H
hugom
Hi,
My lab has a torque transducer+Reading unit used to calibrate torque wrenches.
The last certificate of calibration showed that it is no longer inside its specification (+/-0,3% FS). So I was thinking of correcting the errors through an interpolated polynomial correction (using Excel).
But now I don´t know what should be the uncertainties of my equipament, when I calibrate torque wrenches, in the total uncertainties budget (besides uncertainties associated to resolutions).
Should it be:
Uncertainty associated to the correction of values read by of my transducer+Reading unit
+Uncertainty associated of calibration of my transducer+Reading unit
+Uncertainty associated of drift between calibrations of my transducer+Reading unit
Is it complete ?
Despite the correction, shouldn´t I consider in some way the specified accuracy ot the torque transducer (I´m not even thinking of the specified accuracy ot the reading unit itself) ??
Does anyone have a idea ?
Thank you,
Hugo
My lab has a torque transducer+Reading unit used to calibrate torque wrenches.
The last certificate of calibration showed that it is no longer inside its specification (+/-0,3% FS). So I was thinking of correcting the errors through an interpolated polynomial correction (using Excel).
But now I don´t know what should be the uncertainties of my equipament, when I calibrate torque wrenches, in the total uncertainties budget (besides uncertainties associated to resolutions).
Should it be:
Uncertainty associated to the correction of values read by of my transducer+Reading unit
+Uncertainty associated of calibration of my transducer+Reading unit
+Uncertainty associated of drift between calibrations of my transducer+Reading unit
Is it complete ?
Despite the correction, shouldn´t I consider in some way the specified accuracy ot the torque transducer (I´m not even thinking of the specified accuracy ot the reading unit itself) ??
Does anyone have a idea ?
Thank you,
Hugo
Thank you Graeme and Hershal for your advices.