Calibration tolerance for data logger

A

asteroid1122

#1
Hellow dears,

I am trying to calibrate Rotronic data loggers Model HL-20 which is used to record temperature and humidity using a Testo brand Huminator II but I am not sure about the calibration tolerance. In other words what is the acceptance criteria in the calibration.



Appreciate your support if you can share any reference in this matter.
 
Elsmar Forum Sponsor

Eredhel

Quality Manager
#2
This is completely out of my area of expertise but a quick google brought up the manual with a calibration section that has this:

"4. Maintenance
[FONT=Arial,Arial][FONT=Arial,Arial]4.1 Periodic calibration check
[/FONT]
[/FONT]
[FONT=Arial,Arial][FONT=Arial,Arial]
[/FONT]
[/FONT]
[FONT=Arial,Arial][FONT=Arial,Arial]Both the Pt 100 RTD temperature sensor and associated electronics are very stable and should not require any calibration after the initial factory adjustment.
Long term stability of the ROTRONIC Hygromer humidity sensor is typically better than 1 %RH per year. For maximum accuracy, calibration of the HL20 should be verified every 6 to 12 months. Applications where the data logger is exposed to significant pollution may require more frequent verifications.
[/FONT]
[/FONT]Note:
[FONT=Arial,Arial][FONT=Arial,Arial]Calibrating and adjusting the data logger typically require a humidity-temperature generator or other specialized equipment that may not be available to all users.
[/FONT]
[/FONT]Procedure for adjusting the data logger with the ROTRONIC HW4 software:
[FONT=Courier New,Courier New][FONT=Courier New,Courier New]o [/FONT][/FONT][FONT=Arial,Arial][FONT=Arial,Arial]Use cable AC3006 to connect the service connector of the data logger to the USB port of a PC with the ROTRONIC HW4 software installed. Note that the ROTRONIC USB driver must be installed on the PC as explained in the HW4 manual [/FONT][/FONT]E-M-HW4v3-Main
[FONT=Courier New,Courier New][FONT=Courier New,Courier New]o [/FONT][/FONT][FONT=Arial,Arial][FONT=Arial,Arial]Start HW4 on the PC and search for the data logger (HW4 Main Menu Bar > Devices and Groups > Search for USB Masters).
[/FONT]
[/FONT][FONT=Courier New,Courier New][FONT=Courier New,Courier New]o [/FONT][/FONT][FONT=Arial,Arial][FONT=Arial,Arial]After finding the data logger with HW4, expand the device tree to see the data logger functions. Select [/FONT][/FONT]Probe Adjustment[FONT=Arial,Arial][FONT=Arial,Arial].
[/FONT]
[/FONT][FONT=Courier New,Courier New][FONT=Courier New,Courier New]o [/FONT][/FONT][FONT=Arial,Arial][FONT=Arial,Arial]For further instructions see HW4 manual [/FONT][/FONT]E-M-HW4v3-A2-001 "

 
A

asteroid1122

#3
Thank you my friend for your response. I have seen the manual but I didnt find anything regarding the calibration tolerance.
 

dwperron

Trusted Information Resource
#4
I am afraid that you are "upside down" on the tolerances of the instruments involved here.

Your Rotronic has temperature accuracy of 0.3°C and humidity accuracy of 0.8% to 60% and 1.3% from 60 to 100%.

The Testo is accurate to 0.2°C, but only accurate to 1.5% RH.

The Rotronic is more accurate than the instrument you are trying to calibrate it with. This is happening more and more these days with humidity calibrations. The good old 4:1 ratio is getting harder to achieve, in cases like this it is almost impossible.

A couple of suggestions would be either to send the Rotronic out for calibration, or to limit its accuracy to something the Testo can handle. I would not use the Testo for anything with a tolerance less that ±3%, because that still only gives you a 2:1 accuracy ratio.
 
Thread starter Similar threads Forum Replies Date
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 3
WEAVER Electronic Weighing Scale Calibration Tolerance Manufacturing and Related Processes 1
I Determining Calibration Tolerance of a Measurement Device General Measurement Device and Calibration Topics 2
G Assigning a calibration tolerance - An x-y coordinate machine - Uncertainty as my verification tolerance General Measurement Device and Calibration Topics 4
R Feeler Gauge Calibration - What tolerance is used for calibrating feeler gauges? General Measurement Device and Calibration Topics 4
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
D Calibration and Tolerance Ranges of Pressure Gages General Measurement Device and Calibration Topics 8
K Calibration Standards - Master weight having more tolerance than our gram scale General Measurement Device and Calibration Topics 3
T Calibration Laboratory Temperature/Humidity goes out of tolerance due to Power Loss General Measurement Device and Calibration Topics 9
H Determining Calibration Tolerance for new Leak and Occlusion Equipment Measurement Uncertainty (MU) 2
C Out of Calibration vs. Out of Tolerance - Calibration Definitions Nonconformance and Corrective Action 7
H Is there ISO 13485 Guidance for Calibration Tolerance? General Measurement Device and Calibration Topics 14
M Out of Tolerance Calibration Report Document Control Systems, Procedures, Forms and Templates 12
M Calibration Problem - Range of Use, Calibration Acceptance Tolerance IATF 16949 - Automotive Quality Systems Standard 11
B Calibration Accuracy vs. Gage/Part Tolerance General Measurement Device and Calibration Topics 4
N Universal/International Symbol for Tolerance and/or Calibration Levels? General Measurement Device and Calibration Topics 7
B Tolerance for Digital Indicator Calibration - Machinists Indicators General Measurement Device and Calibration Topics 1
R Setting Calibration Tolerance on Gauges - Oven Calibration and Stepped Plug Gage General Measurement Device and Calibration Topics 10
F Accuracy of calibration relationship to tolerance General Measurement Device and Calibration Topics 2
B Electric 0-8" Caliper Calibration Tolerance Question. General Measurement Device and Calibration Topics 9
D Calibration - Out of Tolerance - Trace Back System General Measurement Device and Calibration Topics 1
M Calibration System - Finding Product Made with Out of Tolerance Torque Drivers General Measurement Device and Calibration Topics 7
T Torque tool calibration Oil and Gas Industry Standards and Regulations 3
E Calibration-Conditional Pass Measurement Uncertainty (MU) 2
S Calibration Interval calculation for ECG Machine Calibration Frequency (Interval) 4
Mikey324 External calibration - Finding in our 3rd party audit General Measurement Device and Calibration Topics 58
L A certificate of conformance used to support a calibration General Measurement Device and Calibration Topics 6
A AS9100 - 7.1.5.2 In-House Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 25
B UV-A and Illuminance meter Calibration Setup General Measurement Device and Calibration Topics 0
S Calibration Related Posters General Measurement Device and Calibration Topics 2
F Quality Control in Electrical Calibration ISO 17025 related Discussions 0
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11

Similar threads

Top Bottom