easiest to do. Start doing some research on Elsmar here.
Depending on your instruments and the actual requirement, it maybe more cost effective to calibrate using an outside calibration house.
A
AshleyE
#6
My 30,000 ft view of the requirement is that our internal calibration requirements must be tightened up, or they are looking for all measuring tools to be calibrated by an accredited lab.
Herhsal - Thank you for explaining the differences in the three standards.
Mikishots - That is about what we were thinking. We manufacture components for a variety of industries and this requirement came from an aerospace customer.
Kales Veggie - The actual requirement states "calibration systems shall meet the applicable requirements of..." said standards.
This customer is not a significant amount of our business but we are moving at a glacial pace towards AS9100 and looking to expand our customer base.
Several of our Aerospace customers have pushed down additional requirements recently, not sure if this is just the annual review and updating documents or is something else is kicking around in the industry.
Thanks everyone for your insight. Found a copy of ISO 10012 to start with and move from there.
Mikishots
#7
The requirements wouldn't differ - unless someone's hiding something. The other thing that may be causing confusion is the header - it says "must meet".
I think some clarity from the OP is in order - is certification mandatory under the customer's revised plans? The body says "requesting that Calibration systems meet...."
Kales Veggie
#8
There is a huge cost difference !!!!
A
AshleyE
#9
Customer Purchase Order Conditions state:
Calibration Systems shall meet the applicable requirements of ISO 10012, ISO 17025 or ANSI/NCSL Z540-1.
I'm wondering now, could this be specifically for calibration houses when this company is purchasing calibration services?
Hershal
Metrologist-Auditor
Staff member
Super Moderator
#10
Based on this, your conclusion appears to be likely.
Thread starter Similar threads Forum Replies Date
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate Qualification and Validation (including 21 CFR Part 11) 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate General Measurement Device and Calibration Topics 2
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
W Equipment for Oven Calibration, Uniformity and System Accuracy Tests AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
Gman2 The current state of our Calibration System General Measurement Device and Calibration Topics 28
H Measurement System in Control if the Equipment requires Daily Calibration? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
R Incorporating New Electronic Test Equipment into our Calibration System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
J AS9100C Calibration System Compliance Requirements General Measurement Device and Calibration Topics 8
S QMS (Quality Management System) and Calibration Procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 20
W Calibration Register - Traceability - Data Transfer to a New System General Measurement Device and Calibration Topics 7
3 Vision System Optical Calibration Standard General Measurement Device and Calibration Topics 6
F Calibration Issue - Continuous Temperature Monitoring System was not Acceptable General Measurement Device and Calibration Topics 7
M Druck DPI 620 Advanced Modular Calibration System General Measurement Device and Calibration Topics 1
AnaMariaVR2 Laboratory Equipment Quality System - Calibration, Maintenance and Inspection Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 10
S Do you include Penetremeters used in Radiography in your Calibration System General Measurement Device and Calibration Topics 2
T System Calibration vs. Component Calibration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
R 3 Wire Thread Measuring System - Gage Calibration General Measurement Device and Calibration Topics 8
D Looking for a "User Friendly" calibration system software package Calibration and Metrology Software and Hardware 18
M Constant for thread gage calibration with the three wire system General Measurement Device and Calibration Topics 7
A TS16949 Calibration system - Concerned with Qualification of Administrator. IATF 16949 - Automotive Quality Systems Standard 3
K TS 16949 Calibration System Audit - What questions will they ask me? General Measurement Device and Calibration Topics 11
S Calibration System was In-House now Outside (External Service Provider) General Measurement Device and Calibration Topics 14
J Advice on Handling a Camera Vision System Calibration Master General Measurement Device and Calibration Topics 7
M How to implement a proper Calibration System? General Measurement Device and Calibration Topics 5
C Vision System which verifies the correct part is present - Calibration / Verification Calibration Frequency (Interval) 9
H Employee owned gages not entered in calibration system - AS9100 Clause 7.6 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
F Digital Multimeter - Verification vs. Calibration MSA (Measurement System Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
V Poke Yoke Gage Calibration - Do you have them in your calibration system? General Measurement Device and Calibration Topics 8
D Calibration - Out of Tolerance - Trace Back System General Measurement Device and Calibration Topics 1
M Calibration System - Finding Product Made with Out of Tolerance Torque Drivers General Measurement Device and Calibration Topics 7
J UPS owned shipping carton weight scales - Include this in my Calibration system? General Measurement Device and Calibration Topics 9
J Should weigh count scales be in the Calibration System? QS-9000 - American Automotive Manufacturers Standard 4
J Do Process gauges need to be in Calibration control system? QS-9000 - American Automotive Manufacturers Standard 2
I Calibration Items - Necessary to include the hand crimp tools into the system? General Measurement Device and Calibration Topics 10
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 26
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Similar threads
Top Bottom